- Poster presentation
- Open access
- Published:
Resistance of neonatal endotracheal tubes: a comparison of four commercially available types
Critical Care volume 17, Article number: P160 (2013)
Introduction
In mechanically ventilated neonates the flow-dependent resistance of the endotracheal tube (ETT) causes a noticeable pressure difference between airway and tracheal pressure [1]. This may potentially lead to retardation of the passive driven expiration and dynamic lung inflation consecutively but more importantly increases the work of breathing. The aim of this study was to compare the resistive pressure drop of four commercially available neonatal ETTs with internal diameter (ID) of 2.0 mm.
Methods
The pressure-flow relationship of neonatal ETTs (internal diameter 2.0 mm) of four different manufacturers (Mallinckrodt, Hi-Contour, Covidien, Dublin, Ireland; Portex, Cole's Neonatal Tube, Smith Medical, St Paul, MN, USA; Rusch, Silko Clear, Teleflex Medical, Kernen, Germany; and Vygon, Pediatric Endotracheal Tube, Ecouen, France) was determined in a physical model consisting of a tube connector, an anatomically curved ETT and an artificial trachea. The model was ventilated with a sinusoidal gas flow with an amplitude of ±100 ml/ second and a ventilation rate ranging from 40 to 60 cycles/minute. The coefficients of an approximation equation considering ETT resistance were fitted separately to the measured pressure-flow curves for inspiration and expiration.
Results
The pressure drop profiles of all ETTs were nonlinearly flow dependent. The expiratory pressure drop (Pexp) slightly exceeded the inspiratory one (Pinsp). The ETT of Portex Cole's Neonatal Tube had a significant lower pressure drop (Pinsp 8.8 ± 0.3 cmH2O vs. Pexp -11.2 ± 0.2 cmH2O at an air flow of 100 ml/second) compared with all other ETTs (Mallinckrodt: Pinsp 19.2 ± 0.4 cmH2O vs. Pexp -24.3 ± 0.7 cmH2O; Rüsch: Pinsp 28.3 ± 0.8 cmH2O vs. Pexp -33.0 ± 0.8 cmH2O; Vygon: Pinsp 19.2 ± 0.4 cmH2O vs. Pexp -22.6 ± 0.6 cmH2O, all P < 0.05) for all flow and ventilation rates, respectively.
Conclusion
The ETT resistance highly contributes to the total airway resistance in neonatal ventilation. The flow-dependent pressure drop of shouldered Cole's Neonatal Tube (Portex) was up to 70% less in inspiration (resp. 67% in expiration) compared with straight tubes with an internal diameter corresponding to the narrow part of Cole's tube. We conclude that neonatal intubation with a Cole's tube can clearly reduce the resistive load due to the endotracheal tube and thus potentially prevent additionally work of breathing.
References
Guttmann J, et al.: Crit Care Med. 2000, 28: 1018-1026. 10.1097/00003246-200004000-00018
Author information
Authors and Affiliations
Rights and permissions
This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Spaeth, J., Steinmann, D., Guttmann, J. et al. Resistance of neonatal endotracheal tubes: a comparison of four commercially available types. Crit Care 17 (Suppl 2), P160 (2013). https://doi.org/10.1186/cc12098
Published: