Tracheal wall pressures in the clinical setting: comparing Portex Soft Seal and Lotrach cuffs
Critical Care volume 12, Article number: P339 (2008)
Introduction
The tracheal wall pressure (TWP) exerted by a tracheal tube cuff should normally be kept between 20 and 30 cmH2O. This protects against mucosal injury whilst allowing ventilation without an audible air leak [1]. The Portex Soft Seal high-volume low-pressure (HVLP) cuff has a working intracuff pressure of 30 cmH2O, providing a safe TWP of the same value because there is no tension in the cuff wall material. There are, however, folds within the cuff wall that allow passage of subglottic fluid to the tracheobronchial tree below, increasing the risk of ventilator-associated pneumonia [2]. The Lotrach endotracheal tube was designed to prevent this leakage at an equivalent TWP to that of correctly inflated HVLP cuffs [3]. Each Lotrach cuff is individually calibrated to transmit only 30 cmH2O to the tracheal wall, yet because of the lack of folds in the cuff wall it has been shown to prevent of aspiration of subglottic contents [4]. Although extensively tested in benchtop models and pig tracheas, we wished to demonstrate that the Lotrach cuff had an identical sealing pressure, and therefore TWP, as the HVLP cuff in normal clinical practice.
Methods
One hundred and two ventilated patients were intubated with either the Lotrach (n = 54) or the Portex Soft Seal (n = 48) tubes. Both the Lotrach and Portex Soft Seal cuffs were inflated to their working pressures. Whilst undertaking staged recruitment manoeuvres (up to 40 cmH2O), the positive end-expiratory pressure at which laryngeal air leak occurred was noted.
Results
The seal pressures (TWP) are presented in Table 1.
Conclusion
Both the Portex Soft Seal and Lotrach cuffs exert an equal and safe TWP when inflated to their recommended working pressures.
References
Seegobin RD, et al.: BMJ. 1984, 288: 965-968.
Young PJ, et al.: BJA. 1997, 78: 557-562.
Young PJ, et al.: Med Eng Phys. 2003, 25: 465-473. 10.1016/S1350-4533(02)00252-7
Young PJ, et al.: Anaesthesia. 1999, 54: 559-563. 10.1046/j.1365-2044.1999.00850.x
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Santhirapala, R., Blunt, M., Fletcher, A. et al. Tracheal wall pressures in the clinical setting: comparing Portex Soft Seal and Lotrach cuffs. Crit Care 12 (Suppl 2), P339 (2008). https://doi.org/10.1186/cc6560
Published:
DOI: https://doi.org/10.1186/cc6560