Inspiratory gas temperature in the endotracheal tube when ventilated with a heated humidifier

International Organization of Standardization recommends that inspiratory gas should possess absolute humidity of more than 33 mg/l. In this study we measured the temperature in the endotracheal tube (ETT) with a hypothesis that actual temperature of inspiratory gas is lower than that set by heated humidifier resulting in an inadequate humidification.

Methods

A lung model (TTL, Michigan Instruments) was ventilated using a PB840 ventilator (Puritan-Bennett) via an 8 mm ETT. A heated humidifier (MR290, Fisher & Paykel) was set in the inspiratory limb of a standard ventilator circuit with a target temperature of 37°C. The ventilator settings were: volume control mode, rate 10/min, PEEP 0 cmH2O with various tidal volumes (VT). The temperature of inspiratory gas was measured at 22 cm, 27 cm and 32 cm from the distal end of the ETT using a thin wire thermocouple thermometer.

Results

Mean inspiratory temperatures (°C) are shown in Table 1.

Table 1

Conclusion

The actual mean inspiratory gas temperature was lower than the target temperature set by the humidifier, especially with low VT, which may result in an inadequate humidification.

Author information

Authors and Affiliations

Authors

Rights and permissions

Reprints and permissions

About this article

Cite this article

Fujino, Y., Uchiyama, A., Miyoshi, E. et al. Inspiratory gas temperature in the endotracheal tube when ventilated with a heated humidifier. Crit Care 7 (Suppl 2), P158 (2003). https://doi.org/10.1186/cc2047

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc2047

Keywords