Digitalized acoustic monitoring of lung congestion

Introduction

Changes in lung water are known to change breath sound acoustics [1]. Using two pig models, we observed that continuous elevation of lung sound amplitude may indicate an increase in total lung water content [2]. Here we report three cases of ventilated patients in whom continuous acoustic monitoring was done during extravascular lung water (EVLW) measurements.

Methods

We retrospectively analyzed cases in which EVLWi (PiCCO) and other clinical parameters were measured, during continuous acoustic monitoring (VRI), using eight small sensors adhered to the anterior chest. A transmission factor (TF) was calculated, using the sound transfer function between different sensors. The TF changes in correspondence to changes in tissue density [1]. The difference in TF was calculated between recordings when pulmonary edema was observed (>7 ml/kg threshold accompanied with an increase of 2 ml/kg in the EVLWi) and when absent. Statistical analysis was made using a t test.

Results

A total of 336 continuous acoustic recordings in three patients (acoustic monitoring was applied together with EVLWi measurements) were analyzed (146 recordings when lung edema was present; 190 with no edema). In all patients, the acoustic profile corresponded to changes in the clinical picture. In two of the cases, changes in acoustic profile were similar to the ones in the EVLWi and other clinical parameters (Figure 1). In one case, where there was stability in lung sound acoustics, EVLWi and other clinical parameters were also stable. Significant differences existed between recordings with edema (-3.61 ± 0.39) and without edema (-5.71 ± 0.15) (P < 0.001).

Figure 1
figure 1

VRI (average ± SE) versus PiCCO and CXR.

Conclusion

Changes in lung water tend to result in changes in the sound TF, due to changes in the tissue's density. These preliminary results indicate that monitoring lung sounds has the potential to monitor changes in lung water.

References

  1. Donnerberg : Br J Dis Chest. 1980, 74: 23.

    Article  CAS  PubMed  Google Scholar 

  2. Lev : Crit Care. 2011, 15: P174. 10.1186/cc9594

    Article  PubMed Central  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Lev, S., Wolloch, L., Kagan, I. et al. Digitalized acoustic monitoring of lung congestion. Crit Care 16 (Suppl 1), P83 (2012). https://doi.org/10.1186/cc10690

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc10690

Keywords