- Poster presentation
- Published:
Sidestream dark-field imaging versus orthogonal polarization spectroscopic imaging: a comparative study
Critical Care volume 12, Article number: P63 (2008)
Introduction
Sidestream dark-field (SDF) imaging, a stroboscopic LED ring-based imaging modality for clinical observation of the microcirculation, was validated by quantitative and qualitative comparison with orthogonal polarization spectral (OPS) imaging.
Methods
For OPS imaging a Cytocan-II backfocus type device (Cytometrics, Philadelphia, PA, USA) was used, and for SDF imaging a MicroScan Video Microscope (MicroVision Medical Inc., Amsterdam, The Netherlands) was employed. To validate SDF imaging, nailfold capillary diameters and red blood cell velocities were measured in the exact same capillaries using OPS and SDF imaging. For quantitative comparison of the quality of sublingually acquired microcirculatory images, an image quality quantification system was developed to score venular and capillary contrast and sharpness on scales from 0 to 1.
Results
After introduction of a scaling factor to correct for the slightly higher magnification of the SDF device with respect to the OPS device, equal quantitative results for capillary diameters and red blood cell velocities were obtained. Venular contrast and sharpness were shown to be comparable for OPS and SDF imaging. Capillary sharpness and contrast, however, were shown to be significantly higher using SDF imaging (Figure 1). Venular granularity, in addition, was more clearly observable employing the SDF device.
Conclusion
SDF imaging provided significantly higher image quality by the use of stroboscopic LED ring-based SDF illumination. It is anticipated that SDF imaging will serve as a reliable imaging modality for the clinical assessment of the microcirculation and will enhance computer-aided image analysis.
References
Goedhart PT, et al.: Opt Express. 2007, 15: 15101-15114. 10.1364/OE.15.015101
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Bezemer, R., Goedhart, P., Khalilzada, M. et al. Sidestream dark-field imaging versus orthogonal polarization spectroscopic imaging: a comparative study. Crit Care 12 (Suppl 2), P63 (2008). https://doi.org/10.1186/cc6284
Published:
DOI: https://doi.org/10.1186/cc6284