Abstract

Background

Noninvasive ventilation (NIV) is generally delivered using pneumatically-triggered and cycled-off pressure support (PSP) through a mask. Neurally adjusted ventilatory assist (NAVA) is the only ventilatory mode that uses a non-pneumatic signal, i.e., diaphragm electrical activity (EAdi), to trigger and drive ventilator assistance. A specific setting to generate neurally controlled pressure support (PSN) was recently proposed for delivering NIV by helmet. We compared PSN with PSP and NAVA during NIV using a facial mask, with respect to patient comfort, gas exchange, and patient-ventilator interaction and synchrony.

Methods

Three 30-minute trials of NIV were randomly delivered to 14 patients immediately after extubation to prevent post-extubation respiratory failure: (1) PSP, with an inspiratory support ≥8 cmH2O; (2) NAVA, adjusting the NAVA level to achieve a comparable peak EAdi (EAdipeak) as during PSP; and (3) PSN, setting the NAVA level at 15 cmH2O/μV with an upper airway pressure (Paw) limit to obtain the same overall Paw applied during PSP. We assessed patient comfort, peak inspiratory flow (PIF), time to reach PIF (PIFtime), EAdipeak, arterial blood gases, pressure-time product of the first 300 ms (PTP300-index) and 500 ms (PTP500-index) after initiation of patient effort, inspiratory trigger delay (DelayTR-insp), and rate of asynchrony, determined as asynchrony index (AI%). The categorical variables were compared using the McNemar test, and continuous variables by the Friedman test followed by the Wilcoxon test with Bonferroni correction for multiple comparisons (p < 0.017).

Results

PSN significantly improved patient comfort, compared to both PSP (p = 0.001) and NAVA (p = 0.002), without differences between the two latter (p = 0.08). PIF (p = 0.109), EAdipeak (p = 0.931) and gas exchange were similar between modes. Compared to PSP and NAVA, PSN reduced PIFtime (p < 0.001), and increased PTP300-index (p = 0.004) and PTP500-index (p = 0.001). NAVA and PSN significantly reduced DelayTR-insp, as opposed to PSP (p < 0.001). During both NAVA and PSN, AI% was <10% in all patients, while AI% was ≥10% in 7 patients (50%) with PSP (p = 0.023 compared with both NAVA and PSN).

Conclusions

Compared to both PSP and NAVA, PSN improved comfort and patient-ventilator interaction during NIV by facial mask. PSN also improved synchrony, as opposed to PSP only.

Trial registration

ClinicalTrials.gov, NCT03041402. Registered (retrospectively) on 2 February 2017.

Keywords

Noninvasive ventilation Pressure support ventilation Neurally adjusted ventilatory assist Patient-ventilator interaction Ventilator performance Patient-ventilator asynchrony

Background

Noninvasive ventilation (NIV) is increasingly used for treating acute respiratory failure (ARF) [1, 2] and is commonly applied using a facial mask [3] and pneumatically triggered and cycled-off pressure support (PSP) [4]. Although better tolerated than invasive mechanical ventilation, NIV is characterized by drawbacks such as poor patient-ventilator interaction and discomfort [5], which are major determinants of NIV failure.

In particular, the pneumatic signals, i.e., flow, volume and airway pressure (Paw), are leak-sensitive [6] and frequently cause patient-ventilator asynchrony [7]. The only mode not utilizing pneumatic signals to trigger and drive the ventilator is neurally adjusted ventilator assist (NAVA). In fact, with NAVA the ventilator assistance is under the control of the diaphragm electrical activity (EAdi) [8]. In contrast to PSP, NAVA has been repeatedly shown to improve patient-ventilator interaction and reduce asynchronies, both during invasive ventilation [9, 10] and NIV [4, 1115]. However, NAVA is characterized by a lower rate of pressurization than PSP [4].

Recently, a specific NAVA setting has been proposed to generate EAdi-controlled pressure support (PSN) in patients receiving either invasive ventilation [16] or NIV by helmet [4]. PSN consists of increasing the user-controlled gain factor (NAVA level) at the maximum level, while limiting peak airway pressure (Pawpeak) by adjusting the upper pressure limit [4, 16].

During NIV delivered by helmet, compared to both PSP and NAVA, PSN results in better pressurization and triggering performance, which improves patient comfort while reducing EAdi, without affecting the respiratory rate and gas exchange [4]. Due to the different characteristics of helmets and masks, it is unclear whether these advantages could be extended to NIV delivered by mask. This physiological study aims at comparing PSN with PSP and NAVA, with respect to the patient’s comfort (primary endpoint), breathing pattern, respiratory drive, gas exchange, pressurization and triggering performance and patient-ventilator synchrony (additional endpoints).

Methods

The present physiologic, crossover, randomized study was conducted from March to September 2013 in the Intensive Care Units (ICUs) of the University Hospital “Maggiore della Carità” (Novara, Italy) and the ZhongDa Hospital, Southeast University (Nanjing, China). The study was approved by the local Ethics Committees “A.O.U Maggiore della Carità” in Novara, Italy (protocol n° 64/12) and the Research Ethics Board of Zhongda Hospital, Southeast University, Nanjing, China (2013ZDSYLL097.0). Written informed consent was obtained from the patients for publication of their individual details and accompanying images in this manuscript. The consent forms are held by the authors and are available for review by the Editor-in-Chief. At the time the study was conducted, trial registration was not mandatory for this type of investigation; however, the trial was retrospectively registered at ClinicalTrials.gov (NCT03041402). We followed the Consolidated Standards of Reporting Trials (CONSORT) recommendations for reporting of randomized trials [17].

