Comparison of bedside measurement of cardiac output with the thermodilution method and the Fick method in mechanically ventilated patients

Abstract

Introduction

Bedside cardiac output determination is a common preoccupation in the critically ill. All available methods have drawbacks. We wished to re-examine the agreement between cardiac output determined using the thermodilution method (QT THERM) and cardiac output determined using the metabolic (Fick) method (QT FICK) in patients with extremely severe states, all the more so in the context of changing practices in the management of patients. Indeed, the interchangeability of the methods is a clinically relevant question; for instance, in view of the debate about the risk–benefit balance of right heart catheterization.

Patients and methods

Eighteen mechanically ventilated passive patients with a right heart catheter in place were studied (six women, 12 men; age, 39–84 years; simplified acute physiology scoreII, 39–111). QT THERM was obtained using a standard procedure. QT FICK was measured from oxygen consumption, carbon dioxide production, and arterial and mixed venous oxygen contents. Forty-nine steady-state pairs of measurements were performed. The data were normalized for repeated measurements, and were tested for correlation and agreement.

Results

The QT FICK value was 5.2 ± 2.0 l/min whereas that of QT THERM was 5.8 ± 1.9 l/min (R = 0.840, P < 0.0001; mean difference, -0.7 l/min; lower limit of agreement, -2.8 l/min; upper limit of agreement, 1.5 l/min). The agreement was excellent between the two techniques at QT THERM values <5 l/min but became too loose for clinical interchangeability above this value. Tricuspid regurgitation did not influence the results.

Discussion and conclusions

No gold standard is established to measure cardiac output in critically ill patients. The thermodilution method has known limitations that can lead to inaccuracies. The metabolic method also has potential pitfalls in this context, particularly if there is increased oxygen consumption within the lungs. The concordance between the two methods for low cardiac output values suggests that they can both be relied upon for clinical decision making in this context. Conversely, a high cardiac output value is more difficult to rely on in absolute terms.

Introduction

Estimating cardiac output at the bedside is a common preoccupation in critically ill patients. Many methods are available; some invasive and others not, some operator dependent and others not. The thermodilution cardiac output obtained through right heart catheterization has been the clinical standard for decades [1,2,3]. However, various kinds of metrological limitations are the source of inaccuracies [4,5,6]. The direct Fick method, or metabolic method, relies on the calculation of cardiac output as the ratio of oxygen uptake (V'O2) to the arteriovenous difference in oxygen content. It was originally used to validate the thermodilution method [7] and is often considered the 'physiological' gold standard. It cannot be taken as a clinical gold standard in intensive care practice because, although this has not been precisely assessed, there are possible causes of error specific to this setting, such as an increased oxygen consumption in the lungs in the presence of the acute respiratory distress syndrome or in the presence of pneumonia [8]. In addition, measuring V'O2 is not easy when the inspired fraction of oxygen is high. Another means to estimate cardiac output at the bedside is the echocardiographic approach, particularly from the transesophageal route. By visualizing the heart directly, the echocardiographic approach alleviates several drawbacks of other methods, but it is strongly operator dependent and thus may not always be readily available.

Comparisons of thermodilution cardiac output and metabolic cardiac output have demonstrated statistically significant correlations [7,9,10,11,12,13,14,15], but this does not mean 'agreement' or 'clinical interchangeability'. More recently, a satisfactory agreement has been found between the two methods in stable children [16] and in stable patients with pulmonary hypertension [17]. Other studies, however, have suggested that discrepancies could appear in less stable situations such as exercise [18] or critical illness [19,20].

In the present study, we re-examined the concordance between thermodilution cardiac output and metabolic cardiac output, for several reasons. First, ventilatory management in the intensive care unit has evolved; low tidal volume strategies currently being much more common than a few years ago. The corresponding permissive hypercapnia can have hemodynamic effects [21,22] and can interfere with the results of both the thermodilution and the metabolic methods. A second reason is that the controversy on the risk–benefit balance of right heart catheterization in critically ill patients [23,24] makes it important to gather knowledge about possible alternative methods. Finally, we wished to obtain data in a population of critically ill patients exhibiting indices of extreme severity, in whom cardiac output determination and manipulation are likely to be a more frequent issue than in other subsets of patients.

