Electrical impedance tomography applied to assess matching of pulmonary ventilation and perfusion in a porcine experimental model

Abstract

Introduction

Electrical impedance tomography (EIT) can be used to measure impedance changes related to the thoracic content of air and blood. Few studies, however, have utilised EIT to make concurrent measurements of ventilation and perfusion. This experimental study was performed to investigate the feasibility of EIT to describe ventilation/perfusion (V/Q) matching after acute changes of pulmonary perfusion and aeration.

Methods

Six mechanically ventilated, anaesthetised pigs in the supine position were studied at baseline, after inflation of a balloon in the inferior caval vein (Binfl) to reduce cardiac output and after an increased positive end-expiratory pressure (PEEP) of 20 cmH2O (PEEP20) to increase pulmonary aeration. EIT measurements were performed at the mid-thoracic level to measure the amplitude of impedance changes related to ventilation (ZV) and perfusion (ZQ), both globally and in four defined regions of interest (ROI) extending from the ventral to dorsal distance.

Results

A largely parallel distribution of ZV and ZQ in all four ROIs during baseline conditions corresponded to a bell-shaped frequency distribution of ZV/ZQ ratios with only moderate scatter. Binfl and PEEP20 with unchanged tidal volumes significantly increased the mismatch of regional ZV and ZQ, the scatter of ZV/ZQ ratios and the heterogeneity of the ZV/ZQ frequency distribution. Significant positive and negative correlations were demonstrated between fractional alveolar dead space (r2 = 0.63 [regression coefficient]) and venous admixture (r2 = 0.48), respectively, and the global ZV/ZQ ratio.

Conclusions

EIT may be used to monitor the distribution of pulmonary ventilation and perfusion making detailed studies of V/Q matching possible.

Introduction

Electrical impedance tomography (EIT) is a non-invasive, non-radiant, continuous technique that can be used to perform bedside measurements of impedance changes related to the thoracic content of air and blood [1, 2]. Several clinical and experimental studies have demonstrated that EIT can be applied to monitor respiration and tailor the settings of mechanical ventilation [35]. The pulmonary circulation can also be monitored by EIT as reported both in normal conditions and during changes induced by disease or pharmacological interventions [610]. However, simultaneous measurements of pulmonary ventilation and circulation by EIT have previously been only scarcely reported [1113]. The concurrent determination of ventilation-related and perfusion-related changes in impedance is complicated by the two signals differing in amplitude by up to an order of magnitude and being superimposed on each other with variable frequency harmonics.

We recently described how EIT can be used to monitor beat-to-beat changes in global pulmonary perfusion with adequate precision during acute variations in stroke volume related to changes in cardiac preload of pulmonary and extra-pulmonary origin [14]. This was performed by applying short periods of apnoea. The present investigation was designed to assess the matching of ventilation and perfusion, as reflected by the ventilation/perfusion (V/Q) ratio derived from EIT measurements. In addition to global measurements, the left and right lung ventilation and perfusion were studied separately and regionally in order to provide a comprehensive assessment of overall and regional V/Q matching.

The EIT algorithm generates values for impedance changes relative to a baseline measurement in every element of the reconstructed tomographic slice [15]. Significant changes in baseline conditions, for example sudden reductions in blood volume or the occurrence of pneumothorax or haemothorax, may confound the interpretation of impedance changes in absolute terms at different time points [16]. Therefore, the relation and distribution of impedance changes within and between defined regions of interest for any given time point, as utilised in the present experimental design, may provide a more robust estimate of ventilation and circulation in different parts of the lung. The V/Q results may potentially be helpful in optimising ventilator and haemodynamic therapy, taking into account the complex cardiorespiratory interactions, and could significantly extend the clinical applicability of EIT.

The aim of the experiment was to investigate the feasibility of EIT to describe the dynamic distribution of pulmonary ventilation and perfusion at resting conditions and during interventions designed to generate a wide range of V/Q ratios. The application of 20 cmH2O of positive end-expiratory pressure (PEEP) was used to primarily increase pulmonary aeration with secondary reduction of cardiac preload. Inflation of a balloon in the inferior caval vein was used to acutely reduce cardiac preload with minimal effects on pulmonary ventilation.

We hypothesised that combined measurements by EIT of lung ventilation and circulation could be used to assess V/Q matching including the extremes of venous admixture and wasted ventilation. It was further hypothesised that V/Q ratios by EIT would correlate to shunt and dead space fractions calculated using standard formulae.

