Volume 7 Supplement 2

23rd International Symposium on Intensive Care and Emergency Medicine

Open Access

Assessment of variables affecting the accuracy of extravascular lung water (EVLW) by single versus double indicator dilution techniques in ARDS patients

  • E Maggioni1,
  • F Sala1,
  • A Manfio1,
  • E Saccavino1,
  • B Cortinovis1,
  • G Bellani1 and
  • N Patroniti1
Critical Care20037(Suppl 2):P186

https://doi.org/10.1186/cc2075

Published: 3 March 2003

Introduction

Estimation of extravascular lung water (EVLW) by the single indicator dilution technique (SDT) implemented in the PICCO monitoring system relies on a relationship between intrathoracic blood volume (ITBV) and global end diastolic volume (GEDV), which has been derived from a large population of mixed critically ill patients using the double indicator dilution technique (thermal and indocyanine green dyes, DDT): ITBVsdt = 1.25 × GEDV [1]. Since the difference between ITBV and GEDV corresponds to pulmonary blood volume (PBV), we can write: PBVsdt = 0.25 × GEDV. Any factor influencing this relationship can affect SDT accuracy. The aims of our study were: 1) to compare SDT versus DDT for EVLW measurements in ARDS patients; and 2) to explore factors influencing the PBV and GEDV relationship, and affecting SDT accuracy.

Methods

We studied 21 ARDS patients (PaO2/FiO2 = 165 ± 65), monitored with a Swan–Ganz catheter, and a 4 F thermistore-tipped, fiberoptic catheter inserted through a femoral artery, both connected to a COLD monitoring system. SDT measurements were obtained by mean transit time and down slope time of thermal indicator using standard formula. Statistical analysis was performed by single and multiple linear regression analysis; accuracy was assessed according to Bland and Altman method.

Results

Agreement between EVLWddt and EVLWsdt showed a bias of 15.3 ± 135.8, and a 95% confidence interval of -256 and 287; the correlation coefficient was r = 0.92 (P < 0.001). Correlations of GEDV and PBVsdt with PBVddt, although significant, were poor (r = 0.394, P < 0.01). EVLWddt indexed to body weight (EVLWi) and cardiac index (CI) explained 64% of PBV indexed to body surface area (PBVi) variance (P < 0.001). EVLWi, CI, central venous pressure (CVP), and PaCO2 explained 82% of PBVi to GEDVi ratio variance (P < 0.001). EVLWddt to EVLWsdt differences were highly correlated with PBVddt (r = 0.88, P < 0.001). EVLWi, CI and PaCO2 explained 65% of the differences between EVLWddt and EVLWsdt variance (EVLWddt - EVLWsdt = -263 + 18.64 × EVLWi - 50.9 × CI + 5.3 × PaCO2, r2 = 0.65, P < 0.001).

Conclusions

The SDT gives acceptable estimates of EVLW in ARDS patients. EVLWi, CI and PaCO2 may influence the relationship between PBV and GEDV, and may affect accuracy of EVLW measured by SDT.

Authors’ Affiliations

(1)
Department of Anaesthesia and Intensive Care, University of Milano-Bicocca, S. Gerardo Hospital

References

  1. Sakka SG, et al.: Assessment of cardiac preload and extravascular lung water by single transpulmonary thermodilution. Intensive Care Med 2000, 26: 180-187. 10.1007/s001340050043View ArticlePubMedGoogle Scholar

Copyright

© BioMed Central Ltd 2003

Advertisement