Implementing tight glycaemic control: performance of bedside glucometers
Critical Care volume 11, Article number: P141 (2007)
Introduction
Implementing tight glycaemic control (TGC) in the ICU requires accurate blood glucose (BG) monitoring. We evaluated the performance of two bedside glucometers (GM) in ICU patients.
Methods
Four hundred and fifty-two arterial blood samples were prospectively analysed in 37 adult ICU patients subjected to TGC. Arterial BG was simultaneously determined using a reference test (ABL®) and two GM (Accu-Chek® and HemoCue®). Data were analysed using linear regression and the Bland–Altman (BA) method.
Results
Correlation between the reference method and both GM in the overall BG range was reasonable, but not perfect (r2 ≥ 0.93). This was further underlined by BA analysis (Figures 1 and 2), showing a bias to overestimate BG with GM. In the TGC range (80–110 mg/dl) correlation was low for both GM (r2 ≤ 0.66). This was confirmed by BA analysis, demonstrating broad limits of agreement: +14.2 and -26.6 mg/dl for Accu-Chek® and +5.5 and -31.1 mg/dl for HemoCue®.
Conclusion
The accuracy of the tested GM in our ICU patients was insufficient for safe clinical practice. Therefore, to avoid potentially harmful hypoglycaemia, caution is warranted when TGC is implemented exclusively based on BG results obtained by GM.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Vlasselaers, D., Vandewiele, K., Herpe, T.V. et al. Implementing tight glycaemic control: performance of bedside glucometers. Crit Care 11 (Suppl 2), P141 (2007). https://doi.org/10.1186/cc5301
Published:
DOI: https://doi.org/10.1186/cc5301