Materials and methods

Six Swedish Landrace pigs (weighing 32 to 34 kg) were included in the study following approval by the Ethics Committee for Animal Experiments at the University of Göteborg. Animal care conformed to the principles set forth in the Guide for the Care and Use of Laboratory Animals [17]. A limited dataset from two of the animals included in this study has previously been reported [14].

Anaesthesia

After an overnight fast with free access to water, animals were premedicated using ketamine (13 mg/kg intramuscularly) and midazolam (2 mg/kg intramuscularly). Induction of anaesthesia was performed using an intravenous bolus of 10 mg/kg sodium pentobarbital and anaesthesia was then maintained using a combination of sodium pentobarbital (beginning at 25 mg/kg/hour and gradually lowered to 10 mg/kg/hour) and fentanyl (2.5 μg/kg/hour). Animals were ventilated via a tracheostomy with 40% supplemental oxygen in a volume-controlled mode with no end-inspiratory pause with PEEP set at 5 cmH2O (Servo 300, Siemens-Elema, Sweden), I:E ratio 1:2 and a tidal volume of 10 ml/kg resulting in an end-inspiratory pressure less than 35 cmH2O throughout the study. Normocapnia (5.0 ± 0.8 kPa (mean ± standard deviation)) was maintained as gauged by end-tidal capnometry and repeated arterial blood-gas analyses. Isotonic 2.5% glucose was administered intravenously at 10 ml/kg/hour and all animals were kept normothermic (38 to 39°C) using heating blankets.

Preparation

A pulmonary artery catheter (7.5 F Opticath™, Abbott Laboratories, Chicago, IL, USA) was inserted via the right internal jugular vein to monitor cardiac output (as the mean of triplicate thermodilution measurements randomly throughout the respiratory cycle) and to obtain mixed venous blood samples. An arterial catheter was inserted in the left femoral artery to monitor mean arterial pressure and to obtain arterial blood samples. A Fogarty embolectomy catheter (Edwards Lifesciences Services GmbH, THB 32080810F, Unterschleissheim, Germany) was placed via the right femoral vein in the inferior caval vein with its inflatable balloon just below the diaphragm.

Electric impedance tomography

A 16-electrode silicon belt was positioned at the mid-thoracic level and connected to the EIT monitor (EIT Evaluation Kit 2, Dräger Medical, Lübeck, Germany). A tomographic image of impedance changes was created every 100 ms (10 Hz) by the application of an electrical current of 5 mA and 50 kHz to the electrodes. The image was reconstructed in a 32 by 32 pixel matrix. The tomographic slice by EIT is about 5 cm thick [1], thus well in excess of the reported cranio-caudal displacement of the lungs during ventilation [18, 19].

The unfiltered EIT signal was recorded and converted to ASCII files using dedicated EIT software (Draeger EIT Data Review, Dräger Medical, Lübeck, Germany). Four equally sized regions of interest (ROIs) of eight by four pixels each, together extending the full ventral to dorsal distance, were set either along a left or right axis going through the lung fields, taking care not to include the cardiac area (Figure 1). Measurements of impedance changes within the four ROIs thus represented the ventral (ROI 1), mid-ventral (ROI 2), mid-dorsal (ROI 3) and dorsal (ROI 4) parts of either the left or right side. All ASCII files were analysed off-line using custom-made software (MATLAB 7.4.0.287 (R2007a), Mathworks Inc., Natick, MA, USA) to calculate the amplitude of impedance (Z) changes during ongoing ventilation (ZV, five to eight breaths) and during a short apnoea (ZQ, 15 to 25 pulse beats). The reconstruction algorithm for the EIT signal is based on impedance differences in every pixel measured and is thus dimensionless [15]. Both ZV and ZQ were recorded in arbitrary units (AU).

Figure 1
figure 1

Electrical impedance tomography image at end-inspiration illustrating for the right lung the setting of the four regions of interest. Region of interest (ROI) 1 = ventral; ROI 2 = mid-ventral; ROI 3 = mid-dorsal; ROI 4 = dorsal. Note that ROI 1 and ROI 4 encloses relatively minor parts of aerated lung tissue. The medial border of the ROIs was set just lateral to the cardiac contour identified during apnoea.

