Open Access

Continuous glucose control in the ICU: report of a 2013 round table meeting

  • Jan Wernerman1,
  • Thomas Desaive2,
  • Simon Finfer3,
  • Luc Foubert4,
  • Anthony Furnary5,
  • Ulrike Holzinger6,
  • Roman Hovorka7,
  • Jeffrey Joseph8,
  • Mikhail Kosiborod9,
  • James Krinsley10,
  • Dieter Mesotten11,
  • Stanley Nasraway12,
  • Olav Rooyackers13,
  • Marcus J Schultz14,
  • Tom Van Herpe15, 16,
  • Robert A Vigersky17 and
  • Jean-Charles Preiser18Email author
Critical Care201418:226

DOI: 10.1186/cc13921

Published: 13 June 2014

Abstract

Achieving adequate glucose control in critically ill patients is a complex but important part of optimal patient management. Until relatively recently, intermittent measurements of blood glucose have been the only means of monitoring blood glucose levels. With growing interest in the possible beneficial effects of continuous over intermittent monitoring and the development of several continuous glucose monitoring (CGM) systems, a round table conference was convened to discuss and, where possible, reach consensus on the various aspects related to glucose monitoring and management using these systems. In this report, we discuss the advantages and limitations of the different types of devices available, the potential advantages of continuous over intermittent testing, the relative importance of trend and point accuracy, the standards necessary for reporting results in clinical trials and for recognition by official bodies, and the changes that may be needed in current glucose management protocols as a result of a move towards increased use of CGM. We close with a list of the research priorities in this field, which will be necessary if CGM is to become a routine part of daily practice in the management of critically ill patients.

Review

Introduction

Achieving adequate glucose control in ICU patients is complex and difficult to perform optimally. Until relatively recently, intermittent blood-gas analyzer and central laboratory measurements of blood glucose from arterial blood samples have been the only means of monitoring blood glucose levels [1]. However, intermittent measurements are limited by the workload associated with the sampling process and the potential that between-measurement events may be missed. With growing interest in the possible beneficial effects of continuous over intermittent monitoring and the development of several continuous glucose monitoring (CGM) systems, a round table conference was convened in March 2013 to discuss and, where possible, reach consensus on various aspects related to glucose monitoring and management. Leading experts in the field of glucose control in ICU patients and invited members of interested industry companies joined for presentation and discussion. After the meeting, a draft report was circulated to all participants by email for critical review. Representatives of the invited industry companies were asked to include a brief summary of their devices in the additional file of this report (Additional file 1), but, other than participation in the open discussion periods of the meeting, had no influence on content.

Continuous glucose monitoring

Definitions

CGM has been proposed as a means to improve management of dysglycemia. Although termed ‘continuous’, current systems still sample intermittently, with a measurement interval of a few milliseconds up to 15 minutes. Some systems average the frequent intermittent measurements and display them as a single reading or moving average, updated regularly. Nevertheless, such measurements can be considered as having ‘real-time’ value especially when compared to their intermittent counterparts, although physiological or data processing lag time may be present depending on the sampled body fluid. Two factors can be considered when defining ‘continuous’: the frequency of actual glucose measurements and the immediacy of the data display. Clearly, measurements need to be frequent enough to capture all glucose dynamics. Based on current knowledge of the physiology of glucose and insulin metabolism in non-critically ill patients [2], an interval of 10 to 15 minutes between measurements is the likely maximum interval that would detect most glycemic dynamics, although faster dynamics may be observed when parenteral nutrition is modified and particularly when an intravenous glucose bolus is administered. The Clinical and Laboratory Standards Institute (CLSI) guidelines use 15 minutes as the cutoff for their definition of continuous monitoring [3], but which cutoff should be used to separate ‘continuous’ from ‘frequent intermittent’ sampling is debatable. More data on glucose trends in the critically ill are needed before clinically relevant sampling frequencies can be defined. The real-time output of CGM devices should be as instantaneous as possible, although there will generally be a lag period, the duration of which will depend on the site and frequency of sampling and data processing. The continuous display enables trends to be identified and visualized.

Importantly, the purposes of any such device are to improve clinically relevant outcomes and to reduce associated nursing workload and ideally costs. Although the overall accuracy of many CGM systems is less than that of intermittent systems using central laboratory testing [4], this limitation is to some degree mitigated by the ability to follow the direction of change in glucose levels, theoretically allowing earlier intervention to maintain blood glucose concentrations within acceptable ranges. A less-often cited advantage is the decreased need for multiple finger-pricks or blood pulls with a continuous system, which may reduce patient discomfort and nurse workload [5, 6].

Several CGM systems are now available for clinical use and early results from clinical trials in critically ill adults [714] and children [15, 16] have been published. However, no studies have assessed clinical outcomes using the continuous approach compared to an intermittent system; furthermore, the different sensors used, the different comparators, and the lack of standardized performance metrics make it difficult to compare results.

Overview of techniques for glucose measurement

The three predominant techniques currently used for continuously measuring glucose levels in the ICU involve glucose oxidase, mid-infrared spectroscopy and fluorescence.

The glucose oxidase technique is based on the sensing of hydrogen peroxide (H2O2) released when glucose is converted to glucolactone: the greater the concentration of glucose, the more H2O2 will be released and the stronger the signal. Results can be influenced by interference from molecules other than glucose (for example, uric acid, acetaminophen and salicylic acid) which oxidize the H2O2.

Mid-infrared spectroscopy detects an absorption spectrum for glucose in plasma using different wavelength filters.

Fluorescence techniques rely on quenched chemical fluorescence to measure glucose concentration [17]. Fluorescence glucose sensors are associated with a foreign body response, are sensitive to local pH and/or oxygen, and require a light source.

