Performances of ventilator at simulated altitude

Introduction

We have assessed the ability of three ventilators to deliver to a normal lung model a set tidal volume (Vt) at different simulated cabin altitudes. We studied the performance of the LTV-1200 (Viasys Healthcare, USA), the Elisée 350 (Resmed, Australia) and the Medumat transport (Weinmann, Germany).

Methods

We used a decompression chamber to mimic the hypobaric environment at a range of simulated cabin altitudes of 2,438 and 3,657 m (8,000 and 12,000 feet). Ventilators were tested with a set fraction of inspired oxygen of 50% and Vt set at 450. Respiratory rate was 12 breaths/minute. Comparisons of preset to actual measured values were accomplished using a t test for each altitude. The protocol included 36 measurements for each Vt set at each simulated altitude. A significant difference was defined by P < 0.05.

Results

Figure 1 summarizes the data. Comparisons of actual delivered Vt in altitude and set Vt demonstrated a significant difference for the three ventilators.

figure 1

Figure 1

Conclusion

The LTV-1200 showed a very significant increase in Vt delivered with increasing altitude (suggesting a lack of efficacy of altimetric correction in hypobaric conditions), whereas the Elisée 350 and Medumat transport delivered respectively a stable and a rather stable Vt.

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Forsans, E., Franck, L., Leclerc, T. et al. Performances of ventilator at simulated altitude. Crit Care 16 (Suppl 1), P476 (2012). https://doi.org/10.1186/cc11083

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc11083

Keywords