Patients

We considered any patient eligible who was ≥18 years of age and admitted to the ICU, and who was orally intubated and undergoing invasive mechanical ventilation for at least 48 hours. The inclusion criteria were: (1) consciousness, as indicated by a Glasgow Coma Scale (GCS) of 11 (i.e. spontaneous eye opening, response to command and no verbal response because of the endotracheal tube in place); (2) no infusion of midazolam or propofol in the previous 24 hours or 4 hours, respectively; and (3) readiness for extubation with indication, prior to extubation, to receive NIV to prevent post-extubation respiratory failure. The patients were considered to be eligible for the spontaneous breathing trial if they met the following criteria [18]: (1) GCS ≥8; (2) presence of clearly audible cough during suctioning; (3) tracheal suctioning ≤2/hour; (4) normal sodium blood values; (5) core temperature <38.5 °C during the previous 8 hours; (6) arterial oxygen tension (PaO2) to fraction of inspired oxygen (FiO2) ratio (PaO2/FIO2) ≥200 with positive end-expiratory pressure (PEEP) ≤5 cmH2O; (7) FiO2 ≤ 0.4; (8) heart rate ≤125 beats/min; and (9) systolic blood pressure >90 mmHg without epinephrine or norepinephrine infusion and with dopamine infusion ≤5 mcg/kg/min. The patients considered to be at risk of extubation failure exhibited at least one of the following: (1) more than one consecutive failure of the weaning trial [19]; (2) arterial partial pressure of carbon dioxide (PaCO2) >45 mmHg at the end of the 30-min spontaneous breathing trial [20]; (3) chronic respiratory disorders [19]; and (4) chronic heart failure [19].

The exclusion criteria were as follows: (1) need for analgaesic or sedative drugs; (2) recent cervical spine injury; (3) obstructive sleep apnoea syndrome; (4) pregnancy; (5) contraindications to placement of a nasal-gastric feeding tube; (6) inclusion in other research protocols; and (7) lack of consent.

Study protocol

After the patient’s enrolment in the study, the nasal-gastric feeding tube in place was replaced by the EAdi catheter (Maquet Critical Care, Solna, Sweden) [9]. The correct positioning was ascertained as previously described [9]. The study was performed using a standard Servo-I ventilator (Maquet Critical Care, Solna, Sweden) equipped with NAVA module and NIV software for air leaks. The facial mask was individually selected for each patient based on their anthropometric characteristics to minimize air leaks and optimize patient tolerance; the facial mask was selected from among three different models: FreeMotion RT041 Non Vented Full Face Mask (Fisher and Paykel, Auckland, New Zealand); Ultra Mirage FFM-NV (ResMed, San Diego, CA, USA); and PerforMax Face Mask (Philips Respironics, Murrysville, PA, USA).

Immediately after extubation, we performed a 15-min PSP trial, setting the inspiratory pressure support ≥8 cmH2O to obtain a tidal volume of 6–8 mL · kg-1 of ideal body weight, with the fastest rate of pressurization and I/E cycling at 35% of peak inspiratory flow (PIF). All patients subsequently underwent three 30-min trials in random order: (1) PSP, with the settings obtained in the aforementioned trial; (2) NAVA, adjusting the NAVA level in order to achieve a comparable peak EAdi (EAdipeak) as during the PSP trial, with a safety Paw upper limit of 30 cmH2O [4, 15]; and (3) PSN, setting the NAVA level at its maximum (i.e., 15 cmH2O/μV), and an upper Paw limit to obtain the same overall Paw applied during the PSP trial [4, 16, 21]. During both NAVA and PSN, the trigger sensitivity was set at 0.5 μV while the default cycling-off was 70% EAdipeak, as fixed by the manufacturer [21]. PEEP was set by the attending physicians in a range between 5 cmH2O and 10 cmH2O, and it remained unmodified throughout the entire study period. The FiO2 was regulated to obtain peripheral oxygen saturation (SpO2) between 94% and 96%, before starting the protocol, and it remained unmodified throughout the study period.

The three modes of ventilation were applied according to a computer-generated random sequence using sealed, opaque, numbered envelopes. The envelopes were kept in the head nurse’s office in both institutions. The envelope was opened by the nurse in charge of the patient, and the prescribed sequence of modes was communicated to the investigators.

The predefined criteria for protocol interruption were as follows: (1) need for emergency re-intubation; (2) SpO2 < 90%; (3) acute respiratory acidosis, as defined by PaCO2 > 50 mmHg and pH <7.30; (3) inability to expectorate secretions; (4) hemodynamic instability (i.e., need for continuous infusion of dopamine or dobutamine >5 μg∙kg-1∙min-1, norepinephrine >0.1 μg∙kg-1∙min-1 or epinephrine or vasopressin at any dosage to maintain mean arterial blood pressure >60 mmHg); (5) life-threatening arrhythmias or electrocardiographic signs of ischaemia; or (6) loss of 2 or more points on the GCS.

Data acquisition and analysis

Airflow, Paw and EAdi were acquired from the ventilator using an RS232 interface at a sampling rate of 100 Hz and were recorded on a computer using dedicated software (ServoTracker V. 4.0, Maquet Critical Care, Solna, Sweden). The last minute of each trial was manually analysed off-line using customized software based on Microsoft Excel, as previously described [9].

Comfort was assessed through an 11-point numeric rating scale (NRS), as previously reported [4, 2224]. Before protocol initiation, all patients received a detailed explanation of the NRS. The patients were asked to evaluate their comfort level, indicating a number between 0 (worst possible comfort) and 10 (best possible comfort) using an ICU-adapted large-printed scale including numbers and descriptors [23]. The scores obtained were recorded without additional indications or comments [24].

Breathing pattern was assessed by determining (1) mechanical inspiratory time (TImec), breath duration (TTOTmec) and rate of ventilator cycling (RRmec) from the flow tracing, and (2) the patient’s own (neural) inspiratory time (TIneu), breath duration (TTOTneu) and respiratory rate (RRneu) from the EAdi tracing. The mechanical (TI/TTOTmec) and neural (TI/TTOTneu) inspiratory duty cycles were also calculated [15, 25]. Air leaks were computed over one minute as the difference between inspiratory and expiratory tidal volumes times RRmec and were expressed as percentage of the exhaled volume over one minute [15, 25]. Moreover, we measured Pawpeak, peak inspiratory flow (PIF) and the time to reach PIF from the onset of the patient’s effort (PIFtime). EAdipeak was also determined as an index of respiratory drive [26]. Gas exchange was assessed at the end of each trial by sampling arterial blood from a catheter already inserted for clinical purposes.

