Influence of the oscillometric calibration method on accuracy and precision of continuous non-invasive arterial pressure measurements using the CNAP™ device

  • 494 Accesses

Introduction

The CNAP™ system (CNSystems Medizintechnik AG, Graz, Austria) provides continuous non-invasive arterial pressure (AP) measurements based on the volume clamp method using a finger cuff. Finger AP values are calibrated to oscillometric upper-arm AP measurements. In the present study we investigated the influence of the calibration approach based on oscillometric upper-arm cuff measurements on the accuracy and precision of the CNAP™ device in comparison with invasively obtained AP measurements.

Methods

The datasets of simultaneously recorded invasive (via arterial catheter) and non-invasive (using the CNAP™ system) AP measurements in 43 patients treated in the medical ICU of a university hospital were analyzed in this study. The following comparative analyses between the two AP measurement techniques were performed: (1) comparison of CNAP™-derived AP values with invasive AP (IAP) measurements; (2) comparison of the CNAP™ oscillometric AP values used for the calibration of finger AP with IAP measurements; and (3) computer- aided calibration (CAC) of the CNAP™ finger AP values to IAP values instead of calibration to oscillometric upper-arm AP measurements with IAP measurements.

Results

(1) The comparison of CNAP™-derived AP values with IAP measurements revealed a mean difference (± standard deviation) for mean AP, systolic AP, and diastolic AP of +0.6 mmHg (±10 mmHg), + 11 mmHg (±17 mmHg), and -6 mmHg (±9 mmHg), respectively. (2) The comparison between the oscillometric AP values used for calibration of the CNAPTM device and the corresponding IAP values resulted in a mean difference (± standard deviation) of -0.8 mmHg (±8 mmHg), -5 mmHg (±14 mmHg), and +10 mmHg (±9 mmHg), respectively. (3) CAC of the CNAPTM finger AP values to IAP values instead of calibration to oscillometric upper-arm AP measurements resulted in a mean difference (± standard deviation) of +4 mmHg (±8 mmHg), +5.5 mmHg (±14 mmHg), and +3 mmHg (±7 mmHg), respectively. The accuracy of CAC-CNAPTM-derived systolic and diastolic AP compared with the CNAPTM-derived AP calibrated to oscillometric AP was significantly higher (P = 0.004 and P < 0.001, respectively).

Conclusion

When using the CNAPTM system, calibration to oscillometric upper-arm AP values integrated into the CNAPTM system is a relevant source of difference between CNAPTM-derived continuous non-invasive AP measurements and invasively assessed AP values.

Author information

Correspondence to JY Wagner.

Rights and permissions

Reprints and Permissions

About this article

Cite this article

Wagner, J., Negulescu, I., Schofthaler, M. et al. Influence of the oscillometric calibration method on accuracy and precision of continuous non-invasive arterial pressure measurements using the CNAP™ device. Crit Care 18, P121 (2014). https://doi.org/10.1186/cc13311

Download citation

Keywords

  • Arterial Pressure
  • Calibration Method
  • Arterial Catheter
  • Systolic Arterial Pressure
  • Relevant Source