Monitoring sites: clinical experience

Glucose can be measured in whole blood, plasma, interstitial fluid, and microdialysis fluid and values will vary according to which fluid is used [18]. Generally, plasma glucose is considered the ‘gold standard’. Glucose dissolves in water and because plasma has a higher water concentration (approximately 93%) than do red blood cells (approximately 71%), plasma will have a higher glucose concentration than will whole blood. The difference in laboratory-measured glucose concentration between whole blood and plasma will also vary with the hematocrit. Because some glucose diffuses from the plasma to interstitial fluid and tissues as blood circulates through the capillary system, arterial blood glucose is usually higher than venous glucose. Arterial blood glucose and capillary blood glucose are generally similar, although when blood glucose levels change rapidly, there may be a delay before similar changes are seen in capillary blood. Microdialysis concentrations tend to be slightly lower than those present in the surrounding tissue or blood.

The degree of invasiveness of a CGM technique varies from highly invasive (for example, intravascular devices) through the minimally invasive subcutaneous techniques, to non-invasive transdermal devices. Although studies comparing the accuracy and performance of more versus less invasive CGM systems have not yet been performed, preliminary data suggest that, moving through the spectrum from invasive to non-invasive, accuracy generally decreases as does the risk of complications, including infections. The type of monitor selected should be adjusted to patient characteristics, including the severity of illness of the patient and the type of access available. For example, a severely ill, unstable ICU patient will likely already have arterial and/or central venous lines in situ allowing invasive intravascular monitoring, whereas a stable patient ready for ward transfer can be monitored using a less- or non-invasive device. Moreover, severely ill patients are more likely to be receiving mechanical ventilation and/or sedative agents, making clinical symptoms of hypoglycemia more difficult to detect and perhaps arguing in favor of the more accurate invasive devices. When comparing devices it is essential to state which reference measurement technique is used so that results can be easily compared. Whenever possible, arterial glucose measurements with a blood gas analyzer or by a central laboratory should be used as the comparator as these are the most accurate and reproducible [1]; when this is not possible, or when the device under study uses venous sampling, venous blood glucose should be used as the comparator. When venous sampling is used, the specific vessel should be defined.

Intravascular CGM devices can be divided into three groups: (1) those that have an intravascular sensor inserted into the lumen of an artery or peripheral/central vein and directly measure the blood glucose concentration without consuming blood in the process; (2) those in which a small blood sample is taken from the intravascular catheter and passed over an external sensor; and (3) those in which a blood sample is re-circulated after passing through an external sensor without blood loss. The accuracy of intravascular microdialysis probes will vary according to their position - for example, if integrated into the central venous catheter, a much larger membrane will be possible than if positioned in a smaller peripheral vein catheter, allowing a greater area for equilibration and a more rapid and reliable result [19]. Recent studies using a central venous catheter with a microdialysis membrane have demonstrated good agreement between microdialysis glucose measurements and reference venous and arterial blood gas values in patients undergoing major abdominal surgery or cardiac surgery [20, 21].

Interstitial fluid glucose is generally measured with subcutaneous probes, often inserted on the abdominal wall or upper arm. Interstitial fluid glucose levels depend on the rate of glucose diffusion from plasma to the interstitial fluid and the rate of uptake by subcutaneous tissue cells; hence, they are influenced by blood flow, the metabolic rate of adjacent cells, capillary permeability, degree of hydration or edema, and so on, all of which may be altered in critically ill patients [18]. Several subcutaneous devices have been tested in critically ill patients and have been shown to have good agreement with reference arterial and venous samples [12, 2224]. Moreover, similar accuracy has been reported in critically ill patients with and without shock requiring norepinephrine therapy [22], and in cardiac surgery versus non-critically ill patients [25, 26]. Nevertheless, the accuracy of interstitial fluid monitoring needs to be further investigated, in particular in unstable patients. One concern with subcutaneous interstitial fluid probes is the tissue trauma created during insertion, such that measurements may be less accurate for several hours after insertion. There is a time lag between change in blood glucose and that measured in the interstitial fluid, which is, however, unlikely to result in ineffective treatment in case of an emerging hypo- or hyperglycemic event [2730]. The clinical relevance of this time-lag needs to be contrasted against current practice with a typical delay of 5 to 10 minutes to take the sample and to measure glucose on an analyzer.

Transcutaneous devices are also being developed. One such device uses a biosensor that can measure transdermal glucose flux, which is proportional to the blood glucose concentration. The skin is prepared by microabrasion to remove the dead surface cells and the biosensor then applied, using the glucose oxidase reaction to create a measureable signal for interstitial glucose. In pilot studies of cardiac surgery patients, good agreement with peripheral blood was demonstrated [31, 32].

All techniques have limitations related in part to the sampling site used (venous, arterial or capillary blood, plasma, and interstitial fluid) [18], but also to the need for anticoagulation with some intravascular devices, problems of local inflammation, and need for recalibration. Rice and Coursin [33] recently proposed a list of attributes for the ‘ideal’ CGM system (Box 1). For all CGM systems, specific performance characteristics related to the clinical utility of the system need to be clearly defined (Box 2).

Trend accuracy versus point accuracy

One of the key advantages of CGM systems is their ability to identify and display trends in blood glucose measurement. Hence, when considering the performance of these devices, additional metrics may need to be developed to complement current assessment of accuracy. Point accuracy is defined as the difference between the current displayed blood glucose value and the current true blood glucose value. Trend accuracy is defined as the degree to which an estimate of the rate of change in blood glucose concentration over a given time interval approximates the true rate of change. Further research is required to establish the duration over which trend accuracy should be calculated and the relative importance of point accuracy versus trend accuracy in terms of clinical outcomes.

In theory, the use of trending could have several potential advantages over individual values (Figure 1), including: that it is less sensitive to random noise, because, if present, noise will be filtered out by the trend line, at least when the period used to calculate the trend is long enough; there is little effect of bias - the presence of a constant over- or under-shoot of the value will not affect the trend. However, there are also several disadvantages. First, there is a lag time when calculating the trend that will depend on the frequency of sampling and the number of measurements and time-lapse over which those measures are used. With longer time intervals between measurements, trending will reflect real changes less accurately, certainly when changes are rapid and intervals are long. Second, if there is a lag time or a bias, extrapolation of the trend line can amplify the error. Third, most current glycemic control protocols rely on PID (proportional, integral, derivative) control with insulin rates determined as a function of the current blood glucose (P), accumulation of historical blood glucose values (I), and the trend (rate of change) in blood glucose (D). Hence, for current protocols, all three aspects need to be accurate; it is not sufficient just to have accurate trend accuracy - point accuracy also needs to be good.
Figure 1

Schematic representation of the potential advantages of using trends. (A) If imprecision or noise is random or normally distributed, the trend line will filter it out. (B) If the measurement system has a fixed bias, trend will not be affected but individual values could be. (C) When trying to predict future events, trend may be clinically more important than the current absolute blood glucose value.

