Recent years have seen the place of the pulmonary artery catheter in intensive care increasingly challenged, with one recent study reporting no difference in outcome in patients treated with or without a pulmonary artery catheter. However, this study has several methodological flaws and, although pulmonary artery catheterization should not be performed routinely on all patients, when used correctly by trained personnel in selected patients the pulmonary artery catheter continues to provide valuable information.

Introduction

More than 20 years ago, Shoemaker and coworkers [1, 2] observed that perioperative alterations in oxygen transport were closely related to the development of organ failure and death. Subsequently, several studies reported that perioperative hemodynamic optimization guided by the pulmonary artery catheter may decrease morbidity and mortality [3–6]. Nevertheless, the use of the pulmonary artery catheter has been challenged because of its invasiveness and possibly the unwarranted interventions that may result from its use [7].

In a recent issue of the New England Journal of Medicine, Sandham and coworkers [8] reported the results of a multicenter Canadian study that investigated the effects of right heart catheterization on perioperative complications in high-risk patients undergoing noncardiac surgery. From 1990 to 1999, those authors randomly allocated 1994 American Society of Anesthesiologists class III and IV patients to conventional monitoring and therapy or to right heart catheterization and hemodynamic optimization. They observed that survival (up to 1 year of follow up) and hospital stay did not differ between the two groups. The incidence of perioperative complications was similar in both groups, except for an increased incidence of pulmonary embolism in the pulmonary artery catheter group.

Although Sandham and coworkers [8] must be commended for their important undertaking, the study raises a number of important concerns. First, although the authors claimed that no patient selection was performed, the inclusion rate of a mean of only 22 patients/center per year was surprisingly low. For example, close to 1000 patients with American Society of Anesthesiologists class III and IV are operated on each year in our 760-bed institution. One of the inclusion criteria was the commitment of the surgeon and the anesthesiologist to adhere to the study concept; the most severely ill patients might therefore have been excluded, and this may explain the unexpectedly low mortality rate among the patients studied (a 15% mortality rate was included in the power calculation of the study).

Second, although the authors claimed that preoperative optimization was performed, in the vast majority of patients the goals were achieved only postoperatively (Fig. 2 of the paper, which reported the maximal value for the corresponding period) [8]. Only in a very limited number of patients were the resuscitation goals achieved in the preoperative and intraoperative periods. Unfortunately, the time from catheter insertion to initiation of surgery was not mentioned. Also, there was no indication as to when these values were achieved and for how long. Indeed, the time allocated to achieve optimization is crucial, because insertion of a catheter just before the beginning of surgery does not allow sufficient time to achieve hemodynamic optimization, and the catheter will then only be used to observe hemodynamic alterations. The low rate of attainment of hemodynamic goals contrasts with previous studies reporting beneficial effects of perioperative optimization. For example, all of the patients in the study conducted by Wilson and colleagues [5] achieved an oxygen delivery of at least 600 ml/min per m2 throughout the study. In addition, the protocol used to achieve the hemodynamic end-points is not well defined. The authors used fluids and vasoactive agents, but the type and doses of these agents were not specified. It is likely that these elements varied from one institution to another.

Third, the incidence of pulmonary embolism may have been different in the two groups but it is surprising to read that no pulmonary embolism was diagnosed in the control group. If a pulmonary embolism had been diagnosed in only two of the 997 patients in the control group (certainly a realistic figure), then the statistical significance would disappear. Of note, the physicians involved in the diagnosis of pulmonary embolism were not blinded to catheter insertion. This may counterbalance the somewhat lower incidence of renal failure in the group of patients receiving the pulmonary artery catheter. One should also note that chance alone can explain one positive P value at the 5% confidence level when more than 20 statistical comparisons are performed.

Fourth and most important, the goals were not protocolized and were less than optimal. Why was there a maximal value for cardiac output? Were patients supposed to receive a β-blocking agent when cardiac output exceeded 4.5 l/min per m2? Also, why was such a high level (18 mmHg) of pulmonary artery occlusion pressure taken as an endpoint? The analysis of the cardiac function curves could have resulted in a lower pulmonary artery occlusion pressure, and on this basis a lesser risk for pulmonary oedema. Also why was the mixed venous oxygen saturation (SvO2) not taken as a valuable (and perhaps the best [9, 10]) end-point? SvO2 reflects the balance between oxygen supply and demand, and it is particularly useful when large changes in oxygen demand are expected to occur (e.g. in anesthesia, hypothermia, pain, etc.). The monitoring of SvO2 may also limit the risk of over-treatment with vasoactive agents, because high doses of inotropic agents may worsen the balance between oxygen supply and demand even though cardiac output increases [10]. Oxygen extraction has already been used in various groups of surgical patients to assess the adequacy of resuscitation at the time of the protocol design [11–16]. Recently, Rivers and coworkers [17] reported that mortality can be decreased when the central venous oxygen saturation is used to guide therapy in patients with severe sepsis and septic shock.

We believe that this important study highlights again that pulmonary artery catheters should not be inserted routinely in patients with limited risk for death [18] but it does not imply that it should be avoided in patients with a higher perioperative risk for death. Less invasive monitoring techniques, such as esophageal Doppler or lithium or transpulmonary dilution, may be required to optimize stroke volume in patients with limited risk for death [19].

Abbreviations

SvO:

mixed venous oxygen saturation.

References

  1. Shoemaker WC, Chang PC, Czer LSC, Bland R, Shabot MM, State D: Cardiorespiratory monitoring in postoperative patients: I. Prediction of outcome and severity of illness. Crit Care Med 1979, 7: 237-242.

    Article  CAS  PubMed  Google Scholar 

  2. Shoemaker WC, Appel PL, Kram HB: Tissue oxygen debt as a determinant of lethal and nonlethal postoperative organ failure. Crit Care Med 1988, 16: 1117-1120.

