Colibri coloriometric technology rapidly detects oesophagal intubations
Critical Care volume 3, Article number: P005 (2000)
Methods
An entdotracheal tube was placed both in the trachea and the oesophagus in otherwise healthy patients undergoing elective surgery under general anaesthesia. We compared the four first ventilations of the endotracheal and oesophageal tube using capnography and a Capno Bri indicator with four different colour gradings. (Blue ~ 0.5%, dark green ~ 1.0%, light green ~ 3.0% and yellow ~ 4.0%)
Results
In all patients (n = 9), the indicator confirmed correct placement of the tube in the trachea at the first ventilation (yellow color). The indicator also verified incorrect oesophageal placement at the first ventilation in all patients (blue color).
These results were confirmed by the capnography.
Conclusion
The Colibri technology is a reliable technique for confirmation of correct endotracheal tube placement. It may be especially suitable in emergency situations where capnography is not available
References
Sum Ping ST: Accuracy of the FEFCO2 detector in the assessment of endotracheal tube placement. Anaest Analg. 1992, 74: 415-419.
Sayah AJ: End-tidal CO2 measurement in the detection of esophagus intubation during cardiac arrest. Ann Emerg Med. 1990, 19: 8-10.1016/S0196-0644(05)81557-4.
Singer M: . Colibri: a new means of CO2 detection. ESA Congress in Barcelona. 1998
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Puntervoll, S., Søreide, E., Jacewicz, W. et al. Colibri coloriometric technology rapidly detects oesophagal intubations. Crit Care 3 (Suppl 2), P005 (2000). https://doi.org/10.1186/cc380
Published:
DOI: https://doi.org/10.1186/cc380