Materials and methods

Patients

Eighteen mechanically ventilated patients were studied (Table 1). Criteria for inclusion were: criteria 1, the presence of a flow-directed, balloon-tipped pulmonary artery catheter placed after the decision of the physician in charge of the patient; criteria 2, controlled mechanical ventilation without spontaneous respiratory activity; criteria 3, a stable level of inspired oxygen fraction and positive expiratory pressure when present; and criteria 4, spontaneous or drug-induced clinical unresponsiveness. Criteria 2–4 were set to minimize the risk of variations in oxygen consumption due to nonhemodynamic factors. When the patients received vascular expansion or when a change in the infusion rate of catecholamines was decided, a 10 min period of stability (<10% changes in cardiac frequency and arterial pressure) was required before the measurements were taken.

Table 1 Characteristics of the patients

The patients were recruited on a consecutive basis. The study was a byproduct of another study, relying on the same methods, and that fulfilled the French legal criteria for patient studies. With approval of the appropriate authority, informed consent was not sought because the study-related intervention was noninvasive and bore no risk of interference with the clinical management of the patients.

Measurements and calculations

Metabolic method

Oxygen consumption was determined from the measurements of carbon dioxide and oxygen concentrations in the inspired and expired gases, using a standard portable metabolic monitor (Deltatrac Metabolic Monitor™; Datex Instrumentation Corp., Helsinki, Finland) calibrated prior to each set of measurements with a 96% oxygen–4% carbon dioxide gas mixture. This monitor has been validated for accuracy, sensitivity and reproducibility over a wide range of conditions [18,25]. To retain a given measure for analysis, a 10 min 'metabolic' steady state was required (<5% change in the respiratory quotient [R], in V'O2, and in carbon dioxide production).

Blood gas analysis was performed on simultaneously drawn arterial and mixed venous samples (5 ml aliquots, with an AVL Omni 9™ analyzer; AVL Medical Instruments, Shaffhausen, Switzerland). The hemoglobin concentration and oxygen saturation were measured using the corresponding co-oximeter, as well as arterial and mixed venous oxygen contents (CaO2 and CO2, respectively). Cardiac output determined using the metabolic (Fick) method (QT FICK) was calculated as the ratio of V'O2 to the CaO2 – CO2 difference. Each of the QT FICK values used in the subsequent comparisons corresponded to 10 min measures of V'O2.

Thermodilution

From a flow-directed, balloon-tipped pulmonary artery catheter positioned in a nondependent zone of the lung [26], the cardiac output determined using the thermodilution method (QT THERM) was measured by fast injections of a 10 ml bolus of 5% dextrose solution, at room temperature. All the measurements were performed by the same operator. Each injection was performed at end expiration. The thermal decay curve was visually inspected extemporaneously, and the data were rejected if the curves were obviously aberrant and in the presence of waveform irregularities suggesting technical artifacts. Each of the QT THERM values used in the subsequent comparisons derives from three successive measures normalized according to Poon [27].

Tricuspid regurgitation

Pulsed Doppler echocardiography (parasternal short-axis view) was used to qualitatively detect a regurgitant signal in the right atrium.

Data analysis

Forty-nine paired measurements of QT FICK and of QT THERM were performed either at baseline or after a therapeutic intervention, with a minimum of two sets of measurements in each patient. The statistical association between QT FICK and QT THERM was expressed in terms of the Z coefficient of correlation with the 95% confidence interval. The agreement between the two techniques was studied using a graphical analysis according to Bland and Altman [28] and using the regression method described by Passing and Bablok [29]. This regression was first calculated using the whole data set. Data points lying far off the regression line were then tested for outlier status (data point considered outlier if value above mean + 3 SD of the data set not including this data point). Outliers so defined were removed from the data set and the regression recomputed. The analysis was conducted in the whole study population (18 patients, 49 pairs of measurements), over restricted ranges of cardiac output, and after exclusion of the patients with tricuspid regurgitation (14 patients remaining, 41 pairs of measurements).

The data are expressed as the mean ± SD.

Results

Whole population

The values for QT FICK ranged from 2.2 to 11 l/min (mean ± SD = 5.2 ± 2.0 l/min), whereas the values for QT THERM ranged from 2.8 to 11.2 l/min (mean ± SD = 5.8 ± 1.9 l/min) (R = 0.84, 95% confidence interval = 0.73–0.91, P < 0.0001). After the removal of one data point meeting the outlier definition (see Materials and methods), the mean difference between QT FICK and QT THERM was -0.8 l/min, with a lower limit of agreement (magnitude of underestimation of QT FICK by QT THERM) at -2.3 l/min and an upper limit (magnitude of overestimation of QT FICK by QT THERM) at 0.8 l/min (Fig. 1a). The results of the Passing and Bablok regression of QT FICK against QT THERM are shown in Figure 1b. The 95% confidence interval of the intercept did not include 0(-0.70 to -0.06) and the upper limit of the 95% confidence interval of the slope was equal to 1(0.87–1.00), indicating the existence of a systematic difference between the two techniques [29].