    Article  CAS  PubMed  Google Scholar 

  3. Shoemaker WC, Appel PL, Kram HB, Waxman K, Lee T-S: Prospective trial of supranormal values of survivors as therapeutic goals in high-risk surgical patients. Chest 1988, 94: 1176-1186.

    Article  CAS  PubMed  Google Scholar 

  4. Boyd O, Grounds M, Bennett ED: A randomized clinical trial of the effect of deliberate perioperative increase of oxygen delivery on mortality in high-risk surgical patients. JAMA 1993, 270: 2699-2707. 10.1001/jama.270.22.2699

    Article  CAS  PubMed  Google Scholar 

  5. Wilson J, Woods I, Fawcett J, Whall R, Dibb W, Morris C, McManus E: Reducing the risk of major elective surgery: randomised controlled trial of preoperative optimisation of oxygen delivery. BMJ 1999, 318: 1099-1103.

    Article  PubMed Central  CAS  PubMed  Google Scholar 

  6. Lobo SM, Salgado PF, Castillo VG, Borim AA, Polachini CA, Palchetti JC, Brienzi SL, de Oliveira GG: Effects of maximizing oxygen delivery on morbidity and mortality in high-risk surgical patients. Crit Care Med 2000, 28: 3396-3404. 10.1097/00003246-200010000-00003

    Article  CAS  PubMed  Google Scholar 

  7. Connors AF Jr, Speroff T, Dawson NV, Thomas C, Harrell FE Jr, Wagner D, Desbiens N, Goldman L, Wu AW, Califf RM, Fulkerson WJ Jr, Vidaillet H, Broste S, Bellamy P, Lynn J, Knaus WA: The effectiveness of right heart catheterization in the initial care of critically ill patients. JAMA 1996, 276: 889-897. 10.1001/jama.276.11.889

    Article  PubMed  Google Scholar 

  8. Sandham JD, Hull RD, Brant RF, Knox L, Pineo GF, Doig CJ, Laporta DP, Viner S, Passerini L, Devitt H, Kirby A, Jacka M, Canadian Critical Care Clinical Trials Group: A randomized, controlled trial of the use of pulmonary-artery catheters in high-risk surgical patients. N Engl J Med 2003, 348: 5-14. 10.1056/NEJMoa021108

    Article  PubMed  Google Scholar 

  9. Vincent JL, De Backer D: Cardiac output measurement: is least invasive always the best? Crit Care Med 2002, 30: 2380-2382. 10.1097/00003246-200210000-00035

    Article  PubMed  Google Scholar 

  10. Teboul JL, Graini L, Boujdaria R, Berton C, Richard C: Cardiac index vs oxygen-derived parameters for rational use of dobutamine in patients with congestive heart failure. Chest 1993, 103: 81-85.

    Article  CAS  PubMed  Google Scholar 

  11. Jastremski MS, Chelluri L, Beney KM, Bailly RT: Analysis of the effects of continuous on-line monitoring of mixed venous oxygen saturation on patient outcome and cost-effectiveness. Crit Care Med 1989, 17: 148-153.

    Article  CAS  PubMed  Google Scholar 

  12. Kyff JV, Vaughn S, Yang SC, Raheja R, Puri VK: Continuous monitoring of mixed venous oxygen saturation in patients with acute myocardial infarction. Chest 1989, 95: 607-611.

    Article  CAS  PubMed  Google Scholar 

  13. Prakash O, Meij S, van den Borden B, Saxena PR: Cardiorespiratory monitoring during open heart surgery. Crit Care Med 1981, 9: 530-535.

    CAS  PubMed  Google Scholar 

  14. Shenaq SA, Casar G, Chelly JE, Ott H, Crawford ES: Continuous monitoring of mixed venous oxygen saturation during aortic surgery. Chest 1987, 92: 796-799.

    Article  CAS  PubMed  Google Scholar 

  15. Shoemaker WC, Appel PL, Kram HB: Hemodynamic and oxygen transport effects of dobutamine in critically ill general surgical patients. Crit Care Med 1986, 14: 1032-1037.

    Article  CAS  PubMed  Google Scholar 

  16. Birman H, Haq A, Hew E, Aberman A: Continous monitoring of mixed venous oxygen saturation in hemodynamically unstable patients. Chest 1984, 86: 753-756.

    Article  CAS  PubMed  Google Scholar 

  17. Rivers E, Nguyen B, Havstad S, Ressler J, Muzzin A, Knoblich B, Peterson E, Tomlanovich M, Early Goal-Directed Therapy Collaborative Group: Early goal-directed therapy in the treatment of severe sepsis and septic shock. N Engl J Med 2001, 345: 1368-1377. 10.1056/NEJMoa010307

    Article  CAS  PubMed  Google Scholar 

  18. Vincent JL, Dhainaut JF, Perret C, Suter P: Is the pulmonary artery catheter misused? A European view. Crit Care Med 1998, 26: 1283-1287. 10.1097/00003246-199807000-00035

    Article  CAS  PubMed  Google Scholar 

  19. Sinclair S, James S, Singer M: Intraoperative intravascular volume optimisation and length of hospital stay after repair of proximal femoral fracture: randomised controlled trial. BMJ 1997, 315: 909-912.

    Article  PubMed Central  CAS  PubMed  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Daniel De Backer.

Additional information

Competing interests

None declared.

Rights and permissions

Reprints and permissions

About this article

Cite this article

De Backer, D., Creteur, J. & Vincent, JL. Perioperative optimization and right heart catheterization: what technique in which patient?. Crit Care 7, 201 (2003). https://doi.org/10.1186/cc2177

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc2177

Keywords