Artifactual lactic acidosis from point of care analysis: mind the gap
Critical Care volume 14, Article number: P163 (2010)
Introduction
A patient presented with severe acidosis, point-of-care (POC) lactate of 42 mmol/l. This led to suspicion of mesenteric ischemia and potential need for laparotomy. However, plasma lactates were <4 mmol/and ethylene glycol (EG) ingestion was subsequently diagnosed. We therefore wished to determine why discrepant lactates occur and whether this lactate gap (that is, the difference between lactate measured using two common methods) could be clinically useful.
Methods
We phlebotomized blood, added various concentrations of EG metabolites, and tested with the five most common lactate analyzers.
Results
Glycolate caused massive artifactual lactate elevation with the Radiometer 700 POC analyzer even at low concentrations. Another major EG metabolite, glyoxylate, but not oxalate (or formate), caused similar elevation. The iSTAT and Bayer POC analyzers and Beckman and Vitros laboratory analyzers caused minimal lactate elevations. The lactate gap was determined by comparing the Radiometer with any of the other analyzers.
Conclusions
We demonstrated how inappropriate laparotomy or delayed-therapy might occur if clinicians are unaware of this phenomenon (now dubbed the lactate gap) or if hospitals only have access to a single analyzer. We also showed how the lactate gap can be exploited: to expedite treatment, diagnose late EG ingestion, and terminate dialysis. By comparing lactate results from the iSTAT or Bayer versus the Radiometer, EG ingestion can be diagnosed at POC, thereby further expediting diagnosis and treatment ahead of plasma EG.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Brindley, D., Brindley, P. Artifactual lactic acidosis from point of care analysis: mind the gap. Crit Care 14 (Suppl 1), P163 (2010). https://doi.org/10.1186/cc8395
Published:
DOI: https://doi.org/10.1186/cc8395