Introduction
Implementation of assisted mechanical ventilation in acute lung injury (ALI) patients may decrease ventilator-induced lung injury by redistribution of tidal ventilation towards dependent lung regions. Up to now, tidal ventilation regional distribution has been measured by expensive and complicated methods, not readily available at the bedside. Electrical impedance tomography (EIT) is a relatively new non-invasive bedside method to monitor tidal ventilation distribution, validated in preclinical studies. We verified the feasibility of using EIT to monitor tidal ventilation regional distribution in patients undergoing assisted ventilation and we describe the effect of different pressure support levels on regional ventilation.
Methods
We enrolled 11 consecutive ALI patients admitted to our ICU, intubated and undergoing pressure support (PS) ventilation. We monitored the regional tidal ventilation distribution by means of a new EIT monitor (PulmoVista 500®; Dräger Medical GmbH, Lübeck, Germany), dividing the lung imaging field into four contiguous same-size regions of interest (ROI): ventral right (ROI 1) and left (ROI 2), dorsal right (ROI 3) and left (ROI 4). We randomly performed two steps of PS ventilation for 15 minutes, leaving the positive end-expiratory pressure (PEEP) and FiO2 unchanged: PSlow (p0.1 ≥2 cmH2O) and PShigh (p0.1 <2 cmH2O). At the end of each step, we recorded: ventilation parameters, arterial blood gas analysis and percentage of tidal ventilation distribution in the four ROIs. Analyses were performed by paired t test.
Results
The ALI etiology was: trauma (18%), septic shock (18%), pneumonia (46%) and postoperative respiratory failure (18%). PSlow was set at 3 ± 2 cmH2O and PShigh at 12 ± 3 cmH2O. An increase in PS level determined a significant increase of tidal volume (7 ± 2 vs. 9 ± 3 ml/kg, P = 0.003) and peak inspiratory pressure (12 ± 4 vs. 18 ± 4 cmH2O, P = 0.0001). At PShigh, proportional distribution of tidal ventilation significantly changed in all four ROIs (ROIs 1 to 4: 25 ± 9 vs. 33 ± 10%, P = 0.0003; 32 ± 13 vs. 37 ± 12%, P = 0.02; 20 ± 8 vs. 14 ± 8%, P = 0.0008; 23 ± 8 vs. 16 ± 5%, P = 0.005), moving from dorsal to ventral.
Conclusions
EIT may be a useful tool to monitor lung regional ventilation at the bedside. PS levels that blunt patient effort may promote redistribution of tidal ventilation towards ventral lung regions.
Author information
Authors and Affiliations
Rights and permissions
This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Mauri, T., Bellani, G., Confalonieri, A. et al. Electrical impedance tomography to monitor regional tidal ventilation at different pressure support levels. Crit Care 15 (Suppl 1), P146 (2011). https://doi.org/10.1186/cc9566
Published:
DOI: https://doi.org/10.1186/cc9566
Keywords
- Acute Lung Injury
- Electrical Impedance Tomography
- Pressure Support
- Lung Region
- Pressure Support Ventilation