Evaluation of a continuous blood glucose monitoring system using a central venous catheter with an integrated microdialysis function

Introduction

Glycemic control in critically ill patients has been debated over the last decade. An accurate glucose monitoring system is essential to understand and study this concern. We have evaluated the accuracy and technical feasibility of a continuous glucose monitoring system using intravascular microdialysis.

Methods

Thirty patients undergoing cardiac surgery were monitored using a triple-lumen central venous catheter (Eirus TLC®; Dipylon Medical AB, Sweden) with an integrated microdialysis membrane. The catheter was placed with the tip in the superior vena cava, and functions both as a central venous catheter, enabling blood sampling and administration of medication, while simultaneously measuring glucose. The patients were monitored for up to 48 hours postoperatively in the ICU. As reference, arterial blood samples were taken every hour.

Results

Data were available from all 30 patients. A total of 725 paired (arterial blood gas-microdialysis) samples were obtained. Glucose correlation coefficient was 0.87. Using Clarke error grid analysis, 100% of the paired samples were in region AB and 97.4% in region A (Figure 1). Mean glucose level was 8.6 mmol/l, bias -1.3% and mean absolute relative difference was 4.8%. A total 97.5% of the paired samples were correct according to ISO criteria. Bland-Altman analysis showed bias ± limits of agreement -0.11 ± 1.3 mmol/l (Figure 2).

Figure 1
figure 1

Clarke error grid analysis.

Figure 2
figure 2

Bland-Altman analysis.

Conclusion

Central venous microdialysis is an accurate and reliable method for continuous blood glucose monitoring in critically patients.

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Möller, F., Liska, J., Öwall, A. et al. Evaluation of a continuous blood glucose monitoring system using a central venous catheter with an integrated microdialysis function. Crit Care 16 (Suppl 1), P170 (2012). https://doi.org/10.1186/cc10777

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc10777

Keywords