Thus, at the present time, both good point accuracy and good trend accuracy are required to achieve optimal glycemic control. However, the more continuous the measurement, the clearer and more reliable the trend will become. In the future, use of algorithms designed specifically for CGM may also reduce the need for highly accurate point measurements. The period of time over which trend should be assessed will depend on lag time and may also depend on the type of patient.

Standards for reporting performance

Standards for reporting of clinical trials of CGM systems need to be developed so that results can be easily compared. In this context, we can consider factors related to the patients and the device and those related to the impact of the device on clinical outcomes. In terms of the device and patient, several aspects need to be reported regarding demographics (age, gender, comorbidities, including diabetes, disease severity), use of vasoactive drugs, design (single-center versus multicenter, type of center, number of samples, comparator), glucose targets (target range), definition of hypoglycemia and hyperglycemia, time in range, analytical and clinical accuracy, number of patients unable to monitor and reasons, down-time (time needed for calibration when no signal/reading available) and time to display, and safety (bleeding events, infections, outliers, alarm performance).

In terms of characterization of accuracy of the system being tested versus the comparator, the Bland-Altman plot remains indispensable, showing the difference between the two measurements either as a function of the average of the two measurements or, when there is a ‘gold standard’, as a function of the comparator [34]. The 95% confidence interval (1.96 × standard deviation) of a tested blood glucose meter against a gold standard can be deduced from these plots. Various grid systems have also been proposed, of which the Clarke error grid [35] is currently the most widely used. However, this grid was not designed for CGM systems and does not reflect rapid changes in the blood glucose level or account for potential errors in insulin dosing. As such, the so-called continuous glucose error grid analysis (cEGA) has been proposed, which is designed to capture errors in the rate and direction of change in glucose between measurement methods [36]. This technique, initially developed for outpatient care, is an interesting approach but relatively complex, requiring specific software and frequent sampling [37]. The R-deviation is another potential metric to assess the accuracy of CGM systems [38]. This value is a numerical metric of rate of change accuracy, based on the deviation between the rates of change in reference and test sensor glucose fluctuations.

How to report on the impact of a device when used with a treatment protocol is perhaps less clear. For this purpose, three domains of glycemic control can be considered: hyperglycemia, hypoglycemia, and glycemic variability [39]; glucose complexity has been suggested as a possible fourth domain [40]. The three domains are all associated with increased mortality in critically ill patients [39] and, as such, the number and duration of hypo- and hyperglycemic episodes (using pre-specified parameters), the time in target, the degree of glucose variability (and possibly complexity) should all be reported when assessing the clinical impact of a new device, in addition to clinical outcomes, including mortality and morbidity measures. Further study is needed to determine how best to define trend and hypoglycemia (including sensitivity and specificity) for regulatory approval (see below).

Alarms, warning signals

Alarms on CGM systems generally concern the three domains of glycemic control listed earlier. Determining at which value alarms should be set for each domain remains difficult. The clinical impact of hypo/hyperglycemia will vary according to the degree and time away from normal values (Figure 2), with considerable overlap among individuals. Several studies have suggested that patients with acute coronary syndromes and severe brain injury may be more sensitive to low blood glucose levels [41, 42], at least in the absence of tight glucose control [43]. Therefore, in some groups of critically ill patients, target glucose ranges may need to be set higher than in other groups. Generally, a blood glucose ≤40 mg/dl is considered to represent severe hypoglycemia [1, 44] and a level of 41 to 70 mg/dl moderate hypoglycemia [1], but studies have used different definitions. Hyperglycemia is variably considered as values >140 or 180 mg/dl. Glycemic variability is even more difficult to define; a relatively high value of the coefficient of variation of >20% has been suggested to define high variability, because it is associated with worse outcomes than values <20%. Variability is also related to ongoing therapy. Glycemic targets will also vary according to individual patient characteristics, including age, comorbidities (notably diabetes), type of patient (for example, surgical versus medical), and so on. Alarm settings therefore need to be adjustable for individual patients. Further study is needed to define optimal target ranges for different groups of patients and to clarify the impact of alarms on clinical practice and patient outcomes. With the development of better validated CGM systems and better knowledge of glucose trends in the critically ill, alarms for trend changes will be developed and have the potential to prevent hyper- and hypoglycemia. Predictive alerts are already in use on some devices inserted subcutaneously.
Figure 2

The clinical impact of hypo/hyperglycemia varies according to the degree away from normal values.

Criteria for approval by the official bodies

In terms of safety and effectiveness, it is unclear which metrics should be used to indicate sufficient accuracy and reliability. The CLSI has produced new standards for point-of-care testing [45]; however, these standards may not be applicable to CGM systems. In our 2013 Consensus document, we suggested that the minimum standard for glucose meters to be used in critically ill patients should be that 98% of readings should be within 12.5% of a reference standard (or within ±10 mg/dl for readings <100 mg/dl); the remaining 2% of readings should be within 20% of a reference standard [1]. The mean absolute relative difference should be cited and values should be <14%; values >18% are considered to represent poor accuracy. For trend accuracy there is not yet an accepted metric. The R-deviation may be useful, but further study is needed [38]. Other concerns that need to be addressed include signal stability, drift, variability, and drop-out; potential interferences (for example, acidosis, hematocrit, bilirubin, hemoglobin, medications and intravenous fluids); edema and nutritional status; and number and characterization of outliers. As yet, there are no clearly defined metrics for reporting what is sufficient in terms of accuracy and reliability.

