Maintenance of tracheal tube cuff pressure: where are the limits?

Abstract

Continuous control of tracheal tube cuff inflation using a pneumatic device resulted in severe tracheal wall damage in ventilated piglets. This damage was similar in piglets managed with manual control of cuff inflation. The periodic hyperinflation of the tube cuff used in both groups of this study may explain these results. This manoeuvre should be avoided in clinical practice.

In a previous issue of Critical Care, Nseir and colleagues presented an article regarding continuous control of endotracheal cuff pressure and tracheal wall damage [1].

Among the pathogenic mechanisms responsible for ventilator-associated pneumonia (VAP), oropharyngeal colonization by potentially pathogenic microorganisms and silent aspiration of subglottic secretions around the tracheal tube cuff seem to play a pivotal role [2]. In order to prevent pneumonia, several approaches have been proposed – such as placing patients in the semirecumbent position [3], continuous aspiration of subglottic secretions (CASS) above the tracheal tube cuff [4], oropharyngeal decontamination by antiseptics [5], and the application of antiseptic-impregnated endotracheal tubes [6].

The key element of the proposed pathogenesis of VAP appears to be aspiration of colonized oropharyngeal and subglottic secretions. Appropriate control of the endotracheal tube cuff pressure (Pcuff) may therefore serve as a major prevention target. Intubated patients were recommended to be managed with Pcuff values between 20 and 30 cmH2O to provide a sufficient seal without compromising mucosal perfusion [7]. The routine management of cuff inflation consists of periodic manual checking of the Pcuff, which does not ensure the appropriate maintenance of the Pcuff during continuous tracheal intubation [8]. Moreover, the manual checking of the Pcuff may cause either overinflation or deflation of the cuff and may cause aspiration of contaminated secretions to the lower airway during the manoeuvre. Leaks and loss of Pcuff are frequent in intubated patients, and a persistent Pcuff below 20 cmH2O was an independent risk factor for VAP in one study [8]. Consequently, appropriate maintenance of pressure of the tracheal tube cuff is recommended in recent guidelines [9].

In a previous issue of the journal, Nseir and coworkers describe a pneumatic device for the continuous control of the Pcuff in an animal model [1]. The aim of the study was to assess whether the continuous control of the Pcuff results in reduced tracheal ischaemic lesions in mechanically ventilated piglets. For this purpose, the authors compared the pneumatic device with the manual control of Pcuff in a randomized trial. The pneumatic device provided effective continuous control of the Pcuff, with longer periods of Pcuff within the target values than piglets managed with manual control. This device is therefore potentially useful for clinical practice in order to avoid both excessive inflation and deflation of the cuff. Hyperaemia and haemorrhages in the trachea were observed at the cuff contact area in all animals, however, with no differences between animals with and without the pneumatic device.

Several devices that provide an automatic and continuous effective control of the Pcuff have been described in the literature. Most of these devices are not automatic, some devices need frequent control by the attending staff, and other devices operating in a more automatic and continuous way are complex, requiring the use of special and expensive equipment that may not be available routinely [10]. It is probable that these issues concerning complexity and cost could explain the lack of continuous automatic control of cuff inflation in clinical practice.

We have described a simple and cheap device that is very effective for the routine maintenance of adequate cuff inflation during mechanical ventilation that does not require any specific equipment [11]. A recent randomized clinical trial in mechanically ventilated patients comparing this device with the routine manual control of cuff inflation, however, showed no benefits in the prevention of VAP [12]. These findings suggest that other factors than cuff inflation influence the microaspiration of secretions to the lower airways around the tracheal tube cuff. Commercially available high-volume low-pressure tracheal tubes such as those used in the study often form folds around the cuff, hence allowing leakage of secretions pooled above the tube cuff in studies in vitro, even at Pcuff levels similar to those used by Nseir and colleagues in piglets [13]. Several devices have consequently been recently developed in order to overcome this problem. Among those devices, the Microcuff endotracheal high-volume low-pressure tube features an ultrathin (7 μm) poly-urethane cuff membrane around an inner conventional inflatable cuff. This tube is effective in preventing fluid leakage around the cuff in an in vitro setup [14]. The combination of this device with CASS is effective in preventing both early-onset and late-onset VAP in a recent clinical study [15].

One of the potential concerns of all these devices, particularly CASS, is the potential damage of the tracheal wall. In an animal sheep model, Berra and colleagues demonstrated important tracheal lesions when using CASS [16]. We do not know whether this is applicable to humans. In the study by Nseir and colleagues, the tracheal lesions found could be explained, at least in part, by the high inflation pressure they applied eight times daily via 50 ml during 30 min. This is not the current clinical practice in humans, and after this study it should be completely avoided.

Abbreviations

CASS:

= continuous aspiration of subglottic secretions

cuff:

Pcuff = endotracheal tube cuff pressure

VAP:

= ventilator-associated pneumonia.