To evaluate the pressurization performance, we computed the pressure-time product (PTP) of the first 200 ms from the onset of the ventilator pressurization (PTP200), and the PTP of the first 300 ms and 500 ms from the onset of the neural effort, expressed as the percentage of the area of ideal pressurization (PTP300-index and PTP500-index, respectively) [4, 24, 27, 28]. The ideal PTP was computed considering a perfectly squared rectangle on the Paw-time tracing, with the height of the actual Paw above PEEP and the width of the time window considered (i.e., 0.3 second and 0.5 second from the onset of the inspiratory effort, assessed from the EAdi tracing, for PTP300-index and PTP500-index, respectively) [4, 24, 27, 28]. The triggering performance was evaluated by determining the pressure drop (ΔPtrigger) and PTP of Paw (PTPt) during the triggering phase [4, 24, 27, 28].

To assess patient-ventilator synchrony, we computed the inspiratory trigger delay (DelayTR-insp), as the time lag between the onsets of neural inspiration and ventilator support, and the expiratory trigger delay (DelayTR-exp), as the time lag between the fall towards baseline of EAdi and the end of ventilator support. The time during which respiratory effort and ventilator assistance were synchronous, indexed to the TIneu (Timesynch/TIneu), was also computed [4, 24, 27]. The asynchrony index (AI%) was calculated as the total number of asynchronies (i.e., ineffective efforts, auto-triggers and double-triggers) divided by the sum of triggered and non-triggered breaths [7]. An AI% ≥10% was considered to indicate a clinically relevant rate of asynchronies [7].

Statistical analysis

To detect an increase in comfort of 2.5 [4], with α risk of 0.05 and β risk of 0.20, a sample of 12 patients was deemed necessary. Because this calculation was based on a pairwise comparison and we actually compared three conditions, we applied the Bonferroni correction, which reduced the α risk from 0.05 to 0.017, increasing the sample size up to 14 patients. We used non-parametric tests because of the relatively small number of patients. The data are reported as median values (25–75% interquartile), unless otherwise specified. All continuous variables were compared between modes using the Friedman test and then by the Wilcoxon test; the Bonferroni correction was applied for multiple comparisons (p < 0.017). We compared the categorical data using the McNemar test. The Spearman rank correlation test was used to ascertain the correlation between each individual comfort score and the corresponding PTP300-index, PTP500-index, PTPt, DelayTR-insp, PIF and PIFtime. For these comparisons, we considered two-sided p values <0.05 significant. All statistical analyses were performed using the Sigmaplot v. 12.0 (Systat Software Inc., San Jose, CA, USA). No interim analysis has been planned or conducted.

Results

We enrolled 14 consecutive patients. The patients’ study flow is shown in Fig. 1. All patients completed the study protocol without any complication and were included in the data analysis. No patient required either sedative or analgaesic drugs during the study period. No patients met any criteria for post-extubation respiratory failure requiring re-intubation. The patients’ demographic and anthropometric characteristics are shown in Table 1.
Fig. 1

Enrolment of the study participants. The flow of patient assessment and inclusion in the protocol is shown. A total of 54 patients were considered eligible for the study, having met all inclusion criteria: 40 patients were excluded from the study because they met one or more of the exclusion criteria. Therefore, 14 patients were included in the study. No protocol discontinuations were recorded. EAdi diaphragm electrical activity

Table 1

Patient characteristics at enrolment

Patient

Weight: kg

BMI: kg/m2

Admission pathology

SAPSII

PEEP: cmH2O

PS: cmH2O

FiO2

1

90

27.8

SE-COPD

38

10

14

0.40

2

92

29.1

SE-COPD

34

10

14

0.50

3

70

23.7

Pneumonia

28

10

10

0.40

4

87

28.2

Sepsis

37

5

15

0.35

5

75

24.5

Polytrauma

44

5

12

0.30

6

80

26.1

Polytrauma

29

5

15

0.35

7

64

23.5

Pneumonia

38

5

12

0.40

8

70

25.7

Pneumonia

38

5

8

0.50

9

60

22.0

Pneumonia

27

5

10

0.40

10

67

24.1

SE-COPD

39

7

12

0.35

11

50

19.5

Pneumonia

56

5

10

0.40

12

60

20.8

Pneumonia

57

7

12

0.40

13

58

19.6

CPE

47

8

10

0.40

14

70

25.1

Sepsis

40

5

10

0.40

BMI body mass index, SAPSII Simplified Acute Physiology Score II, PEEP Positive end-expiratory pressure, PS pressure support, FiO 2 inspired fraction of oxygen, SE-COPD severe exacerbation of chronic obstructive pulmonary disease, CPE cardiac pulmonary edema

Comfort

The individual values of the comfort score for all the patients and their median and interquartile range are depicted in Fig. 2. PSN significantly improved patient comfort (7 (7; 8)), compared to both PSP (5 (5; 6); p = 0.001) and NAVA (5 (5; 7)); p = 0.002), with no differences between PSP and NAVA (p = 0.08). Comfort was directly correlated to PTP300-index (ρ =0.51, p < 0.001) and to PTP500-index (ρ =0.46, p=0.002); comfort was also inversely correlated to DelayTR-insp (ρ =-0.58, p < 0.001), PIFtime (ρ =-0.47, p=0.002) and PTPt (ρ =-0.55, p <0.001) while not correlated to PIF (ρ =−0.14, p=0.369).
Fig. 2

Comfort score. Individual values (open circles), median and interquartile range (solid lines) of the comfort score during pneumatically triggered pressure support (PS P ), neurally adjusted ventilatory assist (NAVA) and neurally controlled pressure support (PS N ) are depicted from left to right

