Uneven distribution of ventilation in acute respiratory distress syndrome

Abstract

Introduction

The aim of this study was to assess the volume of gas being poorly ventilated or non-ventilated within the lungs of patients treated with mechanical ventilation and suffering from acute respiratory distress syndrome (ARDS).

Methods

A prospective, descriptive study was performed of 25 sedated and paralysed ARDS patients, mechanically ventilated with a positive end-expiratory pressure (PEEP) of 5 cmH2O in a multidisciplinary intensive care unit of a tertiary university hospital. The volume of poorly ventilated or non-ventilated gas was assumed to correspond to a difference between the ventilated gas volume, determined as the end-expiratory lung volume by rebreathing of sulphur hexafluoride (EELVSF6), and the total gas volume, calculated from computed tomography images in the end-expiratory position (EELVCT). The methods used were validated by similar measurements in 20 healthy subjects in whom no poorly ventilated or non-ventilated gas is expected to be found.

Results

EELVSF6 was 66% of EELVCT, corresponding to a mean difference of 0.71 litre. EELVSF6 and EELVCT were significantly correlated (r2 = 0.72; P < 0.001). In the healthy subjects, the two methods yielded almost identical results.

Conclusion

About one-third of the total pulmonary gas volume seems poorly ventilated or non-ventilated in sedated and paralysed ARDS patients when mechanically ventilated with a PEEP of 5 cmH2O. Uneven distribution of ventilation due to airway closure and/or obstruction is likely to be involved.

Introduction

Decreased functional residual capacity (FRC) and increased pulmonary resistance are hallmarks of acute respiratory distress syndrome (ARDS) [1]. Pathophysiological mechanisms include alveolar flooding and/or collapse, which contribute to shunting of blood and to hypoxaemia [2]. Whether true alveolar collapse or intraluminar oedema with increased impedance dominates is a matter of debate [3]. Furthermore, the expiratory flow limitation observed in ARDS patients has been attributed to the closure of small airways [4]. Pulmonary gas distal to such an airway closure/obstruction may be poorly ventilated or non-ventilated. If so, it might not be included in FRC measurements based on tracer gas dilution. The end-expiratory lung volume determined by tracer gas dilution is termed 'ventilated gas volume' in this paper. Other techniques such as body plethysmography and radiographical methods [5] determine the total end-expiratory volume of pulmonary gas, irrespective of whether it is well ventilated, poorly ventilated or non-ventilated. This volume is termed 'total gas volume' in this report. A difference between the ventilated gas volume and the total gas volume can be interpreted as a volume of gas being poorly ventilated or non-ventilated. This difference is obvious in patients with chronic obstructive airway disease in whom FRC determined by gas dilution might be considerably lower than FRC determined by body plethysmography [6]. However, in mechanically ventilated ARDS patients the volume of poorly ventilated or non-ventilated gas seems not to have been studied in detail.

The aim of the present study was therefore to assess the volume of poorly ventilated or non-ventilated gas in mechanically ventilated ARDS patients, assuming the difference between the ventilated gas volume and the total gas volume to represent poorly ventilated or non-ventilated gas. To validate the methods involved, similar measurements were performed in young healthy subjects in whom no poorly ventilated or non-ventilated gas is expected to be found.

Materials and methods

Ethical approval

The study was approved by the local ethics committee and conducted in accordance with the Helsinki Declaration. Informed consent was obtained from the next-of-kin of the patients and directly from the healthy subjects.

Patients

Twenty-five sedated and mechanically ventilated patients were included from a mixed-adult intensive care unit. The criterion for selection was the eligible ARDS patient [7] having spent the longest time on mechanical ventilation at the time of the once-weekly available opportunity for computed tomography (CT). Patients were eligible for the study only if their arterial oxygenation was stable and between 10 and 26 kPa during mechanical ventilation with the following parameters: fraction of inspired oxygen 0.5; constant flow volume-controlled mode; tidal volume 8 to 10 ml/kg; positive end-expiratory pressure (PEEP) 5 cmH2O. Chronic obstructive pulmonary disease was not an exclusion criterion but was present only in one patient (no. 13). Clinical data are given in Table 1. Twenty healthy non-smoking students independent of the investigating institutions were enrolled and interviewed to rule out any history of tobacco use or obstructive lung disease. Anthropometric data for both groups are given in Table 2.

