An improved Bussignac device for the delivery of noninvasive continuous positive airway pressure: the SUPER-Bussignac
Critical Care volume 12, Article number: P342 (2008)
Introduction
The purpose of this work was to test, in a bench study, the performance of a modified Boussignac continuous positive airway pressure (CPAP) system, aimed at avoiding the drop in inspired oxygen fraction (FiO2) during high inspiratory patient peak flows.
Methods
We modified a Boussignac system (Bssc), inserting between the valve and the face mask a T-piece, connected with a 1.5 l reservoir balloon, supplemented with oxygen independently from the valve itself. During inspiration, the patient inhales oxygen from the reservoir, diminishing the blending with atmospheric air and keeping the FiO2 higher. The performance of the system was evaluated in a bench study, applying a CPAP face mask to a plexiglass cylinder connected with an active lung simulator (ASL 5000; IngMar Medical, USA) generating tidal volumes with increasing inspiratory peak flow rates (V'insp). Two positive end-expiratory pressure (PEEP) levels were tested for the Bssc, SUPER-Boussignac with 10 l/min (SB10) and 30 l/min (SB30) supplementary oxygen flows. The FiO2 was continuously recorded and the lowest value reached during each tidal volume (FiO2) was averaged over 20 breaths.
Conclusion
The SUPER-Boussignac is simple and effective in increasing (up to 30%) the FiO2, and limiting the drop related to V'insp during noninvasive CPAP.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Bellani, G., Foti, G., Spagnolli, E. et al. An improved Bussignac device for the delivery of noninvasive continuous positive airway pressure: the SUPER-Bussignac. Crit Care 12 (Suppl 2), P342 (2008). https://doi.org/10.1186/cc6563
Published:
DOI: https://doi.org/10.1186/cc6563