Determinants of gas exchange during extracorporeal CO2 removal using a novel pump-driven venovenous gas exchange system in a minimally invasive setting
Critical Care volume 19, Article number: P274 (2015)
Introduction
Pump-driven venovenous extracorporeal CO2 removal (ECCO2-R) increasingly takes root in hypercapnic lung failure to minimize ventilation invasiveness or to avoid intubation. A recently developed miniaturized device consisting of a centrifugal pump and a membrane ventilator (iLA Activve®; Novalung, Germany) allows effective decarboxylation via a jugular double lumen cannula. So far no data on gas exchange in this setting exist to date.
Methods
We included 10 patients receiving iLA Activve® due to hypercapnic respiratory failure as bridge-to-transplant or obstructive lung disease. Sweep gas flow was increased in steps from 1 to 14 l/ minute at constant blood flow (phase 1). Similarly, blood flow was gradually increased at constant sweep gas flow (phase 2). At each step, gas transfer via the membrane as well as arterial blood gas samples were obtained.
Results
During phase 1, we observed a significant increase in CO2 transfer together with a decrease in PaCO2 levels from a median of 66 mmHg (range 46 to 85) to 49 (31 to 65) mmHg from 1 to 14 l/ minute sweep gas flow, while arterial oxygenation deteriorated with high sweep gas flow rates. During phase 2, oxygen transfer significantly increased leading to an increase in PaO2 from 67 (49 to 87) at 0.5 l/ minute to 117 (66 to 305) mmHg at 2.0 l/minute. Higher blood flow rates also significantly enhanced decarboxylation. Increasing blood flow to 2.0 l/minute led to negative suction pressures of more than -100 mmHg and signs of hemolysis. See Figure 1.
Conclusion
Increasing sweep gas flow results in effective CO2 removal which can be further reinforced by raising blood flow. The clinically relevant oxygenation effect even in this setting of low invasivity could broaden the range of indications towards hypercapnic lung failure with mild to moderate hypoxia.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Hermann, A., Riss, K., Schellongowski, P. et al. Determinants of gas exchange during extracorporeal CO2 removal using a novel pump-driven venovenous gas exchange system in a minimally invasive setting. Crit Care 19 (Suppl 1), P274 (2015). https://doi.org/10.1186/cc14354
Published:
DOI: https://doi.org/10.1186/cc14354
Keywords
- Obstructive Lung Disease
- Moderate Hypoxia
- Hypercapnic Respiratory Failure
- PaCO2 Level
- Negative Suction