- Poster presentation
- Open Access
- Published:
Errors in the arterial blood pressure measurement
Critical Care volume 16, Article number: P212 (2012)
Introduction
The artefacts affecting arterial wave morphology may compromise recorded values of arterial blood pressure (ABP) and can lead to therapeutic errors. The aim of this study is to evaluate the errors between invasive and noninvasive arterial pressure values, the incidence of artefacts due to an inadequate dynamic response of the transducer-tubing system, and their detection by the ICU staff.
Methods
Seventy-five consecutive patients (50 male, mean age 55 ± 18) admitted to the ICU for heterogeneous pathologies were enrolled. Inclusion criteria were: the presence of an intra-arterial catheter (IAC) for invasive blood pressure monitoring, and age >18 years. Pregnancy was excluded. At admission and every time the IAC was replaced we acquired invasive systolic, diastolic, and medium arterial pressure values (I-SP, I-DP, I-MP) during hemodynamic stability (variations of mean arterial pressure <10%); at the same time, noninvasive systolic and diastolic arterial pressure values (Ni-SP, Ni-DP) were measured with a sphygmomanometer at the same arm of the IAC. Noninvasive medium arterial pressure (Ni-MP) was calculated as follows: (SP + 2DP)/3. At every time of the study, before ABP value acquisition, medical and nursing staff answered a questionnaire on the reliability of the arterial waveform. The staff could perform the fast flush test if considered appropriate. However, the fast flush test was executed by the main investigator at the end of questionnaire in all patients. Bland-Altman analysis was performed.
Results
We compared 130 pairs of Ni-SP, Ni-DP and Ni-MP and I-SP, I-DP and I-MP. The mean bias between Ni-SP and I-SP was -11 mmHg (limit of agreement (LoA) -43.6 to 21.4 mmHg). The mean bias between Ni-DP and I-DP and between Ni-MP and I-MP was 6.1 mmHg (LoA -15.5 to 27.7 mmHg) and 0.37 mmHg (LoA -21.0 to 21.7 mmHg), respectively. