Lung elastance can be determined without esophageal pressure measurements

Introduction

We have previously shown, in an ex vivo porcine model, that lung elastance calculated as the PEEP change divided by lung volume increase (ΔPEEP/ΔEELV) is closely correlated to conventionally measured lung elastance using oesophageal pressure [1]. In this study we hypothesize that the successive change in lung volume during a PEEP-step manoeuvre could be predicted from ΔPEEP and lung elastance as ΔPEEP/EL. The objective of the study was to validate this hypothesis in patients with acute respiratory failure (ARF).

Methods

Thirteen patients with ARF were studied during an incremental PEEP trial, 0-4-8-12-16 cmH2O. ΔEELV was determined as the change in expiratory tidal volume following each PEEP step. Conventional calculation of lung elastance was obtained from tidal variation in airway pressure minus tidal variation in oesophageal pressure divided by tidal volume. Position of the oesophageal catheter was verified according to Baydur [2]. The measured change in end-expiratory lung volume during the PEEP-step manoeuvre using spirometry was compared with the end-expiratory lung volume change calculated from EL and stepwise changes in PEEP as ΔPEEP/EL.

Results

There was a close correlation between the measured build-up of end-expiratory lung volume during a PEEP-step manoeuvre and ΔPEEP/EL where EL was conventionally determined using oesophageal pressure measurements (see Figure 1).

Figure 1
figure 1

Correlation between spirometrically measured and calculated increase in EELV.

Conclusion

Esophageal pressure measurements are difficult to perform [3] and rarely used in routine clinical practice. Our findings indicate that a change in PEEP together with measurements of the resulting change in end-expiratory volume by spirometry in the ventilator could be used to determine lung elastance separately, the relation between lung and chest wall elastance as well as the transpulmonary pressure.

References

  1. Stenqvist O, et al.: Acta Anaesthesiol Scand. 2012, 56: 738-747. 10.1111/j.1399-6576.2012.02696.x

    Article  CAS  PubMed  Google Scholar 

  2. Baydur A: Chest. 2002, 121: 324-326. 10.1378/chest.121.2.324

    Article  PubMed  Google Scholar 

  3. Hedenstierna G: Minerva Anestesiol. 2012, 78: 959-966.

    CAS  PubMed  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and Permissions

About this article

Cite this article

Grivans, C., Lundin, S. & Stenqvist, O. Lung elastance can be determined without esophageal pressure measurements. Crit Care 17 (Suppl 2), P130 (2013). https://doi.org/10.1186/cc12068

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc12068

Keywords

  • Acute Respiratory Failure
  • Tidal Variation
  • Lung Elastance
  • Oesophageal Pressure
  • Lung Volume Increase