A method for detection and quantification of hydroxyethyl starch in plasma

  • Gordon P Otto1, 2Email author,
  • Bianka Wissuwa1, 3,
  • Anne Mehnert1,
  • Sascha Marx1,
  • Michael Raessler4,
  • Wolfgang Lösche1, 3,
  • Ralf A Claus1, 3 and
  • Maik Sossdorf1, 3
Critical Care201216:426

https://doi.org/10.1186/cc11302

Published: 8 May 2012

Keywords

Plasma SampleTrichloroacetic AcidAmylosePerchloric AcidCentral Venous Catheter

The use of hydroxyethylstarch (HES) is a controversial issue due to increasing evidence that HES accumulates in plasma and various tissues and therefore leads to unfavourable outcome in critically ill patients [1, 2]. No simple methods are available for monitoring HES plasma levels; present technologies to measure HES are based on gas chromatography-mass spectrometry, which are time-consuming and need advanced equipment [3]. Here, we applied Lugol's iodine solution (LUGOL; Sigma-Aldrich, Steinheim, Germany) to determinate HES concentrations in plasma and compare the results with those obtained by high-performance liquid chromatography (HPLC).

The study was approved by the institutional ethical committee (Friedrich-Schiller-University, Jena) as well as by the animal welfare committee (Thüringer Landesamt für Lebensmittelsicherheit und Verbraucherschutz, Germany). Blood samples from healthy volunteers were diluted with balanced 6% HES 130/0.4 (Fresenius Kabi, Bad Homburg, Germany) to get concentrations up to 30 mg/ml. Samples were centrifuged (10 minutes, 4,700 × g, 4°C), and plasma aliquots were mixed with 10% trichloroacetic acid and re-centrifuged (5 minutes, 4,700 × g, 4°C). The supernatant was mixed (4:1) with LUGOL and optical density was measured at 530 nm. For in vivo experiments, 5 ml or 10 ml of 6% HES 130/0.4 was infused into wistar rats over 1 hour via a central venous catheter. Blood samples were obtained prior and up to 24 hours after infusion. Plasma samples were prepared as described. For HPLC analysis, plasma samples were diluted 1:10 with water, then mixed with 60% perchloric acid, heated at 90°C for 60 minutes, diluted again (1:50) and subjected to HPLC [4]. For correlation analysis the Pearson correlation coefficient was calculated.

Using the LUGOL method, we found a linear correlation (r2 > 0.99) between calculated and measured plasma HES concentrations after serial dilutions of HES 130/0.4 in human whole blood (Figure 1a). At a concentration of 6 mg/ml the half life of HES 130/0.4 was estimated to be 1.8 hours (Figure 1b). In rats, plasma levels of 2.9 mg/ml and 4.6 mg/ml were measured immediately after infusion of 5 or 10 ml HES 130/0.4. Two hours after infusion the measured levels were only 10% of the initial levels (Figure 2). Comparing the plasma concentrations measured by LUGOL and HPLC revealed a correlation coefficient of r2 > 0.90 (P < 0.001).
Figure 1
Figure 1

Quantification of HES by Lugol's iodine solution. (a) Measurement of HES concentration using LUGOL's iodine solution in samples of human whole blood incubated with different HES 130/0.4 concentrations. (b) Time course of HES concentration in human whole blood after incubating with HES 130/0.4 at a final concentration of 6 mg/ml.

Figure 2
Figure 2

Time course of HES plasma levels after infusion of 5 or 10 ml HES 130/0.4 solution over 1 hour. Significance was tested using ANOVA (*P < 0.05).

The new method described here allows the rapid and simple detection of HES in plasma samples. LUGOL is known to stain polysaccharides, including amylose, amylopectin and glycogen. Amylose and amylopectin, due to their herbal origin, should not be found in blood. However, glycogen might interfere with the measurements, but it is present in only negligible amounts in plasma [5].

Measurement of HES plasma levels in patients could help to gain new insights into HES degradation and plasma accumulation after infusion.

Abbreviations

HES: 

hydroxyethyl starch

HPLC: 

high-performance liquid chromatography

LUGOL: 

Lugol's iodine staining solution.

Declarations

Acknowledgements

The authors would like to thank Barbara Schmidt and Brigitte Specht for her excellent technical support.

Authors’ Affiliations

(1)
Clinic for Anesthesiology and Intensive Care, Jena University Hospital, Jena, Germany
(2)
Clinic for Internal Medicine (KIM III) - Nephrology, Jena University Hospital, Jena, Germany
(3)
Center for Sepsis Control and Care (CSCC), Jena University Hospital,, Jena, Germany
(4)
Max-Planck-Institut fuer Biogeochemie, Jena, Germany

References

  1. Brunkhorst FM, Engel C, Bloos F, Meier-Hellmann A, Ragaller M, Weiler N, Moerer O, Gruendling M, Oppert M, Grond S, Olthoff D, Jaschinski U, John S, Rossaint R, Welte T, Schaefer M, Kern P, Kuhnt E, Kiehntopf M, Hartog C, Natanson C, Loeffler M, Reinhart K, German Competence Network Sepsis (SepNet): Intensive insulin therapy and pentastarch resuscitation in severe sepsis. N Engl J Med 2008, 358: 125-139.View ArticlePubMedGoogle Scholar
  2. Hartog CS, Kohl M, Reinhart K: A systematic review of third-generation hydroxyethyl starch (HES 130/0.4) in resuscitation: safety not adequately addressed. Anesth Analg 2011, 112: 635-645.View ArticlePubMedGoogle Scholar
  3. Deventer K, Van Eenoo P, Delbeke FT: Detection of hydroxyethylstarch (HES) in human urine by liquid chromatography-mass spectrometry. J Chromatogr B Analyt Technol Biomed Life Sci 2006, 834: 217-220.View ArticlePubMedGoogle Scholar
  4. Raessler M, Wissuwa B, Breul A, Unger W, Grimm T: Determination of water-extractable nonstructural carbohydrates, including inulin, in grass samples with high-performance anion exchange chromatography and pulsed amperometric detection. J Agric Food Chem 2008, 56: 7649-7654.View ArticlePubMedGoogle Scholar
  5. Lutkic A, Cepo M, Dodikovic V, Zjacic-Rotkvic V, Goldoni V, Mazuranic V, Cabrijan T, Cizmic A, Jurilj N: Concentration of glycogen in blood of diabetics. Clin Investig 1992, 70: 707.View ArticlePubMedGoogle Scholar

Copyright

© BioMed Central Ltd 2012

Advertisement