Insulin algorithms

An algorithm can be defined as ’a formalized sequence of instructions for solving a complex problem in finite processing steps’ [46]. Algorithms in the field of tight glucose control are used to standardize care, for quality assurance and to avoid intuitive decision making. An optimal system should be accurate, safe, efficacious, simple to use, reliable, flexible for different patient populations, assessable in real-time, fit into workflow, require a low number of glucose measurements (if not based on CGM), and take into account inter- and intrapatient variability and carbohydrate intake. Algorithms should incorporate dynamic scale protocols, instead of static sliding-scale protocols [47]. Although early algorithms were paper-based [48, 49], increasingly, glucose control algorithms are computer-based, enabling more complex protocols to be developed. Several studies have demonstrated improved glucose control with computer-based compared to paper-based algorithms [5052]. A common type of algorithm is the PID algorithm in which deviation of the blood glucose value from the target range is corrected by adjusting the dose of insulin using a linear combination of absolute deviation, trend, and the sum of past deviations [53]. Another main type of glucose algorithm used in critical care is the model-based or model-predictive control algorithm, which adjusts insulin dose according to a mathematical model of the relationship between blood glucose and insulin [54, 55]. Better standardization of algorithm development is needed [56].

Many algorithms for glucose control have been developed. Wilson and colleagues [57] identified 12 different algorithms and, using blood glucose records from an actual hyperglycemic patient, calculated the insulin doses that would have been recommended by each protocol. There was considerable variability among protocols in patterns and ranges of recommended insulin dose (range 27 to 115 units), and adjustments to dose when nearing target glucose. Protocols therefore behave differently and may have greater influence on outcomes than the glucose measurement error. Different algorithms may be better suited to various patient populations or clinical settings.

Clinical testing and comparison of algorithms is resource intensive in terms of patients, staff, time, and costs. Moreover, the majority of algorithms for glycemic control in the ICU are based on intermittent glucose measurements and new algorithms will need to be designed if CGM systems become more widely used. When comparing algorithms, standard glucose-centric outcomes need to be reported, including numbers of hypo- and hyperglycemic episodes. One useful parameter that has been suggested is the cumulative time-in-band, which calculates the percentage of blood glucose values that fall within a specified range. This measure is independent of sampling frequency, can be applied to all algorithms and is simple to calculate; however, it is only useful when comparing algorithms that target the same blood glucose band.

In silico simulation models using ‘virtual’ ICU patients have been suggested to reduce some of the burden of clinical algorithm comparisons and to accelerate the assessment process. These systems can be used to optimize design parameters and safety features, test effects of changes in nutrition or other medications and interventions, and assess effects of measurement errors or delays. At least four currently available ICU simulators are known: the Cambridge [58], Virginia [59], Leuven [60], and Christchurch [61] models and simulation models are beginning to be used in the critical care setting of glycemic control. Wilinska and colleagues used simulation to compare the effects of different algorithms [62] and the performance of a newly proposed algorithm [63]; the study results were reproduced in the simulated populations. Although these systems need further study, it seems likely that the virtual patient will play an increasingly large role in the ongoing development of CGM systems and glycemic control protocols in the ICU setting.

The development of closed loop systems, which deliver insulin in a glucose responsive fashion every 1 to 15 minutes based on CGM measurements, is the most promising approach to improve glucose control once CGM becomes routinely available. Closed loop systems are being aggressively pursued and may help modulate glucose delivery to further reduce the risk of hypoglycemia. Automated closed-loop glucose control based on continuous subcutaneous glucose measurements and model predictive control in critically ill adults was associated with better glycemic control compared to a local sliding scale protocol [64].

Priorities for research

The expert group defined eight areas where research should help to advance glucose monitoring in the near future to the likely benefit of critically ill patients. First, the different devices for CGM need to be better validated in terms of accuracy and reliability. Head-to-head comparisons are needed, particularly for devices sampling different compartments. Second, the clinical relevance of inaccuracies in glucose measurements should be shown in error grids adapted to current therapeutic algorithms. Third, glucose trends in critically ill patients and subgroups need to be more clearly characterized, so that better definitions of the rate of change can be developed and, thereby, the frequency of sampling needed to describe clinically relevant trends. Fourth, the effect of different insulin treatment algorithms on glucose variability should be studied with development of new and enhancement of existing glucose control protocols based on CGM. Fifth, development and validation of metrics for trend accuracy are required. Sixth, universal metrics to assess glycemic control and blood glucose variability that could be used with continuous as well as intermittent data should be defined and agreed upon. Seventh, at a later time-point, randomized controlled trials need to be conducted assessing the effects of CGM versus intermittent systems on outcome in critically ill patients, including assessment of patient-centered outcome measures (glycemic control and morbidity incidence). Eighth, closed loop systems for glucose control in critically ill patients should be developed and eventually validated and assessed in randomized controlled trials as above.

Conclusion

CGM mandates the development of new approaches to the analysis of parameters of glucose regulation, such as glucose variability and glucose complexity, and also provides a tool to help effect these analyses. While CGM systems clearly have the potential to improve blood glucose control and patient outcomes, this remains a potential that has not yet been demonstrated in clinical practice. Future studies may be able to demonstrate real clinical benefits and reveal the optimal use of the different CGM systems (which system for which patient). When discussing how best to assess CGM, different goals can be considered, including maintenance of specified target levels, which may vary in different patient populations; avoidance of hypoglycemic events; assessment of glucose variability; and degree of glucose complexity. Most important, however, will be the impact of each device on clinical outcomes, including better glucose control and fewer hypoglycemic episodes; this is of far more relevance to clinicians and patients than small differences in accuracy.

Box 1. Suggested criteria for the ideal continuous glucose monitoring system [33]

Rapid: very little lag between blood glucose and the measured value.

Accurate: each measurement should be within accuracy guidelines suggested recently [1].

Free of interference: minimal, if any, important interference, such as drugs or physiologic perturbations.

Inert: the sensor should not react with the tissue or form a coating rendering the device inaccurate over time.

Robust: the system must be able to perform within the dynamic and busy ICU and operative setting.