Figure 1
figure 1

Comparison of cardiac output determined using the thermodilution method (QT THERM) and cardiac output determined using the metabolic (Fick) method (QT FICK) according to (a) the Bland and Altman graphic method [28], and (b) the Passing and Bablok regression method [29]. Determined using the whole set of data after removal of one data point identified as an outlier (48 pairs obtained in the 18 patients), irrespective of the cardiac output value and of the presence of a tricuspid regurgitation. CI, confidence interval; SD, standard deviation.

For QT THERM values ≤ 5 l/min (n = 17, range = 2.8–5 l/min, mean ± SD = 3.8 ± 0.7), the correlation between the two methods was extremely strong (R = 0.93, 95% confidence interval = 0.81–0.97, P < 0.0001). The mean difference between QT FICK and QT THERM was -0.6 l/min, with a lower limit of agreement at -1.2 l/min and an upper limit at -0.1 l/min. The 95% confidence interval of the QT THERM versus QT FICK regression intercept included 0 (-0.89 to 0.32) and the 95% confidence interval of the slope included 1(0.77–1.07), indicating the absence of a systematic difference between the two techniques over that range of values (Fig. 2a) [29]. The QT FICK values never exceeded the QT THERM values.

Figure 2
figure 2

Passing and Bablok regression of cardiac output determined using the metabolic (Fick) method (QT FICK) against cardiac output determined using the thermodilution method (QT THERM) [29] restricted to (a) QT THERM values <5 l/min and (b) QT THERM values >5 l/min (after removal of one outlier). CI, confidence interval.

For QT THERM values >5 l/min (n = 34, range = 5.1–11.2 l/min, mean ± SD = 6.8 ± 1.3), the correlation between the two methods was weaker (R = 0.61, 95% confidence interval = 0.34–0.79, P < 0.0001) (Fig. 2b). After removal of the outlier, the mean difference between QT FICK and QT THERM was -0.9 l/min, with a lower limit of agreement at -2.7 l/min and an upper limit at 1.0 l/min.

Population restricted to patients without tricuspid regurgitation (n = 14)

The values for QT FICK ranged from 2.2 to 11 l/min (mean ± SD = 5.0 ± 1.9 l/min), whereas the values for QT THERM ranged from 2.8 to 11.2 l/min (mean ± SD = 5.8 ± 1.8 l/min) (R = 0.83, 95% confidence interval = 0.71–0.91, P < 0.0001). The mean difference between QT FICK and QT THERM was -0.8 l/min, with a lower limit of agreement (magnitude of underestimation of QT FICK by QT THERM) at -2.2 l/min and an upper limit (magnitude of overestimation of QT FICK by QT THERM) at 0.7 l/min (Fig. 3a). The Passing and Bablok regression of QT FICK against QT THERM (Fig. 3b) indicated a systematic difference between the techniques (confidence interval of the intercept =-0.70 to -0.21; confidence interval of the slope = 0.9–1.0). For QT THERM values <5 l/min, the mean difference between QT FICK and QT THERM was -0.6 l/min (range, -1.2 to -0.02 l/min). For QT THERM values >5 l/min, the mean difference between QT FICK and QT THERM was -0.8 l/min (-2.5 to 0.9 l/min).

Figure 3
figure 3

Comparison of cardiac output determined using the metabolic (Fick) method (QT FICK) and cardiac output determined using the thermodilution method (QT THERM) according to (a) the Bland and Altman graphic method [28], and (b) the Passing and Bablok regression method [29]. Restricted to the patients in whom cardiac echography ruled out tricuspid regurgitation (14 patients, 40 pairs of measurements, after removal of one outlier). CI, confidence interval; SD, standard deviation.

Discussion

The present study, conducted in a pragmatic manner to stay close to the clinical practice, shows that the bolus thermodilution method and the metabolic method can provide clinically interchangeable measures of low cardiac output values in mechanically ventilated, critically ill patients. Conversely, there are marked discrepancies between the two approaches for high cardiac output values.

