Skip to main content
  • Poster presentation
  • Published:

Evaluation of the microbial tightness of closed system transfer devices by simulating airborne and touch contamination

Introduction

The use of intravascular catheter devices is often associated with serious bloodstream infections due to microbial contaminations. To minimize risk of such infections NIOSH recommends the use of closed system transfer devices (CSTDs). To evaluate the microbial tightness of CSTDs we developed two methods which simulate the bioburden in ambient air of operating rooms and ICUs.

Methods

The methods simulate airborne and touch contamination. We tested the microbial tightness of the integrated Safeflow® valve of a Mini-Spike® which is used for drug admixture. The airborne contamination was done in an exposure chamber in which a nebulizer distributed defined B. subtilis spore aerosols [1]. A Mini-Spike® was inserted into a vial of 0.9% sodium chloride solution (NaCl). A nebulizer with a suspension of 4.8 × 105 CFU spores of B. subtilis per ml was used to generate an aerosol for 1 minute. The volume of B. subtilis suspension nebulized per minute was 0.278 ml. This corresponds to 1.34 × 103 aerosolized spores in the exposure chamber, which has a volume of 0.24 m3 (5.6 × 103 CFU per m3 air). The used concentration was 100 times higher than the microbial burden found in hospitals [2]. After nebulization the valve was disinfected and NaCl was withdrawn into a syringe at certain time intervals. The NaCl was incubated on tryptic soy agar at 37°C for 48 hours. Results were documented as CFU. For touch contamination, a Mini-Spike® was attached to a vial of NaCl. The valve of the Mini-Spike® was contaminated with 105 CFU Staphylococcus aureus. The subsequent procedure was done as described above.

Results

Out of nine tested valves, none showed transmission of B. subtilis spores after airborne contamination. Three out of nine tested valves were contaminated with S. aureus after touch contamination.

Conclusion

Our study shows that both Methods are suitable for evaluating the microbial tightness of CSTDs.

References

  1. Dunkelberg H, Fleitmann-Glende F: Measurement of the microbial barrier effectiveness of sterilization containers in terms of the log reduction value for prevention of nosocomial infections. Am J Infect Control. 2006, 34: 285-9. 10.1016/j.ajic.2005.09.012.

    Article  PubMed  Google Scholar 

  2. Qudiesat K: Assessment of airborne pathogens in healthcare settings. AJMR. 2009, 3: 66-76.

    Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Gebel, J. Evaluation of the microbial tightness of closed system transfer devices by simulating airborne and touch contamination. Crit Care 19 (Suppl 1), P85 (2015). https://doi.org/10.1186/cc14165

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc14165

Keywords