Experimental design

All animals were investigated in the supine position at steady state during baseline and following either the increase of PEEP from 5 to 20 cmH2O (PEEP20) or the inflation of the Fogarty balloon catheter (Binfl). At each point in the protocol, EIT recordings were made during ongoing ventilation and during a short end-expiratory pause. Arterial and mixed venous blood samples were simultaneously collected and the inspiratory fraction of oxygen and the end-tidal carbon dioxide (AS/3 Anaesthesia Monitor, GE Healthcare, Solna, Sweden) were noted. The arterial and mixed venous blood samples were immediately analysed using an automated blood gas analyser set to analyse pig blood samples (ABL700, Radiometer, Copenhagen, Denmark). The partial pressures of oxygen and carbon dioxide, oxygen saturation and haemoglobin content were entered into the experimental documentation file.

Calculations and statistical analyses

All values are given as the mean and standard deviation. Both ZV and ZQ were multiplied by the respiratory rate and the heart rate, respectively, to provide a common format of AUs per minute. The relative distribution of ZV and ZQ to each ROI was calculated as the proportion of the global, cumulative sum of ROI 1 to 4 for the left and right lung, respectively, and the ZV/ZQ ratio in each ROI was calculated by dividing the relative values for ZV and ZQ. Venous admixture (Qs/Qt) was calculated according to the standard formula based on pulmonary capillary, arterial and mixed venous oxygen contents. Fractional alveolar dead space (VD/VT) was estimated by dividing the arterial to end-tidal carbon dioxide gradient by the arterial carbon dioxide tension [20]. Changes in ZV, ZQ, their distribution and the ZV/ZQ ratios were analysed by paired student's t-test. The correlations between ZV/ZQ ratios by EIT and Qs/Qt and VD/VT were evaluated using Pearson linear correlation, using pooled data from baseline, PEEP20 and Binfl in each animal. Statistical significance was set at a p < 0.05. All statistical analyses were performed using Prism 5 for Mac OSX (GraphPad Software, Inc., San Diego, CA, USA).

Results

Baseline cardiac output was 3.0 ± 0.8 L/minute and decreased to 1.8 ± 0.3 L/minute during PEEP20 and to 1.4 ± 0.3 L/minunte during Binfl, corresponding to stroke volumes of 27 ± 4 mL at baseline, 14 ± 4 mL during PEEP20 and 12 ± 4 mL during Binfl. Mixed venous oxygen saturation decreased from 70 ± 3% at baseline to 41 ± 9% during interventions. Baseline Qs/Qt was 11 ± 2% and decreased to 5 ± 1% during PEEP20, whereas baseline VD/VT increased from 9 ± 4% to 27 ± 11% during Binfl.

The greatest impedance amplitudes for ZV and ZQ at baseline were observed in the mid-ventral region (ROI 2) with no significant differences between the left and right lung. Binfl reduced ZQ in both left and right ROI 2, although an increase was noted in left ROI 1. PEEP20 also decreased ZQ in the left and right ROI 2 and increased ZV in ROI 3.

The relative distribution of ZV and ZQ between the individual ROIs in relation to the total, cumulative signal in ROI 1 to 4 is illustrated in Figures 2 and 3. At baseline conditions, a parallel distribution of ZV and ZQ was observed, with the greatest proportion of ventilation and perfusion appearing in ROI 2 followed by ROI 3 and then ROI 1, while only a minor, although equal, fraction was observed in ROI 4. Binfl decreased the proportion of ZQ distributed to ROI 2 in both the left and right lung and increased the proportional ZQ in the left ROI 1. No changes were observed in the distribution of ZV between the ROI 1 to 4. Hence, ZV increased in relation to ZQ in ROI 2 bilaterally, although the reverse was observed in the left ROI 1 (Figure 2). PEEP20 increased the proportion of ZV in ROI 3 bilaterally while ZV in ROI 2 tended to decrease. A reduction of relative ZQ was observed in ROI 2 and thus ZV increased in relation to ZQ in ROI 2 and with a similar trend in ROI 3 (Figure 3).

Figure 2
figure 2

Relative distribution of ZV and ZQ within ROI 1 to 4 for the left and right lung at baseline and following balloon inflation. * denotes significant difference (p < 0.05) for ZQ compared with baseline. Relative distribution given as a percentage. ZV = ventilation-induced change in thoracic impedance; dotted line. ZQ = perfusion-induced change in thoracic impedance; solid line. Region of interest (ROI) 1 = ventral; ROI 2 = mid-ventral; ROI 3 = mid-dorsal; ROI 4 = dorsal. Binfl = balloon inflation; BL = baseline; LL = left lung; RL = right lung.