Divergences between methods to estimate cardiac output in critically ill patients have been reported. Sherman et al. [19] found in 10 septic patients (average Acute Physiology and Chronic Health Evaluation [APACHE]II score = 18), as opposed to 10 nonseptic patients (average APACHEII score = 12), that the thermodilution cardiac output could overestimate the metabolic cardiac output by more than 6 l, or underestimate it by more than 3 l. In the study of Sherman et al., 17 out of 20 of the cardiac output values were >5 l/min.

Axler et al. [20] compared 45 pairs of measurements obtained in 13 patients of moderate severity (10 discharged alive from the intensive care unit, 3 deceased). In this series, transesophageal echocardiography, bolus thermodilution and the Fick method provided substantially different results. Although the thermodilution cardiac output values and the metabolic cardiac output values were not statistically different, their limits of agreement ranged from -2.7 to 4.8 l/min. From this, the authors insisted on the notion that clinical decision making could not rely on a cardiac output measurement alone, whatever the technique used to obtain it. In this series, only six metabolic cardiac output data points were <5 l/min.

The present study differs from the previous two studies by the extreme severity of the clinical status of the patients, as illustrated by high simplified acute physiology IIscores and a calamitous outcome (Table 1). Such clinical contexts are generally associated with complex hemodynamical situations, which may serve as a justification to the decision of right heart catheterization. Preliminary data obtained in a cohort of about 600 such patients [30] suggest that this procedure is not associated with an increased mortality, as opposed to what has been suspected in less severe patients [23,24]. Dhingra et al. [31] recently published a study similar to the present one regarding motives, design and methods. In 18 mechanically ventilated, critically ill patients with high APACHEII scores, these investigators showed that the thermodilution method and the metabolic method had limits of agreement ranging from -3.30 to 2.96 l/min. For cardiac output values >7 l/min, these limits were -5.67 to 1.87 l/min.

As compared with the data of Sherman et al. [19] and those of Axler et al. [20], the extreme severity of the patients' condition probably explains the relatively large proportion of low cardiac output values in the present data (Fig. 1) and in the data of Dhingra et al. [31]. Although splitting the data set in two parts carries the risks inherent to all post hoc analyses, it can clearly be seen from Figures 1 and 2 that the discrepancies between QT THERM and QT FICK become major only for high cardiac outputs. The agreement between QT THERM and QT FICK at cardiac output values <5 l/min was almost as good as that reported by Capderou et al. in normal individuals [16] (range -0.8 to-0.3 l/min), and QT THERM never underestimated QT FICK. In the study by Dhingra et al. [31], looking at the data suggests that the thermodilution method and the metabolic method were probably interchangeable up to 6 l/min. From a set of 105 measurements, among which 90 provided values <5 l/min, Hoeper et al. [17] reported limits of agreement between -1 and 1.2 l/min.

It appears that, in severely ill patients and in stable patients, a thermodilution cardiac output value <5 l/min probably reflects 'adequately' what this value would have been with the metabolic method, and vice versa. It must be noted that the meaning of 'adequately' here is arbitrary. The Bland and Altman graphical approach to compare two methods of measurements of a given biological value does not determine whether the agreement found between these two methods is 'good'. This depends on the error magnitude that is, arbitrarily, considered clinically acceptable. It seems to us that the degree of agreement reported by ourselves and others is sufficient to render reasonable a decision making process relying on a low cardiac output value, whatever the method used to obtain it. This is clinically relevant because, as emphasized by Dhingra et al. [31], "cardiac output manipulation is likely to have the greatest impact on outcome when cardiac output is low". It must be borne in mind, however, that the thermodilution method is notoriously unreliable when the cardiac output is very low. van Grondelle et al. [15] reported overestimates of cardiac output, with the thermodilution method reaching 35% of the measured value when the cardiac output was <2.5 l/min. Of note, we did not observe such low values in the present patients (Fig. 2).

The situation is different regarding the higher values of the cardiac output range that we observed. The acceptable agreement found at low values is clearly lost (Fig. 2). This is in line with the data of Sherman et al. [19], of Axler et al. [20] and of Dhingra et al. [31]. This is also in line with the results reported for cardiac output values >5 l/min by Koobi et al. [32] in stable adults in the context of a coronary artery bypass, and in line with the observations of Hsia et al. [33] in dogs and of Espersen et al. [18] in healthy humans, who described a dramatic decrease in agreement between the thermodilution method and the metabolic method when going from rest to exercise. The discrepancies between the thermodilution method and the metabolic method may be due to metrological limitations affecting both techniques, particularly in the intensive care setting. Of note, the presence of tricuspid regurgitation did not seem to have a major impact on the present results (Fig. 3), but it was relatively rare in our series.