Table 1 Clinical data of the ARDS patients
Table 2 Anthropometric data

Measurements

The ventilated gas volume was determined in both groups by a gas dilution technique using rebreathing of sulphur hexafluoride. End-expiratory measurements in the ventilated patients were made at a PEEP of 5 cmH2O (EELVSF6) and measurements in the spontaneously breathing healthy subjects were made at the FRC level (FRCSF6). A prototype system (AMIS 2001; Innovision A/S, Odense, Denmark) equipped with a photoacoustic and magnetoacoustic multigas analyser [8] was used. The accuracy of the analyser was checked by comparison with mass spectrometry (AMIS 2000; Innovision A/S) before and after the series of experiments. Before each measurement, the ambient temperature and pressure were registered and correct readings from the gas analyser were verified by supplying room air and the undiluted tracer gas mixture to the gas inlet. The gas sampling rate was 120 ml/min. The rebreathing unit consisted of a bag-in-box system in which the flexible rubber bellows could be manually ventilated by a piston fitted through the distal short end of the cylinder. For operation, the unit was instantly switched into the patient circuit by a pneumatic slide valve without disconnection. The bellows was initially filled with 1.2 litres (ambient temperature and pressure, dry) of a gas mixture of 1.0% SF6 in 5.0% nitrous oxide (N2O) and oxygen (bal; medical grade). The presence of N2O was due to the circulatory monitoring function of the multimodal monitoring system. The SF6 concentration was continuously plotted during 30 s of ventilation at a frequency of 20 breaths per minute (Fig. 1). Allowing for the tubing dead space (101 ml in the subjects, 107 ml in the patients), the ventilated gas volume was calculated from a formula based on standard gas dilution principles for FRC measurements:

Figure 1
figure 1

Concentration of the tracer gas sulphur hexafluoride (SF6) plotted during 30 s of rebreathing in a supine healthy subject.

where Pb is the barometric pressure in torr, T is the ambient absolute temperature and SF6i and SF6e are the initial and equilibrated concentration of SF6 (standard temperature and pressure, dry), respectively, and 1.2 is the bellows volume. FRC symbolises both FRCSF6 in the young healthy subjects and EELVSF6 in the ventilated patients.

The total gas volume was calculated from CT images reconstructed from a scan lasting about 20 s in a high-speed scanner (GE High Speed CT/i; General Electric Medical Systems, Milwaukee, WI, USA). End-expiratory measurements in the patients were made in apnoea at PEEP 5 cmH2O (EELVCT) and measurements in the healthy subjects were made in apnoea at the FRC level (FRCCT). The following exposure parameters were used: 120 kV; 170 mA; rotation time 1.0 s; collimation 1 mm and a matrix of 512 × 512, yielding voxel volumes of 0.25 to 0.49 mm3 depending on the field of view. An initial topogram defined the limits of the lungs, and the first and last scanning levels were positioned at the apical and caudal extremes, respectively. In between, eight more scanning levels were evenly dispersed, making a total of 10 consecutive single exposures with a distance between the scans of 18 to 25 mm, depending on thoracic dimensions. The total effective radiation dose was estimated to equal one standard chest X-ray examination, yielding an average absorbed radiation of 0.8 mGy to the breasts of female subjects. Within each image, the lungs were manually delineated from the thoracic wall in a single region of interest. Within the region of interest, the voxels with attenuation values between – 1,000 and 0 Hounsfield units (HU) were automatically selected for analysis by software (MALUNA 2.02; Peter Herrmann, Mannheim, Germany) on a personal computer, and their gas volume (V) was calculated from the formula [9]

where Vvox is the single-voxel volume of n voxels within the slice. The total gas volume was calculated by interpolating for the volume of gas in the lung tissue between the 10 scan levels by the method of Kvist [10] with the modified formula

where V1 and V2 are the gas volumes of two adjacent slices with the thickness t, separated by the centre distance d. FRC symbolises both FRCCT in the young healthy subjects and EELVCT in the ventilated patients.

During the measurements, the sedated patients were temporarily paralysed and ventilated by means of a mobile ventilator (Servo 900 C; Siemens, Solna, Sweden) with the settings described above. The end-expiratory position was achieved by activation of the expiratory hold function on the ventilator. The patient was then either ventilated from the rebreathing circuit or CT scanned in maintained apnoea. The rebreathing procedure was performed in duplicate before and after a single CT exposure.