Figure 3
figure 3

Relative distribution of ZV and ZQ within ROI 1 to 4 for the left and right lung at baseline and following the application of PEEP 20. * denotes significant difference (p < 0.05) for ZQ compared with baseline. # denotes significant difference (p < 0.05) for ZV compared with baseline. Relative distribution given as a percentage. ZV = ventilation-induced change in thoracic impedance; dotted line. ZQ = perfusion-induced change in thoracic impedance; solid line. Region of interest (ROI) 1 = ventral; ROI 2 = mid-ventral; ROI 3 = mid-dorsal; ROI 4 = dorsal. BL = baseline; LL = left lung; PEEP20 = the application of 20 cmH2O of positive end-expiratory pressure; RL = right lung.

The calculated ZV/ZQ ratios were plotted against ZV and ZQ before and after Binfl (Figures 4 and 5) and PEEP20 (Figures 6 and 7). At baseline, a close to ideal ZV/ZQ ratio of 1 was observed in ROI 2 and similarly in ROI 3 and 4 although with increased scatter, while a trend towards venous admixture was noted in ROI 1. Binfl increased the ZV/ZQ ratio in ROI 2 indicating dead space ventilation while increased venous admixture appeared in ROI 1. Changes in ROI 3 and 4 did not reach statistical significance because of considerable scatter. Following PEEP20, ZV/ZQ ratio increased in ROI 2 consistent with dead space ventilation and a similar trend was noted in ROI 3 while venous admixture appeared in ROI 1. No significant changes appeared in ROI 4 as a result of persistent scatter.

Figure 4
figure 4

Scatterplots of relative ZV and ZQ within ROI 1 to 4 for the left lung at baseline and following balloon inflation in relation to the relative ZV/ZQ ratio. ZV = ventilation-induced change in thoracic impedance; open circle. ZQ = perfusion-induced change in thoracic impedance; solid stars. Values presented as a proportion of global ZV and ZQ. Region of interest (ROI) 1 = ventral; ROI 2 = mid-ventral; ROI 3 = mid-dorsal; ROI 4 = dorsal. Binfl = balloon inflation; BL = baseline.

Figure 5
figure 5

Scatterplots of relative ZV and ZQ within the ROI 1 to 4 for the right lung at baseline and following balloon inflation in relation to the relative ZV/ZQ ratio. ZV = ventilation-induced change in thoracic impedance; open circle. ZQ = perfusion-induced change in thoracic impedance; solid stars. Values presented as a proportion of global ZV and ZQ. Region of interest (ROI) 1 = ventral; ROI 2 = mid-ventral; ROI 3 = mid-dorsal; ROI 4 = dorsal. Binfl = balloon inflation; BL = baseline.

Figure 6
figure 6

Scatterplots of relative ZV and ZQ within the ROI 1 to 4 for the left lung at baseline and following PEEP20 in relation to the relative ZV/ZQ ratio. ZV = ventilation-induced change in thoracic impedance; open circle. ZQ = perfusion-induced change in thoracic impedance; solid stars. Values presented as a proportion of global ZV and ZQ. Region of interest (ROI) 1 = ventral; ROI 2 = mid-ventral; ROI 3 = mid-dorsal; ROI 4 = dorsal. BL = baseline; PEEP20 = the application of 20 cmH2O of positive end-expiratory pressure.

Figure 7
figure 7

Scatterplots of relative ZV and ZQ within the ROI 1 to 4 for the right lung at baseline and following PEEP20 in relation to the relative ZV/ZQ ratio. ZV = ventilation-induced change in thoracic impedance; open circle. ZQ = perfusion-induced change in thoracic impedance; solid stars. Values presented as a proportion of global ZV and ZQ. Region of interest (ROI) 1 = ventral; ROI 2 = mid-ventral; ROI 3 = mid-dorsal; ROI 4 = dorsal. BL = baseline; PEEP20 = the application of 20 cmH2O of positive end-expiratory pressure.

The frequency distribution of all EIT ZV/ZQ ratios calculated at baseline demonstrated a Gaussian distribution with a mean of 1.1 (95% confidence interval = 0.96 to 1.2). This distribution was markedly changed into a non-Gaussian pattern during Binfl and PEEP20 with a mean of 0.86 (95% confidence interval = 0.78 to 1.1; Figure 8). The calculated ZV/ZQ ratios based on global EIT correlated to venous admixture (r2 = 0.48 [regression coefficient]) with a negative slope (y = -194 × +67) and to fractional dead space (r2 = 0.63) with a positive slope (y = 92 × +32; Figure 9).

