Table 4 Calculation of the ∆WOB patient (%) according to two simulated inspiratory demands

From: A new reservoir-based CPAP with low oxygen consumption: the Bag-CPAP

 

Moderate inspiratory demand with Vt target of 300 ml

High inspiratory demand with Vt target of 500 ml

∆WOB patient (%)

∆WOB patient (%)

PB 980

0

0

AirSense CPAP

− 2

− 11

Boussignac CPAP

+ 35

+ 26

O-Two CPAP

+ 50

+ 38

Ventumask CPAP

+ 68

+ 96

Bag-CPAP with Venturia

+ 137

+ 18

+ 106

+ 18

Bag-CPAP with conventional connectorb

+ 92

+ 37

+ 71

+ 66

  1. aOxygen flow set at 5 L/min and 10 L/min
  2. bOxygen flow set at 15 L/min and 20 L/min
  3. ∆WOB patient represents the relative variation of the simulated patient’s work of breathing needed to maintain the tidal volume obtained with the ICU ventilator (PB 980). See Fig. 4 as well. Bag-CPAP was tested in different oxygen flow conditions, with the Venturi connector at 5 L/min and 10 L/min and with the conventional oxygen connector at 15 L/min and 20 L/min