Effects of mask-ventilator interface elements in a home noninvasive portable ventilator. Study in cold hypercapnic patients
Critical Care volume 3, Article number: P22 (2000)
Noninvasive mechanical ventilation (NIMV) in hypercapnic COLD exacerbations with a home portable ventilator with a single ventilatory tube has some technical considerations.
Objective
We analyze influence of elements used in the mask-ventilator interface in hypercapnic COLD with a NMV: To compare differents 1). Design of filters (A, B, C, D), 2) Non-rebreathing expiratory valve: a) Swisper and b) Plateau exhalation valve. 3) Rramp inspiratory time: (0,05, 0.1, 0.5 seg) affect a predetermined level inspiratory positive pressure (IPAP) (15 cmH2O). 2) Hypercapnia (PCO2) control and 3) Subjective responses: a) Dysnea Brog index (low: 2 to 10 high) and psychological dependence at different stages of therapy (low 2 to high 10).
Setting
ICU.
Subjects
Twelve hypercapnic COLD stable patients.
Material
BiPAP ST-D (Resp, Inc). Facial mask.
Results
See Table.
Conclusion
Subjective (dysnea Brog index), objective respiratory response (hypercapnia) and level of IPAP pressure applied during