We wish to emphasize that finding a low level of agreement between the thermodilution method and the metabolic method when the cardiac output is high does not necessarily mean that either of the two methods is closer than the other to the reality. Indeed, many sources of errors have been identified regarding the thermodilution method, and many publications have warned clinicians against them [6,15,34,35]. The metabolic method is also far from being free of criticism. In spite of the availability of easy-to-use metabolic carts, it remains difficult to use at the bedside. There is a risk to cumulate measurement errors (respiratory gas sampling and blood gas analysis). The reliability of the measurement of oxygen consumption can be decreased by metabolic instability, patient–ventilator dyssynchrony, high inspired oxygen fraction, circuit leaks, and so on. In addition, the metabolic method provides an accurate estimate of cardiac output only if the pulmonary artery flow, the mixed venous oxygen content, and the arterial oxygen content are reasonably constant [36], a condition that may not be fulfilled in hemodynamically compromised, mechanically ventilated patients. It is therefore not possible from the available data to designate a gold standard.

In summary, the present data concur with those of Dhingra et al. [31] to suggest that, in daily practice, a low thermodilution or metabolic cardiac output can reasonably be relied on to build a clinical decision, which is novel information. Conversely, both the present study and that of Dhingra et al. [31] confirm that, in critically ill patients, as in other types of patients, the methodological approach chosen to evaluate the cardiac output has an important influence on the result when cardiac output is high. High cardiac output values should thus be treated and used cautiously.

Key messages

  • This study confirms that the method chosen to evaluate cardiac output in critically ill patients can influence the results, and that this metrological dimension must be taken into account when interpreting clinical data

  • The good level of agreement between thermodilution measurement and metabolic measurement at low cardiac output suggests that such a value can be relied on to build a clinical decision, whatever the method used to determine it. This is novel information

  • Conversely, the divergence between methods for high cardiac output values prompts caution in the presence of such results

Abbreviations

APACHE:

Acute Physiology and Chronic Health Evaluation

CaO2:

arterial oxygen content

CvO2:

mixed venous oxygen content

QT FICK:

cardiac output determined using the metabolic (Fick) method

QT THERM:

cardiac output determined using the thermodilution method

R :

respiratory quotient

SD:

standard deviation

V'O2:

oxygen consumption.

References

  1. Ganz W, Donoso R, Marcus HS, Forrester JS, Swan HJ: A new technique for measurement of cardiac output by thermodilution in man. Am J Cardiol. 1971, 27: 392-396.

    Article  CAS  PubMed  Google Scholar 

  2. Goldenheim PD, Kazemi H: Current concepts. Cardiopulmonary monitoring of critically ill patients (1). N Engl J Med. 1984, 311: 717-720.

    Article  CAS  PubMed  Google Scholar 

  3. Goldenheim PD, Kazemi H: Cardiopulmonary monitoring of critically ill patients (2). N Engl J Med. 1984, 311: 776-780.

    Article  CAS  PubMed  Google Scholar 

  4. Levett JM, Replogle RL: Thermodilution cardiac output: a critical analysis and review of the literature. J Surg Res. 1979, 27: 392-404.

    Article  CAS  PubMed  Google Scholar 

  5. Nishikawa T: Hemodynamic changes associated with thermodilution cardiac output determination in canine acute blood loss or endotoxemia. Acta Anaesthesiol Scand. 1993, 37: 602-606.

    Article  CAS  PubMed  Google Scholar 

  6. Pinsky MR: The meaning of cardiac output. Intensive Care Med. 1990, 16: 415-417.

    Article  CAS  PubMed  Google Scholar 

  7. Fegler G: Measurement of cardiac output in anesthetized animals by a thermodilution method. Q J Exp Physiol. 1954, 39: 153-164.

    Article  CAS  PubMed  Google Scholar 

  8. Light RB: Intrapulmonary oxygen consumption in experimental pneumococcal pneumonia. J Appl Physiol. 1988, 64: 2490-2495.

    CAS  PubMed  Google Scholar 

  9. Enghoff E, Michaelsson M, Pavek K, Sjogren S: A comparison between the thermal dilution method and the direct Fick and the dye dilution methods for cardiac output measurements in man. Acta Soc Med Ups. 1970, 75: 157-170.