Before the supine measurements, the nose-clipped, supine and relaxed healthy subjects breathed room air through a mouthpiece connected to the rebreathing system through a three-way valve. At the FRC level, the valve was either switched into the rebreathing system for gas dilution by spontaneous breathing or was closed during the CT examination. The rebreathing procedure was performed in duplicate before and after a single CT exposure.

Statistical analysis

Data are presented as means ± standard deviation if not specified otherwise. The level of significance was defined as P < 0.05. The coefficient of variation (CV) for paired measurements was calculated as the standard deviation of the differences divided by the mean of all measurements [11]. Calculations were performed with the software package Statistica 6.0 (StatSoft Inc., Tulsa, OK, USA) on a personal computer.

Results

In the ARDS patients, EELVSF6 was 66 ± 14% of EELVCT. EELVSF6 was found systematically lower than EELVCT except in one patient (no. 19), in whom they were similar. The mean difference, corresponding to the poorly ventilated or non-ventilated gas volume, was 0.71 ± 0.47 litre. The magnitude of the poorly ventilated or non-ventilated gas volume was not correlated with age or ventilator days. Mean results are given in Table 3. EELVSF6 and EELVCT were significantly correlated (r = 0.85; P < 0.001) (Fig. 2). The CV of duplicate EELVSF6 measurements was 5.6%.

Table 3 Lung volumes
Figure 2
figure 2

Linear regression between EELV measurements by rebreathing of sulphur hexafluoride (EELVSF6) and by computed tomography (EELVCT) obtained in 25 ARDS patients. The dotted line is the regression line EELVSF6 = 0.4EELVCT + 0.3 (r2 = 0.72; P < 0.001).

In the supine healthy subjects, FRCSF6 was 99 ± 9% of FRCCT, and they were closely correlated (r = 0.91; P < 0.001) (Fig. 3). The differences did not depend on the magnitude of FRC (Fig. 4). The CV of duplicate FRCSF6 measurements was 3.1%.

Figure 3
figure 3

Linear regression between FRC measurements by rebreathing of sulphur hexafluoride (FRCSF6) and by computed tomography (FRCCT) in 20 healthy subjects. The dotted line is the regression line: EELVSF6 = 0.9FRCCT + 0.1 (r2 = 0.83; P < 0.001).

Figure 4
figure 4

Bland-Altman plot [24] of supine functional residual capacity measured by rebreathing of sulphur hexafluoride (FRCSF6) and by computed tomography (FRCCT) in 20 healthy subjects. The individual differences of paired measurements (y axis) did not depend on the magnitude of their average values (x axis). The mean difference (solid line; dotted lines represent the mean ± 2SD) was small.

Discussion

This study shows that there is a considerable volume of poorly ventilated or non-ventilated gas present in the lungs of sedated and paralysed ARDS patients when mechanically ventilated with a PEEP of 5 cmH2O.

We assumed that the difference between the ventilated gas volume determined by gas dilution and the total gas volume calculated from CT corresponds to a poorly ventilated or non-ventilated gas volume. The methods used to determine these volumes were validated by comparison of similar measurements in young healthy subjects, in whom they should yield similar results because these lungs are homogeneously ventilated with no obstruction and no airway closure. Indeed, almost identical results were obtained in the young healthy subjects. Furthermore, the CV of duplicate measurements in the healthy subjects indicated a good repeatability. The FRCSF6 values might seem somewhat low compared with predicted FRC values based on a mixed adult population (Table 3), but normal FRC values in supine young subjects are rare and the predictions therefore remain uncertain. The CT interpolation technique has been validated previously for heterogeneously scattered tissue [12] and should be precise enough with 10 scans evenly distributed over the lungs. In summary, we consider that the two methods used were adequate and that the difference between their results in the ARDS patients can be assumed to correspond to a poorly ventilated or non-ventilated gas volume.

