pH testing to confirm nasogastric tube position on the ICU: are we wasting our time?
Critical Care volume 17, Article number: P245 (2013)
Introduction
For such a simple procedure, the insertion and position checking of nasogastric (NG) tubes can be very problematic. The UK's National Patient Safety Agency declared that 'Placement of NG tubes together with confirmation of correct placement can carry significant risks' and recommends that measuring the pH of NG aspirate should be the first-line method of determining correct NG position (safe range ≤5.5) [1].
Methods
In order to assess the usefulness of pH testing of NG aspirates in critical care patients, a prospective survey of pH testing of NG tube aspirate was carried out. This was undertaken both in patients who had a newly placed NG tube and in patients who were having regular/ routine checks of their existing NG tube.
Results
A total of 168 separate pH readings in 41 ICU patients receiving continuous enteral nutrition were recorded. In total, 137/168 patients were receiving proton pump inhibitors (PPIs). Eighteen readings were from newly placed NG tubes and 150 readings from old NG tubes. Fifty-three per cent of routine pH readings were falsely high; that is, pH 6 or above despite the NG tube being in the stomach (Figure 1). Twenty-eight per cent of newly placed NG tubes had falsely high pH readings (Figure 2).
Conclusion
In this population of ICU patients, routine/daily checks of NG pH aspirate appear to be limited. This is almost certainly due to the use of continuous NG feed together with PPIs. The usefulness of pH testing in newly placed NG tubes, however, appears more reliable.
References
National Patient Safety Agency:Reducing the Harm Caused by Misplaced Nasogastric Feeding Tubes. 2011. NPSA/PSA001./PSA002. NPSA; 2005 [http://www.nrls.npsa.nhs.uk]
Author information
Authors and Affiliations
Rights and permissions
This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Temblett, P., George, S. pH testing to confirm nasogastric tube position on the ICU: are we wasting our time?. Crit Care 17 (Suppl 2), P245 (2013). https://doi.org/10.1186/cc12183
Published:
DOI: https://doi.org/10.1186/cc12183
Keywords
- Enteral Nutrition
- Critical Care Patient
- National Patient Safety Agency
- Position Check
- Receive Proton Pump Inhibitor