Continuous central venous oxygenation measurement by CeVOX in patients undergoing off-pump coronary artery bypass grafting
Critical Care volume 10, Article number: P340 (2006)
Introduction
Less invasive measurement of central venous O2 saturation(ScvO2) has shown to be a valuable alternative to the determination of mixed venous O2 saturation for monitoring of O2 supply/demand [1, 2]. The aim of this study was to compare ScvO2 measured continuously by the new CeVOX (Pulsion Medical System, Munich, Germany) device (CScvO2) with ScvO2 determined by blood gas co-oximetry (BScvO2).
Methods
Ten ASA III patients undergoing elective off-pump coronary artery bypass grafting were studied during the operation (OP) and during their ICU stay. In addition to the standard hemodynamic monitoring according to institutional policy, a CeVOX fiber-optic probe was introduced into a standard central venous catheter placed via internal jugular vein access. OP and ICU measurement started after in-vivo calibration of CeVOX. BScvO2 and CScvO2 readings were recorded at intervals of 30 min during OP and of 120 min during ICU. Data were statistically analyzed using Bland-Altman analysis, Pearson correlation and t test for the periods during OP, ICU, a set of three consecutive measurements during OP immediately after calibration (OPcal) and 4 hours later (OP4h) as well as immediately after calibration on the ICU (ICUcal) and 14 hours later (ICU14h). Trend analysis was performed, calculating differences (Δ) between consecutive measurements. P < 0.05 was considered significant.
Results
One hundred and twenty-nine matched sets of data were obtained (OP:n = 78, ICU: n = 51) with a wide range of ScvO2 values (BScvO2 = 48.0–91.0%, CScvO2 = 49.0–94.0%). The OP observation time was 4.0–6.5 hours and the ICU measurement sequence was 14.0–20.0 hours. Bland-Altman analysis revealed an overall mean bias ± 2SD (limits of agreement) of-0.7 ± 7.8% for CScvO2 – BScvO2 during OP and -1.1± 11.6% during ICU (Fig. 1). There was no significant difference between CScvO2 and BScvO2 (OP: P = 0.120, ICU: P = 0.167). The correlation coefficient(r2) for CScvO2 vs BScvO2 was 0.885 (OP) and0.592 (ICU). Statistics for OPcal' OP4hand ICUcal were comparable, whereas for ICU14hthe bias ± 2SD increased and r2 decreased (Table 1). Trend analysis showed no significant difference (OP: ΔBScvO2 = -1.3 ± 9.0%, ΔCScvO2 = -1.1 ± 8.6%, P = 0.663; ICU: ΔBScvO2 = -0.4 ± 6.8%, ΔBScvO2 = -0.4 ± 5.8%, P = 0.828).

Figure 1
Conclusion
These preliminary results indicate that ScvO2 can be reliably assessed by CeVOX. Scheduled recalibration at intervals <14 hours may be mandatory.
References
Respiration. 2001, 685: 279-285. 10.1159/000050511
Acta Anaesthesiol Scand. 1998, 42: 172-177.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Hofer, C., Ganter, M., Fodor, P. et al. Continuous central venous oxygenation measurement by CeVOX in patients undergoing off-pump coronary artery bypass grafting. Crit Care 10 (Suppl 1), P340 (2006). https://doi.org/10.1186/cc4687
Published:
DOI: https://doi.org/10.1186/cc4687