From: Comparison between Flotrac-Vigileo and Bioreactance, a totally noninvasive method for cardiac output monitoring
Systolic blood pressure
Bias
Relative error
> 160 mmHg
2.2 ± 1.5 L/min
51 ± 36%
120 to 160 mmHg
0.6 ± 1.2 L/min.
15 ± 27%
80 to 120 mmHg
-0.1 ± 1.0 L/min
-0.0 ± 21%
< 80 mmHg
-0.9 ± 1.3 L/min
-18 ± 25%