Comparison of the Mapleson C circuit, 500 ml and 1.6 l self-inflating bags for delivering guideline-compliant ventilation during simulated adult cardiopulmonary resuscitation

Introduction

Despite all the research and education that has gone into the field of cardiopulmonary resuscitation (CPR), survival rates remain bleak. A significant problem has been the discrepancy between teachings and witnessed clinical practice. As a result of this, and the deleterious outcomes associated with hyperventilation, we conducted a manikin-based study to evaluate three different ventilating devices and their ability to provide guideline-compliant ventilation during simulated adult CPR.

Methods

A simulated cardiac arrest scenario was undertaken by 33 healthcare professionals (α = 0.05, power = 80%). Participants were asked to ventilate a simulated cardiac arrest patient for a period of 1 minute with all three devices, during which time various ventilatory parameters were recorded using a spirometer. The devices investigated were the Mapleson C circuit, adult (1.6 l) and paediatric (500 ml) self-inflating bags. P < 0.01 was deemed statistical significant, due to multiple comparisons.

Results

The paediatric self-inflating bag performed best, with significant improvement in the mean minute ventilation (P = 0.003), tidal volume (P < 0.001) and peak airway pressure (P < 0.001). Despite the significant differences, the paediatric self-inflating bag still delivered a mean minute ventilation of 7.01 l/minute, which still exceeds the Resuscitation Council's suggested 5 l/minute. See Table 1.

Table 1 Comparison of data on ventilation parameters

Conclusions

Participants were found to be hyperventilating simulated cardiac arrest patients with all devices. The paediatric self-inflating bags delivered the most guideline-compliant ventilation and its use in adult CPR may be a simple measure to ensure delivery of more guideline-consistent ventilation.

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Sherren, P., Lewinsohn, A., Jovaisa, T. et al. Comparison of the Mapleson C circuit, 500 ml and 1.6 l self-inflating bags for delivering guideline-compliant ventilation during simulated adult cardiopulmonary resuscitation. Crit Care 15 (Suppl 1), P290 (2011). https://doi.org/10.1186/cc9710

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc9710

Keywords