We read with interest the recent article by Monge Garcia and colleagues [1]. We have two comments regarding this interesting physiological study.
First, although we support their observation [2], their explanations may lead to some confusion. Indeed, the Edwards Vigileo™ system-FloTrac™ sensor calculates stroke volume (SV) using the equation:

where σAP is the standard deviation of the arterial pressure (AP) curve and Khi a constant quantifying arterial elastance and vascular resistance [3]. As stroke volume variation (SVV):

Which means that SVV (%) ≈ pulse pressure variation (PPV; %) for FloTrac™.
With the present mathematical equation, it can be demonstrated that the SVV calculated by FloTrac™ (SVVFT) is not influenced by Khi, which means that SVVFT calculation does not include effective arterial elastance (Ea). The originality of Garcia and colleagues' finding is that they realized that when measuring PPV conventionally [4], the ratio PPV/SVVFT becomes a mirror of a dynamic Ea, as the conventional PPV selected integrates Khi (Figure 1). However, when the present innovative method is used, PPV and SVV should be sampled during the same period of time.
In conclusion, we believe that dynamic Ea predicting arterial pressure response to volume loading in preload-dependent patients is an interesting physiological concept. However, this demonstration cannot be achieved without taking into consideration the impact of PPV/SVV sampling.
Abbreviations
- AP:
-
arterial pressure
- Ea:
-
arterial elastance
- PP:
-
pulse pressure
- PPV:
-
pulse pressure variation
- SV:
-
stroke volume
- SVV:
-
stroke volume variation.
References
Monge Garcia MI, Gil Cano A, Gracia Romero M: Dynamic arterial elastance to predict arterial pressure response to volume loading in preload-dependent patients. Crit Care 2011, 15: R15. 10.1186/cc9420
Bendjelid K: When to recalibrate the PiCCO? From a physiological point of view, the answer is simple. Acta Anaesthesiol Scand 2009, 53: 689-690. 10.1111/j.1399-6576.2009.01919.x
Manecke GR: Edwards FloTrac sensor and Vigileo monitor: easy, accurate, reliable cardiac output assessment using the arterial pulse wave. Expert Rev Med Devices 2005, 2: 523-527. 10.1586/17434440.2.5.523
Michard F, Chemla D, Richard C, Wysocki M, Pinsky MR, Lecarpentier Y, Teboul JL: Clinical use of respiratory changes in arterial pulse pressure to monitor the hemodynamic effects of PEEP. Am J Respir Crit Care Med 1999, 159: 935-939.
Author information
Authors and Affiliations
Corresponding author
Additional information
Competing interests
The authors declare that they have no competing interests.
Authors’ original submitted files for images
Below are the links to the authors’ original submitted files for images.
Rights and permissions
About this article
Cite this article
Giraud, R., Siegenthaler, N. & Bendjelid, K. Pulse pressure variation, stroke volume variation and dynamic arterial elastance. Crit Care 15, 414 (2011). https://doi.org/10.1186/cc10088
Published:
DOI: https://doi.org/10.1186/cc10088
Keywords
- Standard Deviation
- Arterial Pressure
- Emergency Medicine
- Stroke Volume
- Vascular Resistance