Breathing pattern, respiratory drive and gas exchange

As reported in Table 2, the breathing pattern was not different between modes. Only TI/TTOTmec was significantly lower during PSP, as opposed to both NAVA (p = 0.007) and PSN (p = 0.010). Pawpeak (p = 0.607), air leaks (p = 0.395) and respiratory drive, as indicated by the EAdipeak (p = 0.931), were also not different between modes. PIF did not differ between the three modes of ventilation (p = 0.109), while PIFtime was significantly reduced by PSN, as opposed to both PSP and NAVA (p < 0.001 for both comparison), with no differences between PSP and NAVA (p = 0.217). Figure 3 shows, from top to bottom, Paw, flow and EAdi tracings of one representative patient undergoing PSP (left), NAVA (middle) and PSN (right). The arrow indicates an ineffective inspiratory effort during PSP. The median group values are presented in Table 2.
Table 2

Breathing pattern, respiratory drive, gas exchange, pressurization and triggering performance and patient-ventilator synchrony

 

Friedman test (p value)

PSP

NAVA

PSN

Breathing pattern and respiratory drive

 RRmec (breaths/min)

0.606

23.9 (18.7; 30.6)

26.7 (19.5; 30.6)

27.4 (18.4; 31.7)

 RRneu (breaths/min)

0.931

25.7 (18.6; 32.9)

26.2 (19.6; 30.7)

26.4 (19.3; 30.8)

 TImec (sec)

0.168

0.71 (0.58; 0.87)

0.83 (0.61; 1.11)

0.82 (0.66; 1.04)

 TIneu (sec)

0.606

0.75 (0.56; 1.10)

0.74 (0.59; 1.10)

0.75 (0.59; 0.96)

 TI/TTOTmec

0.030

0.30 (0.27; 0.33)

0.33 (0.31; 0.40)*

0.34 (0.29; 0.41)#

 TI/TTOTneu

0.606

0.32 (0.26; 0.37)

0.32 (0.28; 0.38)

0.30 (0.26; 0.34)

 Pawpeak

0.607

19.3 (15.1; 21.1)

18.8 (15.4; 21.0)

19.0 (15.2; 20.5)

 Leaks %

0.395

21.4 (8.9; 43.2)

35.9 (15.2; 47.6)

23.2 (11.5; 61.9)

 PIF (l/sec)

0.109

1.12 (0.85; 1.42)

1.05 (0.71; 1.22)

1.20 (0.77; 1.38)

 PIFtime (sec)

<0.001

0.41 (0.34–0.48)

0.41 (0.33–0.58)

0.22 (0.19–0.26)

 EAdipeak (μV)

0.257

13.7 (7.7; 21.2)

15.3 (8.4; 25.7)

12.6 (6.9; 19.3)

Gas exchange

 pH

0.4576

7.43 (7.40; 7.45)

7.43 (7.40; 7.45)

7.43 (7.40; 7.45)

 PaCO2

0.5134

44.1 (36.2; 50.3)

44.4 (36.1; 51.5)

43.8 (38.2; 50.8)

 PaO2/FiO2

0.5103

213.6 (197.9; 224.0)

214.6 (188.1; 238.0)

214.4 (199.0; 226.2)

Pressurization and triggering performance

 PTP300-index (%)

0.004

24.7 (4.3; 32.7)

25.3 (19.9; 34.0)

42.0 (32.5; 46.5)

 PTP500-index (%)

0.001

44.2 (23.3; 52.1)

46.4 (33.4; 56.6)

62.6 (54.1; 67.9)

 PTP200 (cmH2O/sec)

0.001

86.7 (77.5; 112.5)

62.1 (45.7; 81.9)*

85.0 (69.6; 127.4)§

 PTPt (cmH2O/sec)

<0.001

9.45 (5.89; 12.31)

0.89 (0.23; 3.23)*

0.59 (0.16; 2.33)#

 ΔPtrigger (cmH2O)

<0.001

−1.16 (−1.40; −0.87)

−0.36 (−0.78; −0.11)*

−0.32 (−0.71; −0.11)#

Patient ventilator synchrony

 DelayTR-insp (sec)

<0.001

0.13 (0.08; 0.27)

0.07 (0.03; 0.06)*

0.05 (0.04; 0.06)#

 DelayTR-exp (sec)

0.395

0.13 (0.05; 0.22)

0.10 (0.09; 0.14)

0.11 (0.10; 0.12)

 Timesynch/TIneu

0.010

0.79 (0.70; 0.88)

0.90 (0.86; 0.94)*

0.94 (0.89; 0.98)#

 AI% (%)

<0.001

6.6 (0.0; 23.4)

0.0 (0.0; 0.0)*

0.0 (0.0; 0.0)#

PS P pneumatically triggered and cycled-off pressure support, NAVA neurally adjusted ventilatory assist, PS N neurally controlled pressure support, RR mec ventilator respiratory rate, RR neu patient’s respiratory rate, TI mec inspiratory time of the ventilator, TI neu inspiratory time of the patient, TI/TOT mec ventilator inspiratory duty cycle, TI/TOT neu patient’s inspiratory duty cycle, Paw peak peak airway pressure, PIF peak inspiratory flow, PIF time time to reach the PIF, EAdi electrical activity of the diaphragm, EAdi peak peak value of EAdi, PaCO 2 arterial partial pressure of carbon dioxide, PaO 2 /FiO 2 ratio between arterial partial pressure and inspired fraction of oxygen, PTP pressure time product, PTP 300-index PTP of the first 300 ms since the effort of the patient indexed to the ideal PTP, PTP 500-index PTP of the first 500 ms since the effort of the patient indexed to the ideal PTP, PTP 200 PTP of the first 200 ms since the beginning of pressurization, PTPt PTP of the trigger, ΔP trigger drop of pressure during triggering phase, Delay TR-insp inspiratory trigger delay, Dealy TR-exp expiratory trigger delay, Time synch /TI neu synchronous time between respiratory effort and ventilator assistance, indexed to the TIneu, AI% asynchrony index. *p < 0.017 PSP vs. NAVA, # p < 0.017 PSP vs. PSN, § p < 0.017 NAVA vs. PSN

Fig. 3

Examples of tracings from one representative patient. From top to bottom, airway pressure (Paw), flow and electrical activity of the diaphragm (EAdi) tracings of a representative patient are shown during pneumatically triggered pressure support (PS P ), neurally adjusted ventilatory assist (NAVA) and neurally controlled pressure support (PS N ). The arrow indicates an ineffective effort during PSP

Gas exchanges were no different between trials (Table 2).