    CAS  PubMed  Google Scholar 

  10. Branthwaite MA, Bradley RD: Measurement of cardiac output by thermal dilution in man. J Appl Physiol. 1968, 24: 434-438.

    CAS  PubMed  Google Scholar 

  11. Olsson B, Pool J, Vandermoten P, Varnauskas E, Wassen R: Validity and reproducibility of determination of cardiac output by thermodilution in man. Cardiology. 1970, 55: 136-148.

    Article  CAS  PubMed  Google Scholar 

  12. Hodges M, Downs JB, Mitchell LA: Thermodilution and Fick cardiac index determinations following cardiac surgery. Crit Care Med. 1975, 3: 182-184.

    Article  CAS  PubMed  Google Scholar 

  13. Hoel BL: Some aspects of the clinical use of thermodilution in measuring cardiac output. With particular reference to the Swan–Ganz thermodilution catheters. Scand J Clin Lab Invest. 1978, 38: 383-388.

    Article  CAS  PubMed  Google Scholar 

  14. Bilfinger TV, Lin CY, Anagnostopoulos CE: In vitro determination of accuracy of cardiac output measurements by thermal dilution. J Surg Res. 1982, 33: 409-414.

    Article  CAS  PubMed  Google Scholar 

  15. van Grondelle A, Ditchey RV, Groves BM, Wagner WW, Reeves JT: Thermodilution method overestimates low cardiac output in humans. Am J Physiol. 1983, 245: H690-H692.

    CAS  PubMed  Google Scholar 

  16. Capderou A, Douguet D, Losay J, Zelter M: Comparison of indirect calorimetry and thermodilution cardiac output measurement in children. Am J Respir Crit Care Med. 1997, 155: 1930-1934.

    Article  CAS  PubMed  Google Scholar 

  17. Hoeper MM, Maier R, Tongers J, Niedermeyer J, Hohlfeld JM, Hamm M, Fabel H: Determination of cardiac output by the Fick method, thermodilution, and acetylene rebreathing in pulmonary hypertension. Am J Respir Crit Care Med. 1999, 160: 535-541.

    Article  CAS  PubMed  Google Scholar 

  18. Espersen K, Jensen EW, Rosenborg D, Thomsen JK, Eliasen K, Olsen NV, Kanstrup IL: Comparison of cardiac output measurement techniques: thermodilution, doppler, CO2-rebreathing and the direct Fick method. Acta Anaesthesiol Scand. 1995, 39: 245-251.

    Article  CAS  PubMed  Google Scholar 

  19. Sherman MS, Kosinski R, Paz HL, Campbell D: Measuring cardiac output in critically ill patients: disagreement between thermodilution-, calculated-, expired gas-, and oxygen consumption-based methods. Cardiology. 1997, 88: 19-25.

    Article  CAS  PubMed  Google Scholar 

  20. Axler O, Tousignant C, Thompson CR, Dall'ava-Santucci J, Phang PT, Russell JA, Walley KR: Comparison of transesophageal echocardiographic, Fick, and thermodilution cardiac output in critically ill patients. J Crit Care. 1996, 11: 109-116.

    Article  CAS  PubMed  Google Scholar 

  21. Rasmussen JP, Dauchot PJ, DePalma RG, Sorensen B, Regula G, Anton AH, Gravenstein JS: Cardiac function and hypercarbia. Arch Surg. 1978, 113: 1196-1200.

    Article  CAS  PubMed  Google Scholar 

  22. Pfeiffer B, Hachenberg T, Wendt M, Marshall B: Mechanical ventilation with permissive hypercapnia increases intrapulmonary shunt in septic and nonseptic patients with acute respiratory distress syndrome. Crit Care Med. 2002, 30: 285-289. 10.1097/00003246-200202000-00003.

    Article  PubMed  Google Scholar 

  23. Connors AF, Speroff T, Dawson NV, Thomas C, Harrell FE, Wagner D, Desbiens N, Goldman L, Wu AW, Califf RM, Fulkerson WJ, Vidaillet H, Broste S, Bellamy P, Lynn J, Knaus WA: The effectiveness of right heart catheterization in the initial care of critically ill patients. Support investigators. JAMA. 1996, 276: 889-897. 10.1001/jama.276.11.889.