Minimally invasive

Cost-effective

Box 2. Performance characteristics related to the clinical utility of continuous glucose monitoring systems that need to be clearly defined for each system

Frequency of sampling

Delay to display

Lag time

Biofilm development

Measurement accuracy

Reliability (time to sensor failure, frequency and duration of data gaps)

Need for and frequency of calibration

Ability to recognize and correct for interference

Automation

Need for anti-coagulation

Safety

Site of access

Handling of outlier values

Alarms

Clinical effectiveness (that is, impact on glucose control and prevention of hypoglycemia)

Cost-effectiveness

Possibility of combining glucose monitoring with other measurements

Abbreviations

CGM: 

Continuous glucose monitoring

CLSI: 

Clinical and Laboratory Standards Institute

PID: 

Proportional, integral, derivative.

Declarations

Acknowledgements

The authors are very grateful to Dr Karen Pickett for her assistance in the writing of this report. Funding: this meeting was supported by an unrestricted educational grant from B Braun, EchoTherapeutics, Edwards, Glumetrics, Glysure, Maquet, Medtronic, Menarini, and Optiscan; these companies were invited to include a brief summary of their devices in the additional file of this report, but had no influence on other content or the decision to publish.

Authors’ Affiliations

(1)
Department of Anesthesiology, Intensive Care Medicine, K32, Karolinska University Hospital
(2)
GIGA - Cardiovascular Sciences, University of Liege, Institute of Physics, B5
(3)
The George Institute for Global Health and Royal North Shore Hospital, University of Sydney
(4)
Department of Anesthesia, Intensive Care Medicine, OLV Clinic
(5)
Starr-Wood Cardiac Group
(6)
Department of Medicine III - Division of Gastroenterology and Hepatology, Medical University of Vienna
(7)
University of Cambridge Metabolic Research Laboratories, Level 4, Wellcome trust MRC Institute of Metabolic Science, Addenbrooke’s Hospital
(8)
Jefferson Artificial Pancreas Center and Anesthesiology Program for Translational Research, Department of Anesthesiology, Jefferson Medical College of Thomas Jefferson University
(9)
Saint-Luke’s Mid America Heart Institute, University of Missouri - Kansas City
(10)
Division of Critical Care, Stamford Hospital and Columbia University College of Physicians and Surgeons
(11)
Department of Intensive Care Medicine, University Hospitals Leuven
(12)
Surgical Intensive Care Units, Tufts Medical Center
(13)
Anesthesiology and Intensive Care Clinic, Karolinska Institute and University Hospital
(14)
Department of Intensive Care Medicine, Academic Medical Center at the University of Amsterdam
(15)
Department of Intensive Care Medicine, University Hospitals Leuven
(16)
Department of Electrical Engineering (STADIUS) - iMinds Future Health Department, Katholieke Universiteit Leuven
(17)
Diabetes Institute, Walter Reed National Military Medical Center
(18)
Department of Intensive Care, Erasme Hospital, Université libre de Bruxelles