The most likely pathophysiological mechanism associated with this volume is airway closure and/or obstruction. Further contribution from atelectasis formation during the inspiratory hold is unlikely with the fraction of inspiratory oxygen used [13]. However, the deep sedation and paralysis of the ARDS patients might have contributed to poor ventilation in the dependent parts of the lungs [14]. The lung injury is unevenly distributed in ARDS [15], which causes an uneven distribution of ventilation including overdistension of non-dependent regions. By definition, open but non-compliant lung units are poorly ventilated or non-ventilated, but this seems unlikely to be of any importance in ARDS patients during ventilation with a PEEP of 5 cmH2O.

The PEEP level applied in the present study was chosen to be clinically relevant [16] but it does not effectively counteract expiratory derecruitment of lung units. In a study of 10 ARDS patients, mechanically ventilated with zero end-expiratory pressure (ZEEP), Koutsoukou and colleagues determined an intrinsic PEEP of 4.1 ± 2.4 cmH2O, and expiratory flow limitation was demonstrated in eight of them [4]. These results suggest the presence of airway closure and/or obstruction at the FRC level in ARDS. In contrast, when closed circuit helium rebreathing and CT were recently compared in a group of 21 ARDS patients, mechanically ventilated with a PEEP of 12 ± 5 cmH2O, similar EELVs were found [17]. This finding indicates that there is no airway closure and/or obstruction when a PEEP of 12 cmH2O is applied. Indeed, it was recently also shown that the intrinsic PEEP and the expiratory flow limitation present at ZEEP can be eliminated by a PEEP of 10 cmH2O [18]. In summary, those studies and the present results indicate that airway closure and/or obstruction occurs at low levels of PEEP or ZEEP and that the distal gas volume is recruitable for more effective ventilation by a moderate increase in PEEP. Accordingly, increasing PEEP from 0 to 15 cmH2O has been shown in a study of pulmonary mechanics to increase pulmonary compliance in some patients, which was associated with the recruitment of lung units with preserved normal compliance [19]. Furthermore, low compliance during the initial phase of inspiration has been attributed to non-collapsed but slowly ventilated lung units, in which the ventilation can be increased by increased PEEP [20]. The gas content of such non-collapsed but poorly ventilated lung units may correspond to the volume of poorly ventilated or non-ventilated gas demonstrated in the present study.

Substantially elevated pressure in the airways is associated with signs of parenchymal overdistension [21]. CT studies have shown that this effect is located to non-dependent well-aerated lung units that become overdistended by the airway pressure required to inflate compressed dependent lung units [22]. Overdistension associated with increased airway pressure seems to be less pronounced when the parenchyma is diffusely affected without regional atelectasis [23], as in our patients. Possibly, the poorly ventilated or non-ventilated gas volume in this type of diffuse ARDS might reflect gas contained in lung units distal to airway closure and/or obstruction. The recruitment of such gas-containing lung units, excluded from effective ventilation by partial compression or oedema, can be expected to require a smaller elevation of transmural pressure than that needed to inflate completely collapsed lung units. If the volume of poorly ventilated or non-ventilated gas is small or non-existent, a moderately raised airway pressure might be ineffective for recruitment and merely contribute to the risk of overdistension.

Conclusion

We conclude that about one-third of the total gas volume is poorly ventilated or non-ventilated in the lungs of sedated and paralysed ARDS patients when mechanically ventilated with a PEEP of 5 cmH2O. This indicates uneven distribution of ventilation due to the presence of small-airway closure and/or obstruction at this PEEP level. Such a poorly ventilated or non-ventilated gas volume might be recruited for more effective ventilation by an increase in airway pressure that is less than the inflation pressure of completely collapsed lung units.

Key messages

  • This study demonstrates uneven distribution of ventilation in 25 sedated and ventilated ARDS patients by comparing the total end-expiratory gas end volume calculated from computed tomography and the ventilated gas volume measured by inert gas rebreathing.

  • The poorly ventilated or non-ventilated volume distal to the possible airway closure and/or obstruction might be recruited for more effective ventilation by an increase in airway pressure that is less than the inflation pressure of completely collapsed lung units.

Abbreviations

ARDS:

= acute respiratory distress syndrome

CT:

= computed tomography

CV:

= coefficient of variation

EELVCTEELV CT:

= total gas volume calculated from computed tomography images in the end-expiratory position

EELVSF6EELV SF6:

= end-expiratory lung volume measured by rebreathing of sulphur hexafluoride

FRC:

= functional residual capacity

FRCCTFRC CT:

= FRC calculated from computed tomography scans

FRCSF6SF6:

= FRC measured by sulfur hexafluoride rebreathing

HU:

Hounsfield unit

PEEP:

= positive end-expiratory pressure

ZEEP:

= zero end-expiratory pressure.