    Article  PubMed  Google Scholar 

  24. Polanczyk CA, Rohde LE, Goldman L, Cook EF, Thomas EJ, Marcantonio ER, Mangione CM, Lee TH: Right heart catheterization and cardiac complications in patients undergoing noncardiac surgery: an observational study. JAMA. 2001, 286: 309-314. 10.1001/jama.286.3.309.

    Article  CAS  PubMed  Google Scholar 

  25. Konishi T, Nakamura Y, Morii I, Himura Y, Kumada T, Kawai C: Comparison of thermodilution and Fick methods for measurement of cardiac output in tricuspid regurgitation. Am J Cardiol. 1992, 70: 538-539.

    Article  CAS  PubMed  Google Scholar 

  26. Teboul J, Besbes M, Andrivet P, Axler O, Douguet D, Zelter M, Lemaire F, Brun-Buisson C: A bedside index assessing the reliability of pulmonary occlusion pressure measurements during mechanical ventilation with positive end-expiratory pressure. J Crit Care. 1992, 7: 22-29.

    Article  Google Scholar 

  27. Poon CS: Analysis of linear and mildly nonlinear relationships using pooled subject data. J Appl Physiol. 1988, 64: 854-859.

    CAS  PubMed  Google Scholar 

  28. Bland J, Altman D: Statistical methods for assessing agreement between two methods of clinical measurement. Lancet. 1986, i: 307-310.

    Article  Google Scholar 

  29. Passing H, Bablok W: A new biometrical procedure for testing the equality of measurements from two different analytical methods. Application of linear regression procedures for method comparison studies in clinical chemistry, part i. J Clin Chem Clin Biochem. 1983, 21: 709-720.

    CAS  PubMed  Google Scholar 

  30. Richard C, Warzawski J: French multicenter clinical trial of efficacy and safety of pulmonary artery catheter (PAC) in shock and/or ARDS: preliminary results [abstract]. Am J Respir Crit Care Med. 2002, 165: A22-

    Article  Google Scholar 

  31. Dhingra VK, Fenwick JC, Walley KR, Chittock DR, Ronco JJ: Lack of agreement between thermodilution and Fick cardiac output in critically ill patients. Chest. 2002, 122: 990-997. 10.1378/chest.122.3.990.

    Article  PubMed  Google Scholar 

  32. Koobi T, Kaukinen S, Ahola T, Turjanmaa VM: Non-invasive measurement of cardiac output: whole-body impedance cardiography in simultaneous comparison with thermodilution and direct oxygen Fick methods. Intensive Care Med. 1997, 23: 1132-1137. 10.1007/s001340050469.

    Article  CAS  PubMed  Google Scholar 

  33. Hsia CC, Herazo LF, Ramanathan M, Johnson RL: Cardiac output during exercise measured by acetylene rebreathing, thermodilution, and Fick techniques. J Appl Physiol. 1995, 78: 1612-1616.

    CAS  PubMed  Google Scholar 

  34. Nishikawa T, Dohi S: Errors in the measurement of cardiac output by thermodilution. Can J Anaesth. 1993, 40: 142-153.

    Article  CAS  PubMed  Google Scholar 

  35. Bazaral MG, Petre J, Novoa R: Errors in thermodilution cardiac output measurements caused by rapid pulmonary artery temperature decreases after cardiopulmonary bypass. Anesthesiology. 1992, 77: 31-37.

    Article  CAS  PubMed  Google Scholar 

  36. Wood E, Bowers D, Shepherd J, Fox I: Oxygen content of mixed venous blood in man during various phases of the respiratory and cardiac cycles in relation to possible errors in measurements of cardiac output by conventional application of the Fick method. J Appl Physiol. 1955, 7: 621-628.

    CAS  PubMed  Google Scholar 

Download references

Acknowledgements

The authors are indebted to the nursing staff for having made this clinical research possible. The study was supported by Association pour le Développement et l'Organisation de la Recherche en Pneumologie (ADOREP), Paris, France. JG was a scholar of the Société de Pneumologie de Langue Française, Paris, France.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Thomas Similowski.

Additional information

Competing interests

None declared.

Rights and permissions

Reprints and permissions

About this article

Cite this article

Gonzalez, J., Delafosse, C., Fartoukh, M. et al. Comparison of bedside measurement of cardiac output with the thermodilution method and the Fick method in mechanically ventilated patients. Crit Care 7, 171 (2002). https://doi.org/10.1186/cc1848