References

  1. Finfer S, Wernerman J, Preiser JC, Cass T, Desaive T, Hovorka R, Joseph JI, Kosiborod M, Krinsley JS, MacKenzie I, Mesotten D, Schulz M, Scott MG, Slingerland R, van den Berghe G, Van Herpe T: Consensus recommendations on measurement of blood glucose and reporting glycemic control in critically ill adults. Crit Care 2013, 17: 229. 10.1186/cc12537PubMedPubMed CentralView ArticleGoogle Scholar
  2. Gough DA, Kreutz-Delgado K, Bremer TM: Frequency characterization of blood glucose dynamics. Ann Biomed Eng 2003, 31: 91-97.PubMedView ArticleGoogle Scholar
  3. Klonoff DC, Bernhardt P, Ginsberg BH, Josepg J, Mastrototaro J, Parker D, Vesper H, Vigersky R: POCT05-A: Performance Metrics for Continuous Interstitial Glucose Monitoring; Approved Guideline. Wayne, PA: Clinical and Laboratory Standards Institute; 2008.Google Scholar
  4. Inoue S, Egi M, Kotani J, Morita K: Accuracy of blood glucose measurements using glucose meters and arterial blood gas analyzers in critically ill adult patients: systematic review. Crit Care 2013, 17: R48. 10.1186/cc12567PubMedPubMed CentralView ArticleGoogle Scholar
  5. Aragon D: Evaluation of nursing work effort and perceptions about blood glucose testing in tight glycemic control. Am J Crit Care 2006, 15: 370-377.PubMedGoogle Scholar
  6. Inzucchi SE, Kosiborod M: Continuous glucose monitoring during critical care. Anesthesiology 2011, 114: 18-19. 10.1097/ALN.0b013e3181ff419ePubMedView ArticleGoogle Scholar
  7. Schierenbeck F, Franco-Cereceda A, Liska J: Evaluation of a continuous blood glucose monitoring system using central venous microdialysis. J Diabetes Sci Technol 2012, 6: 1365-1371. 10.1177/193229681200600615PubMedPubMed CentralView ArticleGoogle Scholar
  8. Lorencio C, Leal Y, Bonet A, Bondia J, Palerm CC, Tache A, Sirvent JM, Vehi J: Real-time continuous glucose monitoring in an intensive care unit: better accuracy in patients with septic shock. Diabetes Technol Ther 2012, 14: 568-575. 10.1089/dia.2012.0008PubMedPubMed CentralView ArticleGoogle Scholar
  9. Holzinger U, Warszawska J, Kitzberger R, Wewalka M, Miehsler W, Herkner H, Madl C: Real-time continuous glucose monitoring in critically ill patients: a prospective randomized trial. Diabetes Care 2010, 33: 467-472. 10.2337/dc09-1352PubMedPubMed CentralView ArticleGoogle Scholar
  10. De Block C, Manuel-Y-Keenoy B, Van Gaal L, Rogiers P: Intensive insulin therapy in the intensive care unit: assessment by continuous glucose monitoring. Diabetes Care 2006, 29: 1750-1756. 10.2337/dc05-2353PubMedView ArticleGoogle Scholar
  11. Yue XY, Zheng Y, Cai YH, Yin NN, Zhou JX: Real-time continuous glucose monitoring shows high accuracy within 6 hours after sensor calibration: a prospective study. PLoS One 2013, 8: e60070. 10.1371/journal.pone.0060070PubMedPubMed CentralView ArticleGoogle Scholar
  12. Brunner R, Kitzberger R, Miehsler W, Herkner H, Madl C, Holzinger U: Accuracy and reliability of a subcutaneous continuous glucose-monitoring system in critically ill patients. Crit Care Med 2011, 39: 659-664. 10.1097/CCM.0b013e318206bf2ePubMedView ArticleGoogle Scholar
  13. Rooyackers O, Blixt C, Mattsson P, Wernerman J: Continuous glucose monitoring by intravenous microdialysis. Acta Anaesthesiol Scand 2010, 54: 841-847. 10.1111/j.1399-6576.2010.02264.xPubMedView ArticleGoogle Scholar
  14. Price GC, Stevenson K, Walsh TS: Evaluation of a continuous glucose monitor in an unselected general intensive care population. Crit Care Resusc 2008, 10: 209-216.PubMedGoogle Scholar
  15. Steil GM, Langer M, Jaeger K, Alexander J, Gaies M, Agus MS: Value of continuous glucose monitoring for minimizing severe hypoglycemia during tight glycemic control. Pediatr Crit Care Med 2011, 12: 643-648. 10.1097/PCC.0b013e31821926a5PubMedPubMed CentralView ArticleGoogle Scholar
  16. Agus MS, Steil GM, Wypij D, Costello JM, Laussen PC, Langer M, Alexander JL, Scoppettuolo LA, Pigula FA, Charpie JR, Ohye RG, Gaies MG: Tight glycemic control versus standard care after pediatric cardiac surgery. N Engl J Med 2012, 367: 1208-1219. 10.1056/NEJMoa1206044PubMedPubMed CentralView ArticleGoogle Scholar
  17. Klonoff DC: Overview of fluorescence glucose sensing: a technology with a bright future. J Diabetes Sci Technol 2012, 6: 1242-1250. 10.1177/193229681200600602PubMedPubMed CentralView ArticleGoogle Scholar
  18. Cengiz E, Tamborlane WV: A tale of two compartments: interstitial versus blood glucose monitoring. Diabetes Technol Ther 2009,11(Suppl 1):S11-S16.PubMedGoogle Scholar
  19. Rooyackers O, Blixt C, Mattsson P, Wernerman J: Continuous glucose monitoring by intravenous microdialysis: influence of membrane length and dialysis flow rate. Acta Anaesthesiol Scand 2013, 57: 214-219. 10.1111/j.1399-6576.2012.02787.xPubMedView ArticleGoogle Scholar
  20. Blixt C, Rooyackers O, Isaksson B, Wernerman J: Continuous on-line glucose measurement by microdialysis in a central vein. A Pilot Study Crit Care 2013, 17: R87.PubMedGoogle Scholar
  21. Schierenbeck F, Owall A, Franco-Cereceda A, Liska J: Evaluation of a continuous blood glucose monitoring system using a central venous catheter with an integrated microdialysis function. Diabetes Technol Ther 2013, 15: 26-31. 10.1089/dia.2012.0169PubMedView ArticleGoogle Scholar
  22. Holzinger U, Warszawska J, Kitzberger R, Herkner H, Metnitz PG, Madl C: Impact of shock requiring norepinephrine on the accuracy and reliability of subcutaneous continuous glucose monitoring. Intensive Care Med 2009, 35: 1383-1389. 10.1007/s00134-009-1471-yPubMedView ArticleGoogle Scholar
  23. Rabiee A, Andreasik V, Abu-Hamdah R, Galiatsatos P, Khouri Z, Gibson BR, Andersen DK, Elahi D: Numerical and clinical accuracy of a continuous glucose monitoring system during intravenous insulin therapy in the surgical and burn intensive care units. J Diabetes Sci Technol 2009, 3: 951-959. 10.1177/193229680900300443PubMedPubMed CentralView ArticleGoogle Scholar
