Benchtop study comparing leakages across cuffs of three endotracheal tubes

Introduction

The aim was to compare two novel endotracheal tubes (ETT), Mallinckrodt TaperGuard (TG, tapered polyvinyl chloride (PVC) cuff) and KimVent Microcuff (MC, cylindrical polyurethrane cuff), with conventional Portex (PT, globular PVC cuff) in leakages across cuffs (microaspiration) under simulated clinical situations. It has been shown that globular PVC cuffs protect poorly against leakages due to microchannels formed from infolding of redundant cuff material [1]. We hypothesized that TG and MC better prevent microaspiration, which is a major mechanism of ventilator-associated pneumonia (VAP).

Methods

Each ETT was inserted into a silicone cylinder of 2 cm wide inclined at 35°. Then 20 ml water was added above the cuff and leakage measured every minute under five different simulated clinical conditions: mechanical ventilation for acute severe asthma (positive end-expiratory pressure (PEEP) 0 cmH2O), normal lungs (PEEP 5 cmH2O) and acute respiratory distress syndrome (PEEP 10 cmH2O), and disconnection from the ventilator with and without spontaneous breathing effort. Spontaneous breathing was simulated with a respiratory gas exchange simulator. Suction was applied at 200 cmH2O sustained for 3 minutes at the Murphy eye. Each scenario was repeated with cuff pressures (Pcuff) 10, 20 and 30 cmH2O maintained by a Pcuff maintenance device.

Results

PT leaked grossly in all scenarios without PEEP and at PEEP 5 cmH2O in the presence of suction irrespective of Pcuff (Pcuff 30 cmH2O: PT 19.7 ± 0.4, TG 0.0 ± 0.1, MC 0.0 ± 0.0, P < 0.001; Pcuff 20 cmH2O: PT 19.9 ± 0.4, TG 7.4 ± 6.2, MC 0.0 ± 0.0, P < 0.001; Pcuff 10 cmH2O: PT 20.0 ± 0.0, TG 12.7 ± 5.1, MC 0.9 ± 0.8, P < 0.001). Leakage under these scenarios can be reduced in TG and prevented in MC by Pcuff ≥20 cmH2O (Figure 1).

Figure 1
figure 1

Downward leak in PEEP 0 cmH 2 O without suction. Each point represents the mean and standard deviation of eight repeated measures.

Conclusion

Microcuff outperformed the others in preventing microaspiration, while Portex leaked grossly even at a recommended Pcuff of 20 to 30 cmH2O whenever PEEP was lost. The effect of ETT type on the incidence of VAP warrants further investigation.

References

  1. Fernandez JF, et al.: Chest. 2012, 142: 231-238. 10.1378/chest.11-2420

    Article  PubMed Central  PubMed  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Lam, S., Lau, C. & Yan, W. Benchtop study comparing leakages across cuffs of three endotracheal tubes. Crit Care 17 (Suppl 2), P152 (2013). https://doi.org/10.1186/cc12090

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc12090

Keywords