Performances of ventilator at simulated altitude

Introduction

We have assessed the ability of three ventilators to deliver to a normal lung model a set tidal volume (Vt) at different simulated cabin altitudes. We studied the performance of the LTV-1200 (Viasys Healthcare, USA), the Elisée 350 (Resmed, Australia) and the Medumat transport (Weinmann, Germany).

Methods

We used a decompression chamber to mimic the hypobaric environment at a range of simulated cabin altitudes of 2,438 and 3,657 m (8,000 and 12,000 feet). Ventilators were tested with a set fraction of inspired oxygen of 50% and Vt set at 450. Respiratory rate was 12 breaths/minute. Comparisons of preset to actual measured values were accomplished using a t test for each altitude. The protocol included 36 measurements for each Vt set at each simulated altitude. A significant difference was defined by P < 0.05.

Results

Figure 1 summarizes the data. Comparisons of actual delivered Vt in altitude and set Vt demonstrated a significant difference for the three ventilators.

figure1

Figure 1

Conclusion

The LTV-1200 showed a very significant increase in Vt delivered with increasing altitude (suggesting a lack of efficacy of altimetric correction in hypobaric conditions), whereas the Elisée 350 and Medumat transport delivered respectively a stable and a rather stable Vt.

Author information

Affiliations

Authors

Corresponding author

Correspondence to E Forsans.

Rights and permissions

Reprints and Permissions

About this article

Cite this article

Forsans, E., Franck, L., Leclerc, T. et al. Performances of ventilator at simulated altitude. Crit Care 16, P476 (2012). https://doi.org/10.1186/cc11083

Download citation

Keywords

  • Oxygen
  • Public Health
  • Respiratory Rate
  • Emergency Medicine
  • Tidal Volume