  24. Kosiborod M, Gottlieb R, Sekella J, Peterman D, Grodzinsky A, Kennedy P, Borkon M: Performance of the Medtronic Sentrino continuous glucose management system in the cardiac ICU [abstract]. Crit Care 2013,17(Suppl 2):462.Google Scholar
  25. Siegelaar SE, Barwari T, Hermanides J, Stooker W, van der Voort PHJ, Devries JH: Accuracy and reliability of continuous glucose monitoring in the intensive care unit: a head-to-head comparison of two subcutaneous glucose sensors in cardiac surgery patients. Diabetes Care 2011, 34: e31. 10.2337/dc10-1882PubMedPubMed CentralView ArticleGoogle Scholar
  26. Van Herpe T, Mesotten D: Blood glucose measurements in critically ill patients. J Diabetes Sci Technol 2012, 6: 22-28. 10.1177/193229681200600104PubMedPubMed CentralView ArticleGoogle Scholar
  27. Kamath A, Mahalingam A, Brauker J: Analysis of time lags and other sources of error of the DexCom SEVEN continuous glucose monitor. Diabetes Technol Ther 2009, 11: 689-695. 10.1089/dia.2009.0060PubMedView ArticleGoogle Scholar
  28. Garg SK, Voelmle M, Gottlieb PA: Time lag characterization of two continuous glucose monitoring systems. Diabetes Res Clin Pract 2010, 87: 348-353. 10.1016/j.diabres.2009.11.014PubMedView ArticleGoogle Scholar
  29. Wei C, Lunn DJ, Acerini CL, Allen JM, Larsen AM, Wilinska ME, Dunger DB, Hovorka R: Measurement delay associated with the Guardian RT continuous glucose monitoring system. Diabet Med 2010, 27: 117-122. 10.1111/j.1464-5491.2009.02887.xPubMedPubMed CentralView ArticleGoogle Scholar
  30. Basu A, Dube S, Slama M, Errazuriz I, Amezcua JC, Kudva YC, Peyser T, Carter RE, Cobelli C, Basu R: Time lag of glucose from intravascular to interstitial compartment in humans. Diabetes 2013, 62: 4083-4087. 10.2337/db13-1132PubMedPubMed CentralView ArticleGoogle Scholar
  31. Chuang H, Trieu MQ, Hurley J, Taylor EJ, England MR, Nasraway SA Jr: Pilot studies of transdermal continuous glucose measurement in outpatient diabetic patients and in patients during and after cardiac surgery. J Diabetes Sci Technol 2008, 2: 595-602. 10.1177/193229680800200410PubMedPubMed CentralView ArticleGoogle Scholar
  32. Nasraway SA, Ehsan A, Melanson AM, Menzie W, Trieu MQ, Berlin J, Hurley J, Krystyniak K, Segal S: Accuracy of a novel non-invasive transdermal continuous glucose monitor in critically ill patients [abstract]. Crit Care Med 2012, 40: S305. 10.1097/CCM.0b013e318232646bView ArticleGoogle Scholar
  33. Rice MJ, Coursin DB: Continuous measurement of glucose: facts and challenges. Anesthesiology 2012, 116: 199-204. 10.1097/ALN.0b013e318236abf6PubMedView ArticleGoogle Scholar
  34. Bland JM, Altman DG: Statistical methods for assessing agreement between two methods of clinical measurement. Lancet 1986, 1: 307-310.PubMedView ArticleGoogle Scholar
  35. Clarke WL, Cox D, Gonder-Frederick LA, Carter W, Pohl SL: Evaluating clinical accuracy of systems for self-monitoring of blood glucose. Diabetes Care 1987, 10: 622-628. 10.2337/diacare.10.5.622PubMedView ArticleGoogle Scholar
  36. Kovatchev BP, Gonder-Frederick LA, Cox DJ, Clarke WL: Evaluating the accuracy of continuous glucose-monitoring sensors: continuous glucose-error grid analysis illustrated by TheraSense Freestyle Navigator data. Diabetes Care 2004, 27: 1922-1928. 10.2337/diacare.27.8.1922PubMedView ArticleGoogle Scholar
  37. Wentholt IM, Hoekstra JB, Devries JH: A critical appraisal of the continuous glucose-error grid analysis. Diabetes Care 2006, 29: 1805-1811. 10.2337/dc06-0079PubMedView ArticleGoogle Scholar
  38. Clarke WL, Kovatchev B: Continuous glucose sensors: continuing questions about clinical accuracy. J Diabetes Sci Technol 2007, 1: 669-675. 10.1177/193229680700100510PubMedPubMed CentralView ArticleGoogle Scholar
  39. Krinsley JS, Egi M, Kiss A, Devendra AN, Schuetz P, Maurer PM, Schultz MJ, van Hooijdonk RT, Kiyoshi M, Mackenzie IM, Annane D, Stow P, Nasraway SA, Holewinski S, Holzinger U, Preiser JC, Vincent JL, Bellomo R: Diabetic status and the relation of the three domains of glycemic control to mortality in critically ill patients: an international multicenter cohort study. Crit Care 2013, 17: R37. 10.1186/cc12547PubMedPubMed CentralView ArticleGoogle Scholar
  40. Brunner R, Adelsmayr G, Herkner H, Madl C, Holzinger U: Glycemic variability and glucose complexity in critically ill patients: a retrospective analysis of continuous glucose monitoring data. Crit Care 2012, 16: R175. 10.1186/cc11657PubMedPubMed CentralView ArticleGoogle Scholar
  41. Kosiborod M, Inzucchi SE, Krumholz HM, Xiao L, Jones PG, Fiske S, Masoudi FA, Marso SP, Spertus JA: Glucometrics in patients hospitalized with acute myocardial infarction: defining the optimal outcomes-based measure of risk. Circulation 2008, 117: 1018-1027. 10.1161/CIRCULATIONAHA.107.740498PubMedView ArticleGoogle Scholar
  42. Oddo M, Schmidt JM, Carrera E, Badjatia N, Connolly ES, Presciutti M, Ostapkovich ND, Levine JM, Le Roux P, Mayer SA: Impact of tight glycemic control on cerebral glucose metabolism after severe brain injury: a microdialysis study. Crit Care Med 2008, 36: 3233-3238. 10.1097/CCM.0b013e31818f4026PubMedView ArticleGoogle Scholar
  43. van den Berghe G, Schoonheydt K, Becx P, Bruyninckx F, Wouters PJ: Insulin therapy protects the central and peripheral nervous system of intensive care patients. Neurology 2005, 64: 1348-1353. 10.1212/01.WNL.0000158442.08857.FCPubMedView ArticleGoogle Scholar
  44. Ichai C, Preiser JC: International recommendations for glucose control in adult non diabetic critically ill patients. Crit Care 2010, 14: R166.PubMedPubMed CentralGoogle Scholar
  45. Sacks DB, Bruns DE, Horton J, Lindberg S, Mahoney JJ, Manzella S, Reilly ET, Scott MG: POCT12-A3: Point-of-Care Blood Glucose Testing in Acute and Chronic Care Facilities; Approved Guideline. 3rd edition. Wayne, PA: Clinical and Laboratory Standards Institute; 2013.Google Scholar