Figure 8
figure 8

Frequency distribution of ZV/ZQ ratios at baseline (top) and following balloon inflation or increase to PEEP20 (bottom). PEEP20 = the application of 20 cmH2O of positive end-expiratory pressure; ZQ = perfusion-induced change in thoracic impedance; ZV = ventilation-induced change in thoracic impedance.

Figure 9
figure 9

Linear regression plots for the global ZV/ZQ ratio versus the Qs/Qt (top) or the VD/VT (bottom). Qs/Qt = venous admixture; VD/VT = fractional dead space; ZQ = perfusion-induced change in thoracic impedance; ZV = ventilation-induced change in thoracic impedance.

Discussion

This study demonstrated the use of EIT to study pulmonary ventilation and perfusion in parallel. Using an experimental model to generate a wide range of stroke volumes for a constant tidal volume, we were able to investigate the dynamic relation between ZV and ZQ. By defining four ROIs along the ventral to dorsal axis for each of the left and right sides separately, an extensive dataset of ZV/ZQ ratios was collected. A largely parallel proportion of ZV and ZQ in all four ROIs during baseline conditions corresponded to a bell-shaped frequency distribution of ZV/ZQ ratios with only moderate scatter. Acute changes in stroke volume, established by inflating a balloon in the inferior caval vein or increasing PEEP from 5 to 20 cmH2O, while tidal volume was maintained, significantly increased the mismatch of regional, proportional ZV and ZQ, the scatter of ZV/ZQ ratios and the heterogeneity of the ZV/ZQ frequency distribution. Significant positive and negative correlations were demonstrated between fractional alveolar dead space and venous admixture, respectively, and the global ZV/ZQ ratio.

Although an expanding body of evidence supports the use of EIT to monitor pulmonary ventilation and perfusion separately, only few studies have been published on their contemporaneous measurements. Deibele and colleagues, using similar EIT equipment, focused on technical aspects to separate the cardiac and respiratory related signals in two healthy, spontaneously breathing volunteers [12]. Kunst and colleagues investigated left-to right division of ventilation and perfusion in 14 spontaneously breathing patients scheduled for lobectomy or pneumectomy for lung cancer [13]. Frerichs and colleagues studied pulmonary perfusion in three pigs using a hypertonic saline solution as an electrical impedance contrast agent [21]. Thus, this study extends previous observations by adding regional separation and V/Q matching during anaesthesia and mechanical ventilation. Venous admixture was assumed to be an inherent feature of the model using anaesthetised animals studied in the supine position with low PEEP. The balloon inflation mimics a situation of acute onset hypovolaemia and increased PEEP is commonly applied in mechanically ventilated patients to prevent or reduce lung collapse.

The method to assess pulmonary perfusion used in this study has been described previously [14] but warrants some comments. Global ZQ represents an approximation of stroke volume but, like any other method, has an inherent bias and variability, although within clinically acceptable limits. The physiology behind cardiac-related changes in ZQ is still imperfectly understood and, for example, the relative influence of large pulmonary vessels compared with pulmonary capillaries needs further research. Furthermore, ZQ relates to a change in air/blood content by the stroke volume rather than flow velocity and no separation of the latter in terms of alveolar capillary flow or shunt flow can be made.

The definition of ROI is an essential point in the data analysis. In this study, four equally sized ROIs were set by default, covering the full anterioposterior distance and the ZQ and ZV signals were analysed off-line using a stand-alone computer algorithm. It remains possible that setting the four ROIs within the pulmonary boundaries of the individual EIT image might have resulted in other ZV and ZQ amplitudes. In particular the most dorsal ROI 4 may be less representative because of the low impedance signal in this region [22]. The default full anterioposterior approach using equally sized ROIs was chosen in the design of this study to minimise operator-related bias. Initial pilot data indicated that placing the ROIs individually in each analysis significantly increased inter- and intraobserver variability.

The distribution of ventilation and perfusion at baseline differed along the ventral to dorsal distance with the greatest proportion delivered to the mid-ventral followed by the mid-dorsal regions. This is in line with the concept of gravitational influence on distribution and appeared even more clearly if all four ROIs were individually adjusted to the apparent pulmonary field of the EIT image (data not shown). Importantly, the distribution of ZV and ZQ were very closely matched, as would be expected in stable baseline conditions. Balloon inflation significantly reduced ZQ in ROI 2 leading to apparent dead space ventilation in this region, while an apparent redistribution of ZQ towards ROI 1 in the left lung would contribute to venous admixture. However, it should be noted that ROI 1 represents a moderate volume of tissue with the heart and lungs in close proximity.

