Super high-flux continuous hemodialysis: an efficient compromise for blood purification in sepsis

Introduction

High cut-off membranes are proposed for blood purification therapy in septic shock. However, albumin loss related to these membranes is a major drawback limiting their clinical acceptance. Super High-Flux membranes with an optimized cut-off may combine enhanced middle molecule clearances (inflammatory mediators) with limited albumin loss. The aim of our study was to compare small, middle molecule clearances and albumin loss between continuous hemodialysis using a Super High-Flux membrane (SHF-HD) and conventional continuous hemofiltration (CVVH).

Methods

After approval by the ethics committee, patients were enrolled in a single-blind RCT. Patients with septic shock and acute kidney injury received either SHF-HD (EMiC2® filter; Fresenius Medical Care) (cut-off = 40 kDa, dialysate flow rate of 40 ml/kg/hour) or conventional CVVH (cut-off = 30 kDa, UF flow rate of 40 ml/kg/hour). Each patient received a maximum of three sessions of 48 hours each. Creatinine (113 Da), β2-microglobulin (β2-M) (11.8 kDa), kappa free light chain of immunoglobulins (κ-FLC) (23 kDa) and albumin (68 kDa) clearances were measured at 15 minutes, 1 hour, 4 hours, 12 hours, 24 hours and 48 hours. β2-M and κ-FLC were chosen as a middle molecular weight marker. A linear mixed-effects model compared clearances between groups.

Results

Twenty-four patients were included, 12 in the SHF-HD group (32 sessions) and 12 in CVVH (30 sessions). κ-FLC and albumin clearances were higher in the SHF-HD group over time. No difference was observed for creatinine (P = 0.18) and β2-M (P = 0.98) clearances. Plasma albumin levels and the amount of albumin infused did not differ between groups. See Figure 1.

figure 1

Figure 1

Conclusion

The removal of middle molecular weight molecules is higher with SHF-HD. Albumin loss was limited in both groups, even with SHF-HD. Therefore, SHF membranes seem to represent an alternative to high cut-off membranes for blood purification therapies.

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Rimmelé, T., Page, M., Ber, C. et al. Super high-flux continuous hemodialysis: an efficient compromise for blood purification in sepsis. Crit Care 16 (Suppl 1), P377 (2012). https://doi.org/10.1186/cc10984

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc10984

Keywords