Variable | Closed-loop oxygen control | Manual oxygen titration | Difference estimate (95% CI) or mean difference or OR | p value |
---|---|---|---|---|
Primary outcome | ||||
Time in optimal SpO2 range (% of the total recording time) | 96.5 (93.5 to 98.9) | 89.0 (77.4 to 95.9) | 10.40 (5.20 to 17.20) | < 0.0001 |
Secondary outcomes | ||||
Time in suboptimal SpO2 range (% of the total recording time) | ||||
Low | 0.6 (0.1 to 1.1) | 1.0 (0.1 to 4.0) | − 1.20 (− 4.95 to − 0.25) | 0.0053 |
High | 2.3 (0.8 to 5.2) | 4.2 (1.1 to 16.4) | − 5.25 (− 14.45 to − 1.40) | 0.0004 |
Time out of SpO2 range (% of the total recording time) | ||||
Low | 0 (0 to 0) | 0 (0 to 0.1) | − 0.10 (− 0.35 to 0.10) | 0.3972 |
High | 0 (0 to 0.1) | 0 (0 to 0.3) | − 0.50 (− 2.15 to − 0.05) | 0.0166 |
Percentage of time with SpO2 signal available | 99.7 | 99.8 | − 0.10 (− 0.35 to 0.30) | 0.6235 |
Mean SpO2/FIO2 during the recording time | 206 (176 to 254) | 210 (165 to 235) | 8.21 (− 7.08 to 23.38) | 0.2715 |
ROX index | ||||
At the beginning | 10.30 (3.22) | 10.30 (2.87) | Mean difference 0.009 (− 0.83 to 0.85) | 0.9829 |
At the end | 11.50 (4.48) | 10.60 (3.52) | Mean difference 0.88 (− 0.46 to 2.23) | 0.1916 |
ΔROXa | 1.20 (4.20) | 0.33 (2.37) | Mean difference 0.87 (− 0.77 to 2.52) | 0.2917 |
Time with SpO2 < 88% (% of the total recording time) | 0 (0 to 0) | 0 (0 to 0) | − 0.10 (− 0.20 to 0.05) | 0.2470 |
Time with SpO2 < 85% (% of the total recording time) | 0 (0 to 0) | 0 (0 to 0) | − 0.10 (− 0.55 to 0.10) | 0.3741 |
Number of events (SpO2 < 88%) per hour | 0 (0 to 0) | 0 (0 to 0) | 0.05 (− 0.38 to 0.38) | 0.9055 |
Number of events (SpO2 < 85%) per hour | 0 (0 to 0) | 0 (0 to 0) | − 0.25 (− 0.51 to 0.25) | 0.5862 |
Mean FIO2 during the recording time | 0.5 (38.1 to 55.8) | 0.5 (40.0 to 56.6) | 0.75 (− 2.80 to 4.55) | 0.6155 |
Time in different FIO2 (% of the total recording time) | ||||
FIO2 < 0.4 | 21.7 (0 to 66.0) | 0 (0 to 0) | 32.29 (12.05 to 50.30) | 0.0025 |
FIO2 0.4–0.6 | 46.0 (10.3 to 70.2) | 100 (63.2 to 100) | − 41.08 (− 54.80 to − 23.80) | < 0.0001 |
FIO2 > 0.6 | 4.2 (0.0 to 29.8) | 0.0 (0.0 to 0.0) | 16.43 (6.40 to 35.20) | < 0.0001 |
Number of manual adjustments per hour | 0 (0 to 0) | 0.21 (0 to 0.47) | − 0.49 (− 0.72 to − 0.36) | < 0.0001 |
Patients with ≥ 1 alarms/h for low SpO2 | 16 (35.6%) | 22 (48.9%) | OR 0.57 (0.25–1.34) | 0.2003 |
Patients with ≥ 1 alarms/h for high SpO2 | 1 (2.2%) | 1 (2.2%) | OR 1.00 (0.06–16.49) | 1.0000 |
ΔComfort scorea | 0 (0 to 1.0) | 0 (− 0.25 to 0.125) | 1.0 (0 to 1.50) | 0.0468 |