The spatial resolution of EIT is inferior to that of other imaging techniques, for example computed tomography, and cannot exclusively separate impedance changes originating from the cardiopulmonary border. It is thus possible that the recorded increase in ZQ in ROI 1 during balloon inflation results from an increased cardiac contribution rather than an increase in pulmonary blood flow. Increased PEEP to 20 cmH2O shifted ventilation towards the more dorsal region, as previously reported [5]. At the same time, ZQ was reduced, leading to a relative increase in dead space ventilation. Thus, the well-matched distribution of ZQ and ZV was partially lost following the interventions. The relative distribution of ZQ and ZV highlights general changes in V/Q matching. In order to gain a more detailed picture, the individual relative ZQ and ZV were plotted against the ZV/ZQ ratio. This approach to illustrate V/Q matching is less prone to artefacts induced by the selection of ROIs because all data comes from within the same ROI, as opposed to the distribution plot that relates one ROI to all others.

Again, ventilation and perfusion were closely matched at baseline, although a scatter of ZV/ZQ ratios appeared in the most ventral and dorsal parts of the lung. Balloon inflation increased the scatter in most ROIs and increased dead space ventilation in ROI 2, consistent with the acute reduction of blood flow in this region. PEEP ventilation increased dead space in ROI 2 and 3, whereas venous admixture appeared in ROI 1, which is in agreement with the dorsal redistribution of ventilation. Furthermore, the ZV/ZQ plots demonstrated the heterogeneity including both wasted ventilation and perfusion that turned out as significant correlations to shunt and dead space calculated independently of EIT. The relative ease of operation, low cost and lack of invasive procedures and radiation of EIT compares favourably with other methods to assess V/Q matching such as radionuclide scanning [23] or multiple inert gas elimination technology [24].

This was primarily a feasibility study of EIT to determine V/Q relations, and as such has some important limitations. The results are based on a small group of animals and pooled data from multiple observations in each animal was used, which might exaggerate the statistical significance demonstrated. All animals were without lung pathology and thus the results regarding PEEP should not be directly extrapolated to the clinical context when PEEP is applied during mechanical ventilation for respiratory failure. The interruption of ventilation when measuring perfusion-related impedance changes excludes the effects of tidal volume changes on pulmonary circulation. Although the acquisition of EIT data could be achieved on-line within minutes, the off-line analyses were more time consuming and would not be directly available at the bedside. No comparisons were made to other methods of assessing V/Q matching such as radionuclide scanning [23] or the multiple inert gas technique [24], and hence data on precision and bias of EIT in this respect is lacking.

Conclusion

We conclude that EIT might be used to monitor the distribution of pulmonary ventilation and perfusion making detailed studies of V/Q matching possible. The technique holds substantial potential although further work is necessary to position EIT in the context of clinical critical care.

Key messages

  • EIT can be applied to monitor ZV and ZQ in isolated ROIs.

  • A largely parallel, proportional distribution of ZV and ZQ in all ROIs during baseline conditions corresponded to a bell-shaped frequency distribution of ZV/ZQ ratios with only moderate scatter.

  • Acute changes in stroke volume, established by inflating a balloon in the inferior caval vein or increasing PEEP from 5 to 20 cmH2O, while tidal volume was maintained, significantly increased the mismatch of regional, proportional ZV and ZQ, the scatter of ZV/ZQ ratios and the heterogeneity of the ZV/ZQ frequency distribution.

  • Significant positive and negative correlations were demonstrated between fractional alveolar dead space and venous admixture, respectively, and the global ZV/ZQ ratio.

  • EIT might be used to monitor the distribution of pulmonary ventilation and perfusion making detailed studies of V/Q matching possible.

Abbreviations

AU:

arbitrary units

Binfl:

the inflation of a Fogarty balloon in the inferior caval vein

EIT:

electrical impedance tomography

PEEP20:

the application of 20 cmH2O of positive end-expiratory pressure

Qs/Qt:

venous admixture

ROI:

region of interest

VD/VT:

fractional alveolar dead space

V/Q:

ventilation/perfusion ratio

ZQ:

perfusion-induced change in thoracic impedance

ZV:

ventilation-induced change in thoracic impedance.

