Techniques to measure cardiac output: minimally invasive method versus thermodilution
Critical Care volume 16, Article number: P216 (2012)
Introduction
Hemodynamic monitoring is important to manage critically ill patients. The thermodilution pulmonary catheter is considered the gold standard; however, it is invasive and associated with the onset of complications. Our study compared cardiac output (CO) obtained with the MostCare (COMC), which uses the pressure recording analytical method, to CO obtained with a Swan-Ganz (COSG) catheter in hemodynamically unstable patients.
Methods
We conducted a prospective clinical study in our cardiosurgical ICU. Sixteen post-cardiosurgical adult patients were enrolled. They had a Swan-Ganz catheter and were mechanically ventilated. The Swan-Ganz catheter was connected to the monitor Vigilance Edwards®, while the MostCare was connected to the patient's artery. For each patient three measurements of CO have been carried out and the mean was considered for statistical analysis. The correlation coefficient, Bland-Altman test and percentage of error were measured.
Results
The correlation coefficient between COSG and COMC was 0.824 (0.567 to 0.935, 95% CI; P < 0.001) The Bland-Altman analysis showed a mean difference between the two methods (bias) of 0.22 ± 0.55 l/minute/m2 with lower and upper 95% limits of confidence of -0.87 and 1.30 l/minute/m2 respectively. The percentage of error was of 25%.
Conclusion
This study demonstrated a good correlation between the two methods. MostCare is resulted to be reliable and accurate even in hemodynamically unstable patients. It would be interesting to study the new device before and after having modified the therapy, such as fluid challenge or inotropic therapy or the use of vasopressors.
References
Zangrillo A, et al.: J Cardiothorac Vasc Anesth. 2010, 24: 265-269. 10.1053/j.jvca.2009.09.019
Scolletta S, et al.: Br J Anaesth. 2005, 95: 159-165. 10.1093/bja/aei154
Romano SM, et al.: Crit Care Med. 2002, 30: 1834-1841. 10.1097/00003246-200208000-00027
Author information
Authors and Affiliations
Rights and permissions
This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Donati, A., Tondi, S., Carsetti, A. et al. Techniques to measure cardiac output: minimally invasive method versus thermodilution. Crit Care 16 (Suppl 1), P216 (2012). https://doi.org/10.1186/cc10823
Published:
DOI: https://doi.org/10.1186/cc10823