Pressurization and triggering performance

Figure 4 depicts Paw profiles of individual breaths during PSP (solid line), NAVA (dotted line) and PSN (dashed line) from another patient. The arrow indicates the beginning of the patient’s own (neural) effort. PSP and PSN have similar Paw profiles, characterized by a fast rate of pressurization; however, during PSN the beginning of pressurization is notably anticipated and closer to the onset of the patient’s effort. NAVA is characterized by a slower rate of pressurization. Consistent with these findings, PSN improved both PTP300-index and PTP500-index, as opposed to both PSP and NAVA (Table 2), whereas PTP200 was lower during NAVA, as compared to both PSP and PSN (p < 0.001 for both comparisons), with no significant difference between PSP and PSN (p = 0.761). Shown also in Table 2, NAVA and PSN significantly reduced DelayTR-insp, PTPt and ΔPtrigger, in contrast to PSP (p < 0.001 for all comparisons). DelayTR-exp was no different between modes (p = 0.395).
Fig. 4

Pressure airway profiles. Airway pressure (Paw) profile of single breaths during pneumatically triggered pressure support (solid line), neurally adjusted ventilatory assist (dotted line) and neurally controlled pressure support (dashed line) from another patient. The arrow indicates the beginning of the patient’s effort. See main text for additional explanation

Patient-ventilator synchrony

Compared to PSP, both NAVA (p = 0.005) and PSN (p = 0.002) improved Timesynch/TIneu, with no differences between the two (p = 0.08) (Table 2). The median values of AI% are reported in Table 2. As expected, during both NAVA and PSN, the AI% was <10% in all patients, whereas it was ≥10% in 7 patients (50%) with PSP (p = 0.023, compared to both NAVA and PSN).

Discussion

This physiologic study shows that in patients receiving NIV by facial mask, compared to both PSP and NAVA, PSN improves pressurization and triggering performance, resulting in better comfort, while not affecting respiratory drive, Arterial Blood Gases ABGs and respiratory rate. Both PSN and NAVA equally improve patient-ventilator synchrony, in contrast to PSP.

To the best of our knowledge, this investigation is the first to evaluate PSN for delivery of NIV using a mask. In a study evaluating intubated patients with COPD and intrinsic PEEP, compared to PSP, PSN improved patient-ventilator interaction and synchrony, and counterbalanced the extra load due to intrinsic PEEP without the need for externally applied PEEP [16]. In healthy volunteers, comfort was reduced when increasing the level of support [29], whereas it was improved by EAdi triggering, as opposed to pneumatic triggering, during NIV delivered by helmet [30]. In a recent study comparing PSN with PSP and NAVA during NIV delivered by helmet in an analogous patient population, PSN improved comfort, pressurization and triggering performance, and reduced EAdi, without affecting gas exchange [4].

Consistent with the results of these investigations, in the present study PSN outperforms PSP with respect to PTP300-index and PTP500-index, PTPt [4, 16, 30], DelayTR-insp, Timesynch/TIneu and AI [4, 16, 30], and comfort [4, 30]. In accordance with Cammarota et al. [4], who compared the same three modes delivering NIV by helmet, PSN improved pressurization PTP300-index and PTP500-index, and comfort with respect to both PSP and NAVA, while in contrast to that study, PSN neither increased PTP200, compared to PSP, nor reduced EAdi, compared to both PSP and NAVA. These discrepancies are likely due to the different physical properties of mask and helmet, the latter being characterized by more problematic triggering and pressurization performance [31]. Nonetheless, we found improvements in triggering and pressurization performance to ameliorate comfort, which is a major determinant of NIV outcome. Indeed, NIV can be complicated by discomfort, which is associated with increased rate of failure and worsened patient outcome [32].

PIF was not different between modes, while PIFtime was shortened by PSN, as opposed to both PSP and NAVA. In intubated patients with acute on chronic respiratory failure undergoing PSP, Bonmarchand et al. evaluated the effects of varying Paw rates of pressurization; they found that the fastest rate generated the highest PIF and was associated with greater reduction in the work of breathing [33]. Similar results were obtained during invasive PSN in restrictive patients [34] and in patients recovering from hypoxemic ARF [35].

To explain the differences between these studies and our investigation, it is important to note the different computational approach to the pressurization indexes [27]. PTP200 reflects the sole rate of pressurization rate, i.e., the slope of Paw after triggering, which affects the PIF. Both PTP300-index and PTP500-index instead consider not only the pressurization rate but also the triggering performance, which influences PIFtime, without affecting PIF. We found PTP200 no different between PSN and PSP, while triggering performance was significantly improved by PSN, as indicated by the lower values of PTPt and DelayTR-insp. Notably, while patient comfort is improved when flow delivery by the ventilator meets the patient’s demand [36], excessively high PIF may worsen the patient’s comfort during both invasive ventilation [37] and NIV [36].

Our study has two limitations. First, the patient sample is small, a limitation that we share with the majority of earlier physiological investigations [4, 9, 1113, 15, 24, 37, 38]. Second, consistent with the results of previous research [4, 2224], we applied the 11-point NRS to assess comfort, although this scale has been formally validated for pain [39, 40] and dyspnoea [41] only.

Conclusions

Compared to both PSP and NAVA, in patients receiving NIV by facial mask, PSN improves triggering performance and patient-ventilator synchrony, thereby ameliorating the patient’s comfort. It remains to be determined whether these physiologic benefits may also occur in other categories of patients and translate into improved clinical outcomes.