References

  1. Adler A, Amyot R, Guardo R, Bates JH, Berthiaume Y: Monitoring changes in lung air and liquid volumes with electrical impedance tomography. J Appl Physiol 1997, 83: 1762-1767.

    CAS  PubMed  Google Scholar 

  2. Frerichs I: Electrical impedance tomography (EIT) in applications related to lung and ventilation: a review of experimental and clinical activities. Physiol Meas 2000, 21: R1-21. 10.1088/0967-3334/21/2/201

    Article  CAS  PubMed  Google Scholar 

  3. Luepschen H, Meier T, Grossherr M, Leibecke T, Karsten J, Leonhardt S: Protective ventilation using electrical impedance tomography. Physiol Meas 2007, 28: S247-260. 10.1088/0967-3334/28/7/S18

    Article  CAS  PubMed  Google Scholar 

  4. Meier T, Leibecke T, Eckmann C, Gosch UW, Grossherr M, Bruch HP, Gehring H, Leonhardt S: Electrical impedance tomography: changes in distribution of pulmonary ventilation during laparoscopic surgery in a porcine model. Langenbecks Arch Surg 2006, 391: 383-389. 10.1007/s00423-006-0034-8

    Article  CAS  PubMed  Google Scholar 

  5. Meier T, Luepschen H, Karsten J, Leibecke T, Grossherr M, Gehring H, Leonhardt S: Assessment of regional lung recruitment and derecruitment during a PEEP trial based on electrical impedance tomography. Intensive Care Med 2008, 34: 543-550. 10.1007/s00134-007-0786-9

    Article  PubMed  Google Scholar 

  6. Victorino JA, Borges JB, Okamoto VN, Matos GF, Tucci MR, Caramez MP, Tanaka H, Sipmann FS, Santos DC, Barbas CS, Carvalho CR, Amato MB: Imbalances in regional lung ventilation: a validation study on electrical impedance tomography. Am J Respir Crit Care Med 2004, 169: 791-800. 10.1164/rccm.200301-133OC

    Article  PubMed  Google Scholar 

  7. Vonk-Noordegraaf A 2nd, Janse A, Marcus JT, Bronzwaer JG, Postmust PE, Faes TJ, De Vries PM: Determination of stroke volume by means of electrical impedance tomography. Physiol Meas 2000, 21: 285-293. 10.1088/0967-3334/21/2/308

    Article  PubMed  Google Scholar 

  8. Smit HJ, Handoko ML, Vonk Noordegraaf A, Faes TJ, Postmus PE, de Vries PM, Boonstra A: Electrical impedance tomography to measure pulmonary perfusion: is the reproducibility high enough for clinical practice? Physiol Meas 2003, 24: 491-499. 10.1088/0967-3334/24/2/359

    Article  CAS  PubMed  Google Scholar 

  9. Smit HJ, Vonk-Noordegraaf A, Marcus JT, Weijden S, Postmus PE, de Vries PM, Boonstra A: Pulmonary vascular responses to hypoxia and hyperoxia in healthy volunteers and COPD patients measured by electrical impedance tomography. Chest 2003, 123: 1803-1809. 10.1378/chest.123.6.1803

    Article  PubMed  Google Scholar 

  10. Smit HJ, Vonk Noordegraaf A, Roeleveld RJ, Bronzwaer JG, Postmus PE, de Vries PM, Boonstra A: Epoprostenol-induced pulmonary vasodilatation in patients with pulmonary hypertension measured by electrical impedance tomography. Physiol Meas 2002, 23: 237-243. 10.1088/0967-3334/23/1/324

    Article  CAS  PubMed  Google Scholar 

  11. Smit HJ, Vonk Noordegraaf A, Marcus JT, Boonstra A, de Vries PM, Postmus PE: Determinants of pulmonary perfusion measured by electrical impedance tomography. Eur J Appl Physiol 2004, 92: 45-49. 10.1007/s00421-004-1043-3

    Article  PubMed  Google Scholar 

  12. Deibele JM, Luepschen H, Leonhardt S: Dynamic separation of pulmonary and cardiac changes in electrical impedance tomography. Physiol Meas 2008, 29: S1-14. 10.1088/0967-3334/29/6/S01

    Article  CAS  PubMed  Google Scholar 

  13. Kunst PW, Vonk Noordegraaf A, Hoekstra OS, Postmus PE, de Vries PM: Ventilation and perfusion imaging by electrical impedance tomography: a comparison with radionuclide scanning. Physiol Meas 1998, 19: 481-490. 10.1088/0967-3334/19/4/003

