Volume 5 Supplement 1

21st International Symposium on Intensive Care and Emergency Medicine

Open Access

Comparison of the P/V curve obtained by the supersyringe and the optoelectronic plethysmography

  • D Chiumello1,
  • E Calvi1,
  • E Noe'1,
  • L Civardi1,
  • E Carlesso1,
  • A Aliverti2 and
  • R Dellacà2
Critical Care20015(Suppl 1):P025

https://doi.org/10.1186/cc1093

Received: 15 January 2001

Published: 2 March 2001

The pressure volume (P-V) curve of the total respiratory system is drawn assuming the changes of chest wall (?Vcw) equal to the volume displaced from the syringe (?Vgas). We compared ?Vgas and ?Vcw during P-V curves obtained by supersyringe and opto-electronic plethysmography [1]. Eight sedated paralysed intubated ALI/ARDS patients (5 M/3 F, age 70 ± 13 years, BMI 25.6 ± 3 kg/m2, PaO2/FiO2 220 ± 76) were studied. During the manoeuvre ?Vcw was recorded by optoelectronic plethysmography. The volume injected and withdrawn by the supersyringe step by step (100 ml) was corrected by temperature, humidity, pressure and gas exchange [2]. The discrepancy was computed as the difference between the volume of air inflated and the chest volume measured. The compliance of the total respiratory system was measured between zero and maximum airway pressure values on the inflation (Crsinf) and deflation (Crsdef) limbs of the P-V curves. Hysteresis loops of the corrected P-Vgas and P-Vcw curves were calculated as the percentage ratio between the area of the P-V curves and the product of maximal volume by maximal pressure.

Even considering thermodynamics and gas exchange correction, ?Vgas values were systematically higher than ?Vcw probably due to blood shifts from the thorax to the extremities. As a consequence, the standard supersyringe method provides an overestimation of the inspiratory and expiratory compliance of the total respiratory system on the inflation limb and an overestimation of the hysteresis area.

Table 1

 

Discrepancy

Crsinf

Crs def

Hysteresis

 

(ml)

(ml/cmH2O)

(ml/cmH2O)

(%)

Pvgas

130 ± 83

66 ± 21*

63 ± 21*

19.2 ± 5.9*

PVcw

-

59 ± 19

58 ± 18

15.2 ± 5.4

The data are expressed as mean ± SD.* Paired t-test P < 0.05 vs PVCW.

Authors’ Affiliations

(1)
Istituto di Anestesia e Rianimazione, Ospedale Policlinico, IRCCS, Università di Milano
(2)
Dipartimento di Bioingegneria, Politecnico di Milano, Centro di Bioingegneria, Fondazione Don Gnocchi IRCCS

References

  1. Aliverti A, Dellacà R, Pelosi P, et al.: Am J Respir Crit Care Med 2000, 161: 1546-1552.PubMedView ArticleGoogle Scholar
  2. Gattinoni L, Mascheroni D, Basilico E, et al.: Intensive Care Med 1987, 13: 19-25. 10.1007/BF00263552PubMedView ArticleGoogle Scholar

Copyright

© The Author(s) 2001