References

  1. Ramachandran PR, Fairley HB: Changes in functional residual capacity during respiratory failure. Can Anaesth Soc J 1970, 17: 359-369.

    Article  CAS  PubMed  Google Scholar 

  2. Ware LB, Matthay MA: The acute respiratory distress syndrome. N Engl J Med 2000, 342: 1334-1349. 10.1056/NEJM200005043421806

    Article  CAS  PubMed  Google Scholar 

  3. Hubmayr RD: Perspective on lung injury and recruitment: a skeptical look at the opening and collapse story. Am J Respir Crit Care Med 2002, 165: 1647-1653. 10.1164/rccm.2001080-01CP

    Article  PubMed  Google Scholar 

  4. Koutsoukou A, Armaganidis A, Stavrakaki-Kallergi C, Vassilakopoulos T, Lymberis A, Roussos C, Milic-Emili J: Expiratory flow limitation and intrinsic positive end-expiratory pressure at zero positive end-expiratory pressure in patients with adult respiratory distress syndrome. Am J Respir Crit Care Med 2000, 161: 1590-1596.

    Article  CAS  PubMed  Google Scholar 

  5. Kendrick AH: Comparison of methods of measuring static lung volumes. Monaldi Arch Chest Dis 1996, 51: 431-439.

    CAS  PubMed  Google Scholar 

  6. Rodenstein DO, Stanescu DC: Reassessment of lung volume measurement by helium dilution and by body plethysmography in chronic air-flow obstruction. Am Rev Respir Dis 1982, 126: 1040-1044.

    CAS  PubMed  Google Scholar 

  7. Bernard GR, Artigas A, Brigham KL, Carlet J, Falke K, Hudson L, Lamy M, Legall JR, Morris A, Spragg R: The American-European Consensus Conference on ARDS. Definitions, mechanisms, relevant outcomes, and clinical trial coordination. Am J Respir Crit Care Med 1994, 149: 818-824.

    Article  CAS  PubMed  Google Scholar 

  8. Clemensen P, Christensen P, Norsk P, Gronlund J: A modified photo- and magnetoacoustic multigas analyzer applied in gas exchange measurements. J Appl Physiol 1994, 76: 2832-2839.

    CAS  PubMed  Google Scholar 

  9. Gattinoni L, Pesenti A, Avalli L, Rossi F, Bombino M: Pressure-volume curve of total respiratory system in acute respiratory failure. Computed tomographic scan study. Am Rev Respir Dis 1987, 136: 730-736.

    Article  CAS  PubMed  Google Scholar 

  10. Kvist H, Sjostrom L, Tylen U: Adipose tissue volume determinations in women by computed tomography: technical considerations. Int J Obes 1986, 10: 53-67.

    CAS  PubMed  Google Scholar 

  11. Hankinson JL, Stocks J, Peslin R: Reproducibility of lung volume measurements. Eur Respir J 1998, 11: 787-790.

    CAS  PubMed  Google Scholar 

  12. Kvist H: Adipose tissue volume determinations by computed tomography. In Dissertation. Göteborg University, Sweden; 1988.

    Google Scholar 

  13. Edmark L, Kostova-Aherdan K, Enlund M, Hedenstierna G: Optimal oxygen concentration during induction of general anesthesia. Anesthesiology 2003, 98: 28-33. 10.1097/00000542-200301000-00008

    Article  CAS  PubMed  Google Scholar 

  14. Frerichs I, Hahn G, Golisch W, Kurpitz M, Burchardi H, Hellige G: Monitoring perioperative changes in distribution of pulmonary ventilation by functional electrical impedance tomography. Acta Anaesthesiol Scand 1998, 42: 721-726.

    Article  CAS  PubMed  Google Scholar 

  15. Kallet RH, Katz JA: Respiratory system mechanics in acute respiratory distress syndrome. Respir Care Clin N Am 2003, 9: 297-319.

