Skip to main content
  • Meeting abstract
  • Published:

Complications of monitoring of jugular bulb venous saturation

Introduction

Neurological monitoring is very important to change the prognosis of a critical neurological patient. The jugular venous bulb saturation is extremely important to evaluate the consumption and delivery of oxygen.

Objective

To describe complications during insertion and the permanence period of a jugular venous bulb catheter.

Materials and methods

A prospective, observational study of 21 patients from June 2000 to September 2002 in an intensive care unit. All patients were monitored with an intracranial pressure device. The jugular venous line was cannulated independent of which side. The catheter flow was sustained by continuous saline infusion (rate, 3 ml/hour). The monitor used was a VIGILANCE (Baxter). Complications observed were: during insertion, arterial puncture, bleeding, and misplacement; during catheter permanence, obstruction and infection (daily examination); and after decannulation, thrombosis detected through Doppler examination, which was performed after 24 hours. All the catheter tips were sent for bacteriological examination.

Results

The mean time of cannulation was 5 days. The thrombosis rate detected by Doppler examination was 31.6% (without clinical compromise). The catheter obstruction rate was 15.8% and the infection rate was 10.5%.

Conclusion

Strict control with Doppler examination is very important to warrant optimal flow. The catheter must be changed every 5 days in order to avoid infection.

Authors

Rights and permissions

Reprints and permissions

About this article

Cite this article

Rodrigues, M., Salgado, D., Resende, V. et al. Complications of monitoring of jugular bulb venous saturation. Crit Care 7 (Suppl 3), P79 (2003). https://doi.org/10.1186/cc2275