Parameter | Initiation settings |
---|---|
FiO2 | Same FiO2 that patient was receiving during conventional ventilation, adjust to SpO2 > 90% |
Mean airway pressure (mPaw) | 3–5 cmH2O higher than patient was receiving during conventional ventilation, titrate upward to reduce FiO2 below 0.6 |
Bias flow | 40 l/min, titrate to exceed any spontaneous inspiratory efforts |
Pressure amplitude of oscillation ('power' or ΔP) | Titrate to produce a 'wiggle', or body movement, from shoulders to midthigh |
Frequency | 5 Hz, then titrate to PaCO2 |
Percentage inspiratory time | 33% |