Table 4 Comparison of quality of chest compression measures in the manual and load-distributing band (LDB) phases

From: Improved neurologically intact survival with the use of an automated, load-distributing band chest compression device for cardiac arrest presenting to the emergency department

Variables measured

Manual CPR

(n = 29)

LDB CPR

(n = 151)

Mean difference

95% CI for differencee

Median NFT 0 to 5 minutes (IQR), sec

84 (43-107)

114 (75-164)

-40

(-66 to -14)

Mean NFR 0 to 5 minutes (SD)

0.29 (0.17)

0.42 (0.38)

-0.13

(-0.22 to -0.05)

Median NFT 5 to 10 minutes (IQR), sec

76 (50-137)

53 (37-97)

25

(-3 to 52)

Mean NFR 5 to 10 minutes (SD)

0.33 (0.213)

0.24 (0.194)

0.09

(-0.01 to 0.18)

Mean CPR ratio (SD)

36 (21)

50 (28)

-14

(-24 to -3)

Mean compression ratio (SD)

34 (19)

46 (26)

-13

(-22 to -3)

Mean compression rate/minute (SD)

126 (14)

104 (30)

22

(11 to 33)

Mean compression/min (SD)

40 (23)

46 (28)

-6

(-17 to 5)

Mean ventilation rate/minute 0 to 5 minutes (SD)

13 (4)

8 (4)

4

(2 to 6)

Mean ventilation/minute 0 to 5 minutes (SD)

12 (4)

9 (4)

3

(2 to 5)

Mean ventilation rate/minute 5 to 10 minutes (SD)

13 (5)

7 (3)

5

(4 to 7)

Mean ventilation/minute 5 to 10 minutes (SD)

12 (5)

8 (3)

5

(3 to 6)

  1. For elapsed time variables, median (inter-quartile range) is reported. For other variables, mean (SD) is reported.