- Poster presentation
- Published:
Does the cycling-off criteria of pressure support change the respiratory parameters in intensive care unit patients?
Critical Care volume 11, Article number: P63 (2007)
Introduction
In modern mechanical ventilators it is possible to modify the flow cycling-off criteria of pressure support ventilation. The changes in the flow cycling-off criteria of pressure support ventilation can modify the synchrony between the mechanical and neural inspiration termination.
Objective
To compare the effects of two different flow cycling-off criteria of pressure support ventilation on the respiratory parameters of ICU mechanically ventilated patients.
Methods
We prospectively evaluated 20 intubated and mechanically ventilated adult ICU patients recovering from acute respiratory failure that could be comfortably ventilated with pressure support of 15 cmH2O, PEEP of 5 cmH2O and FIO2 of 40%. The patients were ventilated at two different cycling-off criteria of pressure support of 25% and 40% of the peak inspiratory flow. We evaluated the respiratory rate (RR), expiratory tidal volume (VTe), minute ventilation (MV), VCO2, VTCO2, ETCO2, SpO2, mean arterial pressure (MAP) and heart rate (HR).
Results
Comparisons between different cycling-off values did not result in any statistically significant changes for the evaluated variables. There were no significant changes in blood pressure or HR under any experimental conditions (Table 1).
Conclusion
The changes in cycling of criteria from 25% to 40% of the peak flow did not affect the respiratory parameters in mechanically ventilated patients in a mixed ICU.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Correa, T., Passos, R., Kanda, S. et al. Does the cycling-off criteria of pressure support change the respiratory parameters in intensive care unit patients?. Crit Care 11 (Suppl 3), P63 (2007). https://doi.org/10.1186/cc5850
Published:
DOI: https://doi.org/10.1186/cc5850