    Article  CAS  PubMed  Google Scholar 

  14. Fagerberg A, Stenqvist O, Åneman A: Monitoring pulmonary perfusion by electrical impedance tomography: an evaluation in a pig model. Acta Anaesthesiol Scand 2009, 53: 152-158. 10.1111/j.1399-6576.2008.01847.x

    Article  CAS  PubMed  Google Scholar 

  15. Yorkey TJ, Webster JG: A comparison of impedance tomographic reconstruction algorithms. Clin Phys Physiol Meas 1987, 8: 55-62. 10.1088/0143-0815/8/4A/007

    Article  PubMed  Google Scholar 

  16. Hahn G, Just A, Dudykevych T, Frerichs I, Hinz J, Quintel M, Hellige G: Imaging pathologic pulmonary air and fluid accumulation by functional and absolute EIT. Physiol Meas 2006, 27: S187-198. 10.1088/0967-3334/27/5/S16

    Article  CAS  PubMed  Google Scholar 

  17. National Academy of Sciences: Guide for the care and use of laboratory animals. Washington DC: National Academy Press; 1996.

    Google Scholar 

  18. Zinserling J, Wrigge H, Neumann P, Muders T, Magnusson A, Hedenstierna G, Putensen C: Methodologic aspects of attenuation distributions from static and dynamic thoracic CT techniques in experimental acute lung injury. Chest 2005, 128: 2963-2970. 10.1378/chest.128.4.2963

    Article  PubMed  Google Scholar 

  19. Rylander C, Högman M, Perchiazzi G, Magnusson A, Hedenstierna G: Oleic acid lung injury: a morphometric analysis using computed tomography. Acta Anaesthesiol Scand 2004, 48: 1123-1129. 10.1111/j.1399-6576.2004.00482.x

    Article  CAS  PubMed  Google Scholar 

  20. Hardman JG, Aitkenhead AR: Estimating alveolar dead space from the arterial to end-tidal CO(2) gradient: a modeling analysis. Anesth Analg 2003, 97: 1846-1851. 10.1213/01.ANE.0000090316.46604.89

    Article  PubMed  Google Scholar 

  21. Frerichs I, Hinz J, Herrmann P, Weisser G, Hahn G, Quintel M, Hellige G: Regional lung perfusion as determined by electrical impedance tomography in comparison with electron beam CT imaging. IEEE Trans Med Imaging 2002, 21: 646-652. 10.1109/TMI.2002.800585

    Article  PubMed  Google Scholar 

  22. Pulletz S, van Genderingen HR, Schmitz G, Zick G, Schadler D, Scholz J, Weiler N, Frerichs I: Comparison of different methods to define regions of interest for evaluation of regional lung ventilation by EIT. Physiol Meas 2006, 27: S115-127. 10.1088/0967-3334/27/5/S10

    Article  PubMed  Google Scholar 

  23. Harris RS, Schuster DP: Visualizing lung function with positron emission tomography. J Appl Physiol 2007, 102: 448-458. 10.1152/japplphysiol.00763.2006

    Article  PubMed  Google Scholar 

  24. Wagner PD: The multiple inert gas elimination technique (MIGET). Intensive Care Med 2008, 34: 994-1001. 10.1007/s00134-008-1108-6

    Article  PubMed  Google Scholar 

Download references

Acknowledgements

The EIT equipment was kindly provided by Dräger Medical AG within an unrestricted evaluation agreement. The valuable technical discussion with Mr. Eckhard Teschner, Dräger Medical AG, is gratefully acknowledged. This study received financial support from the LUA-project at Sahlgrenska University Hospital, the Swedish Research Council and the Göteborg Medical Society. Part of this work was presented at the 20th Annual Congress of the European Society of Intensive Care Medicine, 7 to 10 October, 2007, Berlin, Germany.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Anders Åneman.

Additional information

Competing interests

OS has received lecturing reimbursements from Dräger Medical. AF and AÅ declare that they have no competing interests.

Authors' contributions

All authors equally contributed to the design, data acquisition, data analysis and manuscript preparation. All authors have read and approved the final manuscript.

Authors’ original submitted files for images

Rights and permissions

Reprints and permissions

About this article

Cite this article

Fagerberg, A., Stenqvist, O. & Åneman, A. Electrical impedance tomography applied to assess matching of pulmonary ventilation and perfusion in a porcine experimental model. Crit Care 13, R34 (2009). https://doi.org/10.1186/cc7741

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/cc7741

Keywords