Caution-closed suction systems!

Context

Closed suction systems have become popular in the airway toilet of critically ill patients who are intubated because they may reduce cross infection and prevent lung collapse and desaturation. The aim of this study was to examine their effect on ventilation using a plastic model of the lungs and trachea.

Significant findings

Open suctioning through a 7 mm endotracheal tube (ETT) resulted in alveolar pressures of -7 cm H20 with a 12 Fr catheter. This fell to -24 cm H2O if a 14 Fr catheter was used. During volume controlled modes, intrinsic positive end-expiratory pressure (PEEP) developed at levels up to 22 cm H2O greater than externally applied PEEP during suction through a 7 mm tracheal tube with a 14 Fr catheter. Peak inspiratory pressures rose to as high as 50 cm H2O.

Pressure controlled ventilation by contrast was much safer. Intrinsic PEEP levels of just 5 cm H2O greater than PEEP were generated, associated with a reduction in alveolar ventilation. Continuous positive airway pressure (CPAP) appeared best with minimal changes in alveolar pressure during closed system suctioning.

Comments

A vital aspect of protective lung ventilation as advocated in the ARDSnet study is alveolar recruitment and the maintenance of PEEP. Contrary to perceived wisdom closed suction systems may actually cause more marked swings in alveolar pressures, particularly during volume controlled modes of ventilation. Further comparisons between closed and open suction catheter systems are therefore necessary to determine the overall effect of closed suction systems with a variety of ventilatory modes from alternative ventilators.

Methods

Lung model with rigid plastic trachea, ventilation with a Servo, alveolar and tracheal pressures were measured

Additional information

References

  1. Stenqvist O, Lindgren S, Karason S, Sondergaard S, Lundin S: Warning! Suctioning. A lung model evaluation of closed suctioning systems. Acta Anaesthiol Scand. 2001, 45: 167-172.

    Article  CAS  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

Reprints and permissions

About this article

Cite this article

Bewley, J. Caution-closed suction systems!. Crit Care 5, 73200 (2001). https://doi.org/10.1186/ccf-2001-73200

Download citation

  • Received:

  • Published:

  • DOI: https://doi.org/10.1186/ccf-2001-73200

Keywords