Abbreviations

AI%: 

Asynchrony index

ARF: 

Acute respiratory failure

COPD: 

Chronic obstructive pulmonary disease

DelayTR-exp

Expiratory trigger delay

DelayTR-insp

Inspiratory trigger delay

EAdi: 

Diaphragm electrical activity

EAdipeak

Peak of electrical activity of the diaphragm

FiO2

Inspiratory oxygen fraction

GCS: 

Glasgow Coma Scale

ICUs: 

Intensive Care Units

NAVA: 

Neurally adjusted ventilatory assist

NIV: 

Noninvasive ventilation

NRS: 

Numeric rating scale

PaCO2

Arterial partial pressure of carbon dioxide

Paw: 

Airway pressure

Pawpeak

Peak of airway pressure

PEEP: 

Positive end-expiratory pressure

PIF: 

Peak inspiratory flow

PIFtime

Time to reach the peak inspiratory flow from the onset of patient’s effort

PSN

Neurally controlled pressure support

PSP

Pneumatically triggered and cycled-off pressure support

PTP: 

Pressure-time product

PTP200

Pressure-time product of the first 200 ms from the onset of the ventilator pressurization

PTP300-index

Pressure-time product of the first 300 ms from the onset of the neural effort, indexed to the ideal area

PTP500-index

Pressure-time product of the first 500 ms from the onset of the neural effort, indexed to the ideal area

PTPt

Pressure-time product of the triggering phase

RRmec

Rate of ventilator cycling

RRneu

Patient’s own (neural) respiratory rate

SpO2

Peripheral oxygen saturation

TI/TTOTmec

Mechanical inspiratory duty cycle

TI/TTOTneu

Patient’s own (neural) inspiratory duty cycle

TImec

Mechanical inspiratory time

Timesynch/TIneu

Time during which respiratory effort and ventilator assistance are synchronous, indexed to the patient’s own (neural) inspiratory time

TIneu

Patient’s own (neural) inspiratory time

TTOTmec

Total mechanical respiratory time

TTOTneu

Total patient’s own (neural) respiratory time

ΔPtrigger

Pressure drop of the triggering phase

Declarations

Acknowledgements

None.

Funding

Maquet Critical Care (Solna, Sweden) provided the NAVA module and catheters used for the study. A portion of the results from this study was presented in abstract form at the International Symposium on Intensive Care and Emergency Medicine in Brussels (2014).

Availability of data and materials

The full protocol and raw data are available at longhini.federico@gmail.com.

Authors’ contributions

FL was responsible for conception and design of the study, acquisition, analysis and interpretation of the data and for drafting and revising the article for final approval of the version to be published. CP was responsible for the conception and design of the study, acquisition, analysis and interpretation of data and for drafting and revising the article for final approval of the version to be published. JX and GC were responsible for the acquisition of data and for revising the article for final approval of the version to be published. AB and EG were responsible for analysis and interpretation of the data and for drafting and revising the article for final approval of the version to be published. YY participated in the design of the stud, acquisition and analysis of the data and in revising the article for final approval of the version to be published. PN and HQ were responsible for the conception and design of the study, analysis and interpretation of data and for drafting and revising the article for important intellectual content and final approval of the version to be published. All authors have read and approved the final manuscript.

Ethics approval and consent to participate

The study was approved by the local Ethics Committees “A.O.U Maggiore della Carità” in Novara, Italy (protocol n° 64/12) and the Research Ethics Board of Zhongda Hospital, Southeast University, Nanjing, China (2013ZDSYLL097.0). Written informed consent was obtained from each participant before inclusion in the study, according to the local regulations and principles outlined in the Helsinki declaration. At the time the study was conducted, trial registration was not mandatory for this type of investigation.

Competing interests

PN contributed to the development of the helmet, Next (Castar Next, Intersurgical, Mirandola, Italy), whose license for the patent belongs to Intersurgical S.P.A., and received royalties for that invention. PN’s research laboratory has received equipment and/or grants from Maquet Critical Care (Solna, Sweden), Intersurgical S.p.A. (Mirandola, Italy), Draeger Medical GmbH (Corsico, Italy), Biotest (Trezzano sul Naviglio, Italy) and Hillrom (Bussigny, Switzerland). PN received honoraria/speaking fees from Maquet Critical Care (Solna, Sweden), Covidien AG (Segrate, Italy), Draeger Medical GmbH (Corsico, Italy), Breas (Mölnlycke, Sweden), Hillrom (Chicago, IL, USA), Resmed (Vimercate MB, Italy) and Linde AG (Munich, Germany). All other authors declare that they have no competing interests.

Consent for publication

All patients gave consent for data publication according to national regulations.

Publisher’s Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Open AccessThis article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Authors’ Affiliations

(1)
Anesthesia and Intensive Care, Sant’Andrea Hospital, ASL VC
(2)
Department of Critical Care Medicine, Nanjing Zhong-Da Hospital, Southeast University School of Medicine
(3)
Anesthesia and Intensive Care, “Maggiore della Carità” Hospital
(4)
Intensive Care Unit, University Hospital Mater Domini, Department of Medical and Surgical Sciences, Magna Graecia University

