Cardiac output monitoring in cirrhotic patients: EV1000 versus pulmonary artery catheter - preliminary data

Introduction

The EV1000 platform, a new calibrated device for intermittent and continuous cardiac output monitoring, has recently been introduced into clinical practice [1]. This study aims to assess the level of agreement between intermittent and continuous cardiac output obtained from VolumeView (ICOvv and CCOvv) connected to the EV1000 platform (Edwards Lifesciences, Irvine, CA, USA) and intermittent (ICOvig) and continuous cardiac output (CCOvig) obtained using an advanced pulmonary artery catheter (PAC) connected to the Vigilance System (Edwards Lifesciences) in cirrhotic patients undergoing liver transplantation.

Methods

Seven consecutive patients (seven male, zero female), mean age 56.5 (± 12) years, were enrolled into the study. ICO data were obtained from the two devices after ICU admission (T0) and after 12 (T12) and 24 hours (T24). CCOvig and CCOvv were recorded every hour from T0 up to 48 hours after ICU admission. Agreement and precision between CO values were evaluated with Bland-Altman analysis. The percentage error (PE) was calculated as 2SD/mean CO [2].

Results

Twenty-one ICO data pairs were compared. Two patients were excluded from CCO data analysis for technical reasons. A total 240 CCO data pairs from five patients were analysed. Data yielded were analysed as total and for CO values lower and higher than 8 l/minute (Table 1).

Table 1 Bias, 2SD and PE for all data pairs and for CO higher and lower than 8 l/minute

Conclusion

These data, even if very preliminary, showed low agreement and high PE either for intermittent and continuous CO obtained from the VolumeView. However, for CO data lower than 8 l/minute the PE was improved.

References

  1. Bendjelid , et al.: Crit Care. 2010, 14: R209. 10.1186/cc9332

    Article  PubMed Central  PubMed  Google Scholar 

  2. Cecconi , et al.: Crit Care. 2009, 13: 201. 10.1186/cc7129

    Article  PubMed Central  PubMed  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Costa, G., Cecconet, T., Baron, D. et al. Cardiac output monitoring in cirrhotic patients: EV1000 versus pulmonary artery catheter - preliminary data. Crit Care 16 (Suppl 1), P219 (2012). https://doi.org/10.1186/cc10826

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc10826

Keywords