From: Clinical evaluation of the Life Support for Trauma and Transport (LSTATTM) platform
Recovery room survey questionnaire | Yes | No | Abstained |
---|---|---|---|
LSTAT for patient care in the postanesthetic care unit | Â | Â | Â |
   Did the monitoring equipment in the LSTAT allow proper management of the patient? | 9/9 (100%) | 0/9 (0%) | 0/9 (0%) |
   Were the LSTAT controls accessible and easy to operate? | 5/9 (56%) | 1/9 (44%) | 0/9 (0%) |
   Did you find the table of vital signs useful for filling out your nursing record? | 7/9 (78%) | 1/9 (11%) | 1/9 (11%) |
Critical events | Â | Â | Â |
   In the recovery room, were critical changes in vital signs detected in a timely manner using the LSTAT? | 7/9 (78%) | 0/9 (0%) | 2/9 (22%) |
   Were there any limitations in the LSTAT equipment that prevented you from detecting critical changes in vital signs and adequately addressing them? | 3/9 (33%) | 5/9 (56%) | 1/9 (11%) |
   Were there any features of the LSTAT that improved your ability to detect critical changes in vital signs and address them? | 3/9 (33%) | 5/9 (56%) | 1/9 (11%) |
   If no other medical equipment were available and you were called upon to evaluate and resuscitate a patient with the LSTAT in a remote setting, do you feel it would be sufficient? | 8/9 (89%) | 1/9 (11%) | 0/9 (0%) |
Recovery room to ambulatory surgery discharge area or hospital bed transports | Â | Â | Â |
   Did you notice any particular advantage of the LSTAT during intra-hospital transports? | 2/9 (22%) | 5/9 (56%) | 2/9 (22%) |
   Was the location of the patient connectors and lines allow for unobstructed access to the patient, controls and displays? | 7/9 (78%) | 2/9 (22%) | 0/9 (0%) |
   Was the hand held display useful in monitoring vital signs during transport? | 5/9 (56%) | 2/9 (22%) | 2/9 (22%) |
   Were the patient's vital signs easy to monitor during transport? | 6/9 (67%) | 3/9 (33%) | 0/9 (0%) |