Dead space free flow sensor reduces the ventilation effort in very low birth weight infants: an experimental study
Critical Care volume 8, Article number: P13 (2004)
Introduction
Bronchopulmonary dysplasia is a serious side effect of long-term ventilation in very low birth-weight infants, high tidal volumes being the main cause. However, technical dead space represents the biggest part of the necessary tidal volume. Therefore we studied whether reducing dead space could also reduce the ventilation effort.
Methods
The time needed to eliminate CO2 by mechanical ventilation from a 'preterm infant test lung' flooded with CO2 was measured. Four different Y-pieces (three with flow sensor: integrated [A], small dead space [B], dead-space free (our patent) [C] and one without flow sensor [D]) either connected or not connected to an enclosed suction device were used. Mechanical ventilation was standardized using a ventilation rate of 60/min, a PIP of 16 mbar, a PEEP of 3.5 mbar and a flow of 10 l/min. A capnograph in the expiration-part of the ventilation tube measured CO2 concentrations. The test series were repeated 10 times and the mean (± SD) was used for statistical analysis.
Results
CO2 elimination time was: in A, 47.7 (± 0.82) s with enclosed suctioning device, without 45.5 (± 1.18) s; in B, 42.5 (± 1.27) and 41.1 (± 0.99) s; in C, 37.5 (± 1.18) and 37.4 (± 0.97) s; and in D, 38.3 (± 1.16) s and 36.8 (± 0.79) s, respectively.
Conclusion
The use of the Y-piece with the patented dead space free flow sensor shows the same CO2 elimination time as the Y-piece without flow sensor. This reduced the ventilation effort by 20% compared with the Y-pieces with integrated flow sensor and by 10% with the flow sensor with small dead space. This might also reduce the incidence of bronchopulmonary dysplasia.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Wald, M., Jeitler, V., Lawrenz, K. et al. Dead space free flow sensor reduces the ventilation effort in very low birth weight infants: an experimental study. Crit Care 8 (Suppl 1), P13 (2004). https://doi.org/10.1186/cc2480
Published:
DOI: https://doi.org/10.1186/cc2480
Keywords
- Mechanical Ventilation
- Preterm Infant
- Tidal Volume
- Dead Space
- Ventilation Rate