    Article  PubMed  Google Scholar 

  16. Esteban A, Anzueto A, Alia I, Gordo F, Apezteguia C, Palizas F, Cide D, Goldwaser R, Soto L, Bugedo G, et al.: How is mechanical ventilation employed in the intensive care unit? An international utilization review. Am J Respir Crit Care Med 2000, 161: 1450-1458.

    Article  CAS  PubMed  Google Scholar 

  17. Patroniti N, Bellani G, Manfio A, Maggioni E, Giuffrida A, Foti G, Pesenti A: Lung volume in mechanically ventilated patients: measurement by simplified helium dilution compared to quantitative CT scan. Intensive Care Med 2004, 30: 282-289. 10.1007/s00134-003-2109-0

    Article  PubMed  Google Scholar 

  18. Koutsoukou A, Bekos B, Sotiropoulou C, Koulouris NG, Roussos C, Milic-Emili J: Effects of positive end-expiratory pressure on gas exchange and expiratory flow limitation in adult respiratory distress syndrome. Crit Care Med 2002, 30: 1941-1949. 10.1097/00003246-200209000-00001

    Article  PubMed  Google Scholar 

  19. Ranieri VM, Eissa NT, Corbeil C, Chasse M, Braidy J, Matar N, Milic-Emili J: Effects of positive end-expiratory pressure on alveolar recruitment and gas exchange in patients with the adult respiratory distress syndrome. Am Rev Respir Dis 1991, 144: 544-551.

    Article  CAS  PubMed  Google Scholar 

  20. Vieillard-Baron A, Prin S, Chergui K, Page B, Beauchet A, Jardin F: Early patterns of static pressure–volume loops in ARDS and their relations with PEEP-induced recruitment. Intensive Care Med 2003, 29: 1929-1935. 10.1007/s00134-003-1938-1

    Article  PubMed  Google Scholar 

  21. Vieira SR, Puybasset L, Richecoeur J, Lu Q, Cluzel P, Gusman PB, Coriat P, Rouby JJ: A lung computed tomographic assessment of positive end-expiratory pressure-induced lung overdistension. Am J Respir Crit Care Med 1998, 158: 1571-1577.

    Article  CAS  PubMed  Google Scholar 

  22. Rouby JJ, Puybasset L, Nieszkowska A, Lu Q: Acute respiratory distress syndrome: lessons from computed tomography of the whole lung. Crit Care Med 2003,31(4 Suppl):S285-295. 10.1097/01.CCM.0000057905.74813.BC

    Article  PubMed  Google Scholar 

  23. Puybasset L, Gusman P, Muller JC, Cluzel P, Coriat P, Rouby JJ: Regional distribution of gas and tissue in acute respiratory distress syndrome. III. Consequences for the effects of positive end-expiratory pressure. CT Scan ARDS Study Group. Adult Respiratory Distress Syndrome. Intensive Care Med 2000, 26: 1215-1227. 10.1007/s001340051340

    Article  CAS  PubMed  Google Scholar 

  24. Bland JM, Altman DG: Statistical methods for assessing agreement between two methods of clinical measurement. Lancet 1986, 1: 307-310.

    Article  CAS  PubMed  Google Scholar 

  25. Ibanez J, Raurich JM: Normal values of functional residual capacity in the sitting and supine positions. Intensive Care Med 1982, 8: 173-177.

    Article  CAS  PubMed  Google Scholar 

Download references

Acknowledgements

The results of this study were in part presented at the ESICM meeting in Rome in 1999. The study was supported by departmental funding and by grants from the Gothenburg Medical Association. The inert gas system (AMIS 2001) with consumables was made available by Innovision A/S, Odense, Denmark.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Christian Rylander.

Additional information

Competing interests

The author(s) declare that they have no competing interests.

Authors' contributions

CR, UT and BB conceived the study and designed the protocol. UT, MQ and PH defined the radiographical image analysis. CR and RRN performed measurements. CR, UT and BB wrote and revised the manuscript, which was reviewed and approved by all authors before final submission.

Authors’ original submitted files for images

Rights and permissions

Reprints and permissions

About this article

Cite this article

Rylander, C., Tylén, U., Rossi-Norrlund, R. et al. Uneven distribution of ventilation in acute respiratory distress syndrome. Crit Care 9, R165 (2005). https://doi.org/10.1186/cc3058

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/cc3058

Keywords