Impact of arterial catheter location on the accuracy of cardiac output provided by an endotracheal bioimpedance device

Introduction

With respect to a radial arterial pressure measurement, a more central achievement of this pressure should improve the reliability of endotracheal bioimpedance cardiac output (CO) monitoring. We therefore compared prospectively the impact on accuracy of this device, in comparison with thermodilution (TD) CO.

Methods

Fourteen patients undergoing cardiac surgery with cardiac output monitoring by TD have been enrolled. A specially designed endotracheal tube (ECOM; ConMed) was placed in conjunction with a catheter located either in the brachial (18 G) or in the radial (20 G) artery in each group of seven patients. Six individual measurements have been carried out in each patient at fixed period, resulting in a total of 42 measurements for each subset. The mean CO by TD was compared with CO by ECOM for each operative period and assessed for agreement by linear regression, Bland-Altmann analysis and percentage error methods. The measurement error should not exceed 30% to be considered as valid, according to Critchley and colleagues.

Results

Mean patient age was 71 years (56 to 89) (13 male, one female). R2 values of 0.47 (P < 0.01) and 0.63 (P < 0.01) in the linear regressions and errors of 41% and 50% were found for the radial and brachial catheter data, respectively. See Figures 1 and 2.

Figure 1
figure 1

abstract P224

Figure 2
figure 2

abstract P229

Conclusion

Accuracy was considerably improved using a brachial artery catheter. Nevertheless, measurement errors between TD and ECOM using either a radial or brachial catheter both exceed 30%. Based on these results and under the current technical conditions, ECOM should not replace TD in CO monitoring for patients undergoing cardiac surgery.

References

  1. Critchley , et al.: J Clin Monit. 1999, 15: 85-91. 10.1023/A:1009982611386

    Article  CAS  Google Scholar 

  2. Bland , et al.: Lancet. 1986, i: 307-310.

    Article  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Gennart, F., Beckers, S., Verborgh, C. et al. Impact of arterial catheter location on the accuracy of cardiac output provided by an endotracheal bioimpedance device. Crit Care 16 (Suppl 1), P221 (2012). https://doi.org/10.1186/cc10828

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc10828

Keywords