References

  1. Esteban A, Frutos-Vivar F, Muriel A, Ferguson ND, Penuelas O, Abraira V, Raymondos K, Rios F, Nin N, Apezteguia C, et al. Evolution of mortality over time in patients receiving mechanical ventilation. Am J Respir Crit Care Med. 2013;188(2):220–30.View ArticlePubMedGoogle Scholar
  2. Nava S, Navalesi P, Carlucci A. Non-invasive ventilation. Minerva Anestesiol. 2009;75(1-2):31–6.PubMedGoogle Scholar
  3. Crimi C, Noto A, Princi P, Nava S. Survey of non-invasive ventilation practices: a snapshot of Italian practice. Minerva Anestesiol. 2011;77(10):971–8.PubMedGoogle Scholar
  4. Cammarota G, Longhini F, Perucca R, Ronco C, Colombo D, Messina A, Vaschetto R, Navalesi P. New setting of neurally adjusted ventilatory assist during noninvasive ventilation through a helmet. Anesthesiology. 2016;125(6):1181–9.View ArticlePubMedGoogle Scholar
  5. Antonelli M, Conti G, Pelosi P, Gregoretti C, Pennisi MA, Costa R, Severgnini P, Chiaranda M, Proietti R. New treatment of acute hypoxemic respiratory failure: noninvasive pressure support ventilation delivered by helmet–a pilot controlled trial. Crit Care Med. 2002;30(3):602–8.View ArticlePubMedGoogle Scholar
  6. Calderini E, Confalonieri M, Puccio PG, Francavilla N, Stella L, Gregoretti C. Patient-ventilator asynchrony during noninvasive ventilation: the role of expiratory trigger. Intensive Care Med. 1999;25(7):662–7.View ArticlePubMedGoogle Scholar
  7. Vignaux L, Vargas F, Roeseler J, Tassaux D, Thille AW, Kossowsky MP, Brochard L, Jolliet P. Patient-ventilator asynchrony during non-invasive ventilation for acute respiratory failure: a multicenter study. Intensive Care Med. 2009;35(5):840–6.View ArticlePubMedGoogle Scholar
  8. Navalesi P, Longhini F. Neurally adjusted ventilatory assist. Curr Opin Crit Care. 2015;21(1):58–64.View ArticlePubMedGoogle Scholar
  9. Colombo D, Cammarota G, Bergamaschi V, De Lucia M, Corte FD, Navalesi P. Physiologic response to varying levels of pressure support and neurally adjusted ventilatory assist in patients with acute respiratory failure. Intensive Care Med. 2008;34(11):2010–8.View ArticlePubMedGoogle Scholar
  10. Vaschetto R, Cammarota G, Colombo D, Longhini F, Grossi F, Giovanniello A, Della Corte F, Navalesi P. Effects of propofol on patient-ventilator synchrony and interaction during pressure support ventilation and neurally adjusted ventilatory assist*. Crit Care Med. 2014;42(1):74–82.View ArticlePubMedGoogle Scholar
  11. Schmidt M, Dres M, Raux M, Deslandes-Boutmy E, Kindler F, Mayaux J, Similowski T, Demoule A. Neurally adjusted ventilatory assist improves patient-ventilator interaction during postextubation prophylactic noninvasive ventilation. Crit Care Med. 2012;40(6):1738–44.View ArticlePubMedGoogle Scholar
  12. Bertrand PM, Futier E, Coisel Y, Matecki S, Jaber S, Constantin JM. Neurally adjusted ventilatory assist vs pressure support ventilation for noninvasive ventilation during acute respiratory failure: a crossover physiologic study. Chest. 2013;143(1):30–6.View ArticlePubMedGoogle Scholar
  13. Piquilloud L, Tassaux D, Bialais E, Lambermont B, Sottiaux T, Roeseler J, Laterre PF, Jolliet P, Revelly JP. Neurally adjusted ventilatory assist (NAVA) improves patient-ventilator interaction during non-invasive ventilation delivered by face mask. Intensive Care Med. 2012;38(10):1624–31.View ArticlePubMedGoogle Scholar
  14. Doorduin J, Sinderby CA, Beck J, van der Hoeven JG, Heunks L. Automated patient-ventilator interaction analysis during neurally adjusted non-invasive ventilation and pressure support ventilation in chronic obstructive pulmonary disease. Crit Care. 2014;18(5):550.View ArticlePubMedPubMed CentralGoogle Scholar
  15. Cammarota G, Olivieri C, Costa R, Vaschetto R, Colombo D, Turucz E, Longhini F, Della Corte F, Conti G, Navalesi P. Noninvasive ventilation through a helmet in postextubation hypoxemic patients: physiologic comparison between neurally adjusted ventilatory assist and pressure support ventilation. Intensive Care Med. 2011;37(12):1943–50.View ArticlePubMedGoogle Scholar
  16. Liu L, Xia F, Yang Y, Longhini F, Navalesi P, Beck J, Sinderby C, Qiu H. Neural versus pneumatic control of pressure support in patients with chronic obstructive pulmonary diseases at different levels of positive end expiratory pressure: a physiological study. Crit Care. 2015;19:244.View ArticlePubMedPubMed CentralGoogle Scholar
  17. Altman DG, Schulz KF, Moher D, Egger M, Davidoff F, Elbourne D, Gotzsche PC, Lang T. The revised CONSORT statement for reporting randomized trials: explanation and elaboration. Ann Intern Med. 2001;134(8):663–94.View ArticlePubMedGoogle Scholar
  18. Navalesi P, Frigerio P, Moretti MP, Sommariva M, Vesconi S, Baiardi P, Levati A. Rate of reintubation in mechanically ventilated neurosurgical and neurologic patients: evaluation of a systematic approach to weaning and extubation. Crit Care Med. 2008;36(11):2986–92.View ArticlePubMedGoogle Scholar
  19. Nava S, Gregoretti C, Fanfulla F, Squadrone E, Grassi M, Carlucci A, Beltrame F, Navalesi P. Noninvasive ventilation to prevent respiratory failure after extubation in high-risk patients. Crit Care Med. 2005;33(11):2465–70.View ArticlePubMedGoogle Scholar
  20. Ferrer M, Sellares J, Valencia M, Carrillo A, Gonzalez G, Badia JR, Nicolas JM, Torres A. Non-invasive ventilation after extubation in hypercapnic patients with chronic respiratory disorders: randomised controlled trial. Lancet. 2009;374(9695):1082–8.View ArticlePubMedGoogle Scholar
