Device for measuring bronchodilator delivery and response in resource-limited settings
Critical Care volume 17, Article number: P165 (2013)
Introduction
Effective delivery of aerosolized bronchodilators for patients with asthma is crucial for adequate therapy in critical care and emergent settings. Often administered with pressure-metered dose inhalers (pMDIs), bronchodilator delivery depends on the correct patient technique during administration [1] and the ability to measure treatment response, which are difficult to monitor at the point of care and particularly so in resource-poor settings where standard in-hospital monitoring is unavailable [2].
Methods
A point-of-care device for airflow measurement during bronchodilator delivery was designed and tested for use in resourcelimited settings. The handheld device was constructed from a clinical aerosol delivery tube with a bidirectional sensor for pressure differential detection about the aerosol element (Figure 1). The custom low-cost (<$11) electronics were designed such that no power supply would be needed apart from a computer or mobile device. Collection of airflow signal and calibration was performed with a standard 3 l syringe with flow volume measurement on one adult subject.
Results
Calibration of the assembled device over range of 0.5 to 3 l/ second enabled conversion of pressure difference to air flow with a mean measured tube resistance of 0.258 (kPa*second)/l. Robust signal responses to an adult subject's continuous respiratory maneuvers on the tube itself were demonstrated. Subject performance of the pMDI technique with subsequent tidal volume breathing was recorded and analyzed (Figure 1).
Conclusion
Design and calibration of a novel low-cost monitoring device for bronchodilator delivery monitoring enabled detection and recording of characteristic flow volume respiratory patterns for point-of-care diagnostics in resource-limited settings. Future work will require clinical testing and automated detection of the correct pMDI patient technique.
References
Ari A, et al.: Respir Care. 2012, 57: 613-626. 10.4187/respcare.01756
Riviello E, et al.: Crit Care Med. 2011, 39: 860. 10.1097/CCM.0b013e318206d6d5
Author information
Authors and Affiliations
Rights and permissions
This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Carspecken, C., Talmor, D. Device for measuring bronchodilator delivery and response in resource-limited settings. Crit Care 17 (Suppl 2), P165 (2013). https://doi.org/10.1186/cc12103
Published:
DOI: https://doi.org/10.1186/cc12103