References

  1. Nseir S, Duguet A, Copin MC, De Jonckheere J, Zhang M, Similowski T, Marquette CH: Continuous control of endotracheal cuff pressure and tracheal wall damage: a randomized controlled animal study. Crit Care 2007, 11: R109. 10.1186/cc6142

    Article  PubMed Central  PubMed  Google Scholar 

  2. Bonten MJ, Kollef MH, Hall JB: Risk factors for ventilator-associated pneumonia: from epidemiology to patient management. Clin Infect Dis 2004, 38: 1141-1149. 10.1086/383039

    Article  PubMed  Google Scholar 

  3. Drakulovic MB, Torres A, Bauer TT, Nicolas JM, Nogue S, Ferrer M: Supine body position as a risk factor for nosocomial pneumonia in mechanically ventilated patients: a randomised trial. Lancet 1999, 354: 1851-1858. 10.1016/S0140-6736(98)12251-1

    Article  CAS  PubMed  Google Scholar 

  4. Valles J, Artigas A, Rello J, Bonsoms N, Fontanals D, Blanch L, Fernandez R, Baigorri F, Mestre J: Continuous aspiration of sub-glottic secretions in preventing ventilator-associated pneumonia. Ann Intern Med 1995, 122: 179-186.

    Article  CAS  PubMed  Google Scholar 

  5. Chlebicki MP, Safdar N: Topical chlorhexidine for prevention of ventilator-associated pneumonia: a meta-analysis. Crit Care Med 2007, 35: 595-602. 10.1097/01.CCM.0000253395.70708.AC

    Article  CAS  PubMed  Google Scholar 

  6. Rello J, Kollef M, Diaz E, Sandiumenge A, del CY, Corbella X, Zachskorn R: Reduced burden of bacterial airway colonization with a novel silver-coated endotracheal tube in a randomized multiple-center feasibility study. Crit Care Med 2006, 34: 2766-2772. 10.1097/01.CCM.0000242154.49632.B0

    Article  PubMed  Google Scholar 

  7. Sengupta P, Sessler DI, Maglinger P, Wells S, Vogt A, Durrani J, Wadhwa A: Endotracheal tube cuff pressure in three hospitals, and the volume required to produce an appropriate cuff pressure. BMC Anesthesiol 2004, 4: 8. 10.1186/1471-2253-4-8

    Article  PubMed Central  PubMed  Google Scholar 

  8. Rello J, Sonora R, Jubert P, Artigas A, Rue M, Valles J: Pneumonia in intubated patients: role of respiratory airway care. Am J Respir Crit Care Med 1996, 154: 111-115.

    Article  CAS  PubMed  Google Scholar 

  9. Guidelines for the management of adults with hospital-acquired, ventilator-associated, and healthcare-associated pneumonia Am J Respir Crit Care Med 2005, 171: 388-416. 10.1164/rccm.200405-644ST

  10. Miller DM: A pressure regulator for the cuff of a tracheal tube. Anaesthesia 1992, 47: 594-596. 10.1111/j.1365-2044.1992.tb02332.x

    Article  CAS  PubMed  Google Scholar 

  11. Farre R, Rotger M, Ferrer M, Torres A, Navajas D: Automatic regulation of the cuff pressure in endotracheally-intubated patients. Eur Respir J 2002, 20: 1010-1013. 10.1183/09031936.02.02692001

    Article  CAS  PubMed  Google Scholar 

  12. Valencia M, Ferrer M, Farre R, Navajas D, Badia JR, Nicolas JM, Torres A: Automatic control of tracheal tube cuff pressure in ventilated patients in semirecumbent position: a randomized trial. Crit Care Med 2007, 35: 1543-1549. 10.1097/01.CCM.0000266686.95843.7D

    Article  PubMed  Google Scholar 

  13. Young PJ, Rollinson M, Downward G, Henderson S: Leakage of fluid past the tracheal tube cuff in a benchtop model. Br J Anaesth 1997, 78: 557-562.

    Article  CAS  PubMed  Google Scholar 

  14. Dullenkopf A, Gerber A, Weiss M: Fluid leakage past tracheal tube cuffs: evaluation of the new Microcuff endotracheal tube. Intensive Care Med 2003, 29: 1849-1853. 10.1007/s00134-003-1933-6

    Article  PubMed  Google Scholar 

  15. Lorente L, Lecuona M, Alejandro J, Maria M, Antonio S: Influence of an endotracheal tube with polyurethane cuff and subglottic drainage on pneumonia. Am J Respir Crit Care Med 2007, 176: 1979-1783. 10.1164/rccm.200705-761OC

    Article  Google Scholar 

  16. Berra L, De Marchi L, Panigada M, Yu ZX, Baccarelli A, Kolobow T: Evaluation of continuous aspiration of subglottic secretion in an in vivo study. Crit Care Med 2004, 32: 2071-2078. 10.1097/01.CCM.0000142575.86468.9B

    Article  PubMed  Google Scholar 

Download references

Acknowledgements

The present study was supported by CibeRes (ISCiii-CB06/06/0028), FIS 02-0744, SEPAR 2001, and 2005 SGR 00822.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Antoni Torres.

Additional information

Competing interests

The authors declare that they have no competing interests.