Benchtop study comparing leakages across cuffs of three endotracheal tubes
Critical Care volume 17, Article number: P152 (2013)
Introduction
The aim was to compare two novel endotracheal tubes (ETT), Mallinckrodt TaperGuard (TG, tapered polyvinyl chloride (PVC) cuff) and KimVent Microcuff (MC, cylindrical polyurethrane cuff), with conventional Portex (PT, globular PVC cuff) in leakages across cuffs (microaspiration) under simulated clinical situations. It has been shown that globular PVC cuffs protect poorly against leakages due to microchannels formed from infolding of redundant cuff material [1]. We hypothesized that TG and MC better prevent microaspiration, which is a major mechanism of ventilator-associated pneumonia (VAP).
Methods
Each ETT was inserted into a silicone cylinder of 2 cm wide inclined at 35°. Then 20 ml water was added above the cuff and leakage measured every minute under five different simulated clinical conditions: mechanical ventilation for acute severe asthma (positive end-expiratory pressure (PEEP) 0 cmH2O), normal lungs (PEEP 5 cmH2O) and acute respiratory distress syndrome (PEEP 10 cmH2O), and disconnection from the ventilator with and without spontaneous breathing effort. Spontaneous breathing was simulated with a respiratory gas exchange simulator. Suction was applied at 200 cmH2O sustained for 3 minutes at the Murphy eye. Each scenario was repeated with cuff pressures (Pcuff) 10, 20 and 30 cmH2O maintained by a Pcuff maintenance device.
Results
PT leaked grossly in all scenarios without PEEP and at PEEP 5 cmH2O in the presence of suction irrespective of Pcuff (Pcuff 30 cmH2O: PT 19.7 ± 0.4, TG 0.0 ± 0.1, MC 0.0 ± 0.0, P < 0.001; Pcuff 20 cmH2O: PT 19.9 ± 0.4, TG 7.4 ± 6.2, MC 0.0 ± 0.0, P < 0.001; Pcuff 10 cmH2O: PT 20.0 ± 0.0, TG 12.7 ± 5.1, MC 0.9 ± 0.8, P < 0.001). Leakage under these scenarios can be reduced in TG and prevented in MC by Pcuff ≥20 cmH2O (Figure 1).
Conclusion
Microcuff outperformed the others in preventing microaspiration, while Portex leaked grossly even at a recommended Pcuff of 20 to 30 cmH2O whenever PEEP was lost. The effect of ETT type on the incidence of VAP warrants further investigation.
References
Fernandez JF, et al.: Chest. 2012, 142: 231-238. 10.1378/chest.11-2420
Author information
Authors and Affiliations
Rights and permissions
This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Lam, S., Lau, C. & Yan, W. Benchtop study comparing leakages across cuffs of three endotracheal tubes. Crit Care 17 (Suppl 2), P152 (2013). https://doi.org/10.1186/cc12090
Published:
DOI: https://doi.org/10.1186/cc12090
Keywords
- Endotracheal Tube
- Acute Respiratory Distress Syndrome
- Polyvinyl Chloride
- Spontaneous Breathing
- Acute Severe Asthma