  46. Khalil PN, Kleespies A, Angele MK, Thasler WE, Siebeck M, Bruns CJ, Mutschler W, Kanz KG: The formal requirements of algorithms and their implications in clinical medicine and quality management. Langenbecks Arch Surg 2011, 396: 31-40. 10.1007/s00423-010-0713-3PubMedView ArticleGoogle Scholar
  47. Meijering S, Corstjens AM, Tulleken JE, Meertens JH, Zijlstra JG, Ligtenberg JJ: Towards a feasible algorithm for tight glycaemic control in critically ill patients: a systematic review of the literature. Crit Care 2006, 10: R19. 10.1186/cc3981PubMedPubMed CentralView ArticleGoogle Scholar
  48. van den Berghe G, Wouters P, Weekers F, Verwaest C, Bruyninckx F, Schetz M, Vlasselaers D, Ferdinande P, Lauwers P, Bouillon R: Intensive insulin therapy in the critically ill patients. N Engl J Med 2001, 345: 1359-1367. 10.1056/NEJMoa011300PubMedView ArticleGoogle Scholar
  49. Finfer S, Chittock DR, Su SY, Blair D, Foster D, Dhingra V, Bellomo R, Cook D, Dodek P, Henderson WR, Hebert PC, Heritier S, Heyland DK, McArthur C, McDonald E, Mitchell I, Myburgh JA, Norton R, Potter J, Robinson BG, Ronco JJ: Intensive versus conventional glucose control in critically ill patients. N Engl J Med 2009, 360: 1283-1297.PubMedView ArticleGoogle Scholar
  50. Boord JB, Sharifi M, Greevy RA, Griffin MR, Lee VK, Webb TA, May ME, Waitman LR, May AK, Miller RA: Computer-based insulin infusion protocol improves glycemia control over manual protocol. J Am Med Inform Assoc 2007, 14: 278-287. 10.1197/jamia.M2292PubMedPubMed CentralView ArticleGoogle Scholar
  51. Dortch MJ, Mowery NT, Ozdas A, Dossett L, Cao H, Collier B, Holder G, Miller RA, May AK: A computerized insulin infusion titration protocol improves glucose control with less hypoglycemia compared to a manual titration protocol in a trauma intensive care unit. JPEN J Parenter Enteral Nutr 2008, 32: 18-27. 10.1177/014860710803200118PubMedView ArticleGoogle Scholar
  52. Crockett SE, Suarez-Cavelier J, Accola KD, Hadas LA, Harnage DL, Garrett PR, Butler KA, Mulla ZD: Risk of postoperative hypoglycemia in cardiovascular surgical patients receiving computer-based versus paper-based insulin therapy. Endocr Pract 2012, 18: 529-537. 10.4158/EP11337.ORPubMedView ArticleGoogle Scholar
  53. Hoekstra M, Vogelzang M, Verbitskiy E, Nijsten MW: Health technology assessment review: computerized glucose regulation in the intensive care unit - how to create artificial control. Crit Care 2009, 13: 223. 10.1186/cc8023PubMedPubMed CentralView ArticleGoogle Scholar
  54. Chase JG, Shaw G, Le Compte A, Lonergan T, Willacy M, Wong XW, Lin J, Lotz T, Lee D, Hann C: Implementation and evaluation of the SPRINT protocol for tight glycaemic control in critically ill patients: a clinical practice change. Crit Care 2008, 12: R49. 10.1186/cc6868PubMedPubMed CentralView ArticleGoogle Scholar
  55. Van Herpe T, Mesotten D, Wouters PJ, Herbots J, Voets E, Buyens J, De Moor B, van den Berghe G: LOGIC-insulin algorithm-guided versus nurse-directed blood glucose control during critical illness: the LOGIC-1 single-center, randomized, controlled clinical trial. Diabetes Care 2013, 36: 188-194. 10.2337/dc12-0584PubMedPubMed CentralView ArticleGoogle Scholar
  56. Rattan R, Nasraway SA: The future is now: software-guided intensive insulin therapy in the critically ill. J Diabetes Sci Technol 2013, 7: 548-554. 10.1177/193229681300700231PubMedPubMed CentralView ArticleGoogle Scholar
  57. Wilson M, Weinreb J, Hoo GW: Intensive insulin therapy in critical care: a review of 12 protocols. Diabetes Care 2007, 30: 1005-1011. 10.2337/dc06-1964PubMedView ArticleGoogle Scholar
  58. Hovorka R, Chassin LJ, Ellmerer M, Plank J, Wilinska ME: A simulation model of glucose regulation in the critically ill. Physiol Meas 2008, 29: 959-978. 10.1088/0967-3334/29/8/008PubMedView ArticleGoogle Scholar
  59. Boyd JC, Bruns DE: Monte Carlo simulation in establishing analytical quality requirements for clinical laboratory tests meeting clinical needs. Methods Enzymol 2009, 467: 411-433.PubMedView ArticleGoogle Scholar
  60. Van Herpe TV, Espinoza M, Haverbeke N, De Moor BD, van den Berghe G: Glycemia prediction in critically ill patients using an adaptive modeling approach. J Diabetes Sci Technol 2007, 1: 348-356. 10.1177/193229680700100306View ArticleGoogle Scholar
  61. Chase JG, Suhaimi F, Penning S, Preiser JC, Le Compte AJ, Lin J, Pretty CG, Shaw GM, Moorhead KT, Desaive T: Validation of a model-based virtual trials method for tight glycemic control in intensive care. Biomed Eng Online 2010, 9: 84. 10.1186/1475-925X-9-84PubMedPubMed CentralView ArticleGoogle Scholar
  62. Wilinska ME, Blaha J, Chassin LJ, Cordingley JJ, Dormand NC, Ellmerer M, Haluzik M, Plank J, Vlasselaers D, Wouters PJ, Hovorka R: Evaluating glycemic control algorithms by computer simulations. Diabetes Technol Ther 2011, 13: 713-722. 10.1089/dia.2011.0016PubMedView ArticleGoogle Scholar
  63. Wilinska ME, Nodale M: An evaluation of ‘I, Pancreas’ algorithm performance in silico. J Diabetes Sci Technol 2009, 3: 857-862. 10.1177/193229680900300433PubMedPubMed CentralView ArticleGoogle Scholar
  64. Leelarathna L, English SW, Thabit H, Caldwell K, Allen JM, Kumareswaran K, Wilinska ME, Nodale M, Mangat J, Evans ML, Burnstein R, Hovorka R: Feasibility of fully automated closed-loop glucose control using continuous subcutaneous glucose measurements in critical illness: a randomized controlled trial. Crit Care 2013, 17: R159. 10.1186/cc12838PubMedPubMed CentralView ArticleGoogle Scholar

Copyright

© Wernerman et al.; licensee BioMed Central Ltd. 2014

This article is published under license to BioMed Central Ltd. The licensee has exclusive rights to distribute this article, in any medium, for 12 months following its publication. After this time, the article is available under the terms of the Creative Commons Attribution License (http://creativecommons.org/licenses/by/4.0), which permits unrestricted use, distribution, and reproduction in any medium, provided the original work is properly cited. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.