  21. SERVO-i VENTILATOR SYSTEM V6.0, User’s Manual, Rev 14 English edn: Maquet Critical Care AB; 2011.Google Scholar
  22. Chooi CS, White AM, Tan SG, Dowling K, Cyna AM. Pain vs comfort scores after Caesarean section: a randomized trial. Br J Anaesth. 2013;110(5):780–7.View ArticlePubMedGoogle Scholar
  23. Maggiore SM, Idone FA, Vaschetto R, Festa R, Cataldo A, Antonicelli F, Montini L, De Gaetano A, Navalesi P, Antonelli M. Nasal high-flow versus Venturi mask oxygen therapy after extubation. Effects on oxygenation, comfort, and clinical outcome. Am J Respir Crit Care Med. 2014;190(3):282–8.View ArticlePubMedGoogle Scholar
  24. Olivieri C, Longhini F, Cena T, Cammarota G, Vaschetto R, Messina A, Berni P, Magnani C, Della Corte F, Navalesi P. New versus conventional helmet for delivering noninvasive ventilation: a physiologic, crossover randomized study in critically ill patients. Anesthesiology. 2016;124(1):101–8.View ArticlePubMedGoogle Scholar
  25. Navalesi P, Hernandez P, Wongsa A, Laporta D, Goldberg P, Gottfried SB. Proportional assist ventilation in acute respiratory failure: effects on breathing pattern and inspiratory effort. Am J Respir Crit Care Med. 1996;154(5):1330–8.View ArticlePubMedGoogle Scholar
  26. Bellani G, Mauri T, Coppadoro A, Grasselli G, Patroniti N, Spadaro S, Sala V, Foti G, Pesenti A. Estimation of patient’s inspiratory effort from the electrical activity of the diaphragm. Crit Care Med. 2013;41(6):1483–91.View ArticlePubMedGoogle Scholar
  27. Olivieri C, Costa R, Conti G, Navalesi P. Bench studies evaluating devices for non-invasive ventilation: critical analysis and future perspectives. Intensive Care Med. 2012;38(1):160–7.View ArticlePubMedGoogle Scholar
  28. Olivieri C, Costa R, Spinazzola G, Ferrone G, Longhini F, Cammarota G, Conti G, Navalesi P. Bench comparative evaluation of a new generation and standard helmet for delivering non-invasive ventilation. Intensive Care Med. 2013;39(4):734–8.View ArticlePubMedGoogle Scholar
  29. Costa R, Navalesi P, Antonelli M, Cavaliere F, Craba A, Proietti R, Conti G. Physiologic evaluation of different levels of assistance during noninvasive ventilation delivered through a helmet. Chest. 2005;128(4):2984–90.View ArticlePubMedGoogle Scholar
  30. Moerer O, Beck J, Brander L, Costa R, Quintel M, Slutsky AS, Brunet F, Sinderby C. Subject-ventilator synchrony during neural versus pneumatically triggered non-invasive helmet ventilation. Intensive Care Med. 2008;34(9):1615–23.View ArticlePubMedPubMed CentralGoogle Scholar
  31. Costa R, Navalesi P, Spinazzola G, Ferrone G, Pellegrini A, Cavaliere F, Proietti R, Antonelli M, Conti G. Influence of ventilator settings on patient-ventilator synchrony during pressure support ventilation with different interfaces. Intensive Care Med. 2010;36(8):1363–70.View ArticlePubMedGoogle Scholar
  32. Lemyze M, Mallat J, Nigeon O, Barrailler S, Pepy F, Gasan G, Vangrunderbeeck N, Grosset P, Tronchon L, Thevenin D. Rescue therapy by switching to total face mask after failure of face mask-delivered noninvasive ventilation in do-not-intubate patients in acute respiratory failure. Crit Care Med. 2013;41(2):481–8.View ArticlePubMedGoogle Scholar
  33. Bonmarchand G, Chevron V, Chopin C, Jusserand D, Girault C, Moritz F, Leroy J, Pasquis P. Increased initial flow rate reduces inspiratory work of breathing during pressure support ventilation in patients with exacerbation of chronic obstructive pulmonary disease. Intensive Care Med. 1996;22(11):1147–54.View ArticlePubMedGoogle Scholar
  34. Bonmarchand G, Chevron V, Menard JF, Girault C, Moritz-Berthelot F, Pasquis P, Leroy J. Effects of pressure ramp slope values on the work of breathing during pressure support ventilation in restrictive patients. Crit Care Med. 1999;27(4):715–22.View ArticlePubMedGoogle Scholar
  35. Chiumello D, Pelosi P, Taccone P, Slutsky A, Gattinoni L. Effect of different inspiratory rise time and cycling off criteria during pressure support ventilation in patients recovering from acute lung injury. Crit Care Med. 2003;31(11):2604–10.View ArticlePubMedGoogle Scholar
  36. Prinianakis G, Delmastro M, Carlucci A, Ceriana P, Nava S. Effect of varying the pressurisation rate during noninvasive pressure support ventilation. Eur Respir J. 2004;23(2):314–20.View ArticlePubMedGoogle Scholar
  37. Chiumello D, Pelosi P, Croci M, Bigatello LM, Gattinoni L. The effects of pressurization rate on breathing pattern, work of breathing, gas exchange and patient comfort in pressure support ventilation. Eur Respir J. 2001;18(1):107–14.View ArticlePubMedGoogle Scholar
  38. Vargas F, Thille A, Lyazidi A, Campo FR, Brochard L. Helmet with specific settings versus facemask for noninvasive ventilation. Crit Care Med. 2009;37(6):1921–8.View ArticlePubMedGoogle Scholar
  39. Gerbershagen HJ, Rothaug J, Kalkman CJ, Meissner W. Determination of moderate-to-severe postoperative pain on the numeric rating scale: a cut-off point analysis applying four different methods. Br J Anaesth. 2011;107(4):619–26.View ArticlePubMedGoogle Scholar
  40. Gagliese L, Weizblit N, Ellis W, Chan VW. The measurement of postoperative pain: a comparison of intensity scales in younger and older surgical patients. Pain. 2005;117(3):412–20.View ArticlePubMedGoogle Scholar
  41. Gift AG, Narsavage G. Validity of the numeric rating scale as a measure of dyspnea. Am J Crit Care. 1998;7(3):200–4.PubMedGoogle Scholar