Introduction
The sequestration of drugs, such as fentanyl [1,2], thiopental, nitroglycerine and propofol [3] in the extracorporeal circuit in vitro has been well described. This phenomenon can change the pharmacokinetic behaviour of drugs during cardiopulmonary bypass, thus potentially leading to problems in achieving adequate dosing regimens.
Objective
The aim of this in vitro laboratory study was to compare the binding of propofol to different oxygenator membranes, and to examine the effects of the type of prime solution and temperature on the rate of binding.
Methods
Samples of three types of membrane oxygenators were used: the SM-35 (polydimethylsiloxane in sheet form), the CML (polypropylene in sheet form) and the SAFE II (polypropylene in hollow fibre form). 14C-propofol in either crystalloid solution or diluted bovine blood was incubated with samples of membranes at 28°C or 37°C. Membrane samples were removed at 30, 60, 90 and 120 min and rinsed. The mass of propofol bound to the various membranes was then measured using liquid scintillation counting. This experiment was carried out for both types of prime solution at 28°C and 37°C.
Results
See Table 1. The SM-35 membrane bound significantly more propofol than the membranes from the CML and the SAFE II. Binding of propofol in diluted blood was significantly less than in crystalloid solution. Temperature had little effect on propofol binding in either prime solution type.
References
Koren G, Crean P, Klein J, Goresky G, Villamater J, MacLeod SM: Sequestration of fentanyl by the cardiopulmonary bypass (CPBP). Eur J Clin Pharmacol 1984, 27: 51. 10.1007/BF00553154
Skacel M, Knott C, Reynolds F, Aps C: Extracorporeal circuit sequestration of fentanyl and alfentanil. Br J Anaesth 1986, 58: 947-949.
Tarr TJ, Kent AP: Sequestration of propofol in an extracorporeal circuit. J Cardiothorac Anesth 1989, 3(suppl 1): 75. 10.1016/0888-6296(89)90818-1
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Hickey, S., Quasim, I. & Gaylor, J. In vitro variability in propofol absorption by different membrane oxygenators. Crit Care 5, 8 (2001). https://doi.org/10.1186/cc1037
Published:
DOI: https://doi.org/10.1186/cc1037
Keywords
- Fentanyl
- Polydimethylsiloxane
- Nitroglycerine
- Thiopental
- Fibre Form