- Poster presentation
- Published:
Pressure reconstruction method for spontaneous breathing effort monitoring
Critical Care volume 19, Article number: P259 (2015)
Introduction
Estimating respiratory mechanics of mechanically ventilated (MV) patients is unreliable when patients exhibit spontaneous breathing (SB) efforts on top of ventilator support. This reverse triggering effect [1] results in an M-wave shaped pressure wave. A model-based method to reconstruct the affected airway pressure curve is presented to enable estimation of the true underlying respiratory mechanics of these patients.
Methods
Airway pressure and flow data from 72 breaths of a pneumonia patient were used for proof of concept. A pressure wave reconstruction method fills parts of the missing area caused by SB efforts and reverse triggering by connecting the peak pressure and end-inspiration slope (Figure 1). A time-varying elastance model [2] was then used to identify underlying respiratory elastance (AUCE drs ). The area of the unreconstructed M-wave has less pressure, resulting in a lower overall AUCE drs without reconstruction. The missing area of the airway pressure or AUCE drs is hypothesized to be a surrogate of patient-specific inspiratory to assess the strength of SB efforts. AUCE drs and missing area A2 are compared with/without reconstruction.
Results
Median AUCE drs and breath-specific effort using reconstruction were 24.99 (IQR: 22.90 to 25.98) cmH2O/l and 3.64 (IQR: 0.00 to 3.87)% versus AUCE drs of 20.87 (IQR: 15.24 to 27.48) cmH2O/l for unreconstructed M-wave data, indicating significant patient and breath-specific SB effort, and the expected higher elastance (P < 0.05).
Conclusion
A simple reconstruction method enables the real-time measurement of respiratory system properties of a SB patient and measures the surrogate of the SB effort, that latter of which has clinical use in deciding whether to extubate or re-sedate the patient.
References
Akoumianaki E, et al: Chest. 2013, 143: 927-38. 10.1378/chest.12-1817.
Chiew YS, et al: Biomed Eng Online. 2011, 10: 111-10.1186/1475-925X-10-111.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Damanhuri, N., Chiew, Y., Othman, N. et al. Pressure reconstruction method for spontaneous breathing effort monitoring. Crit Care 19 (Suppl 1), P259 (2015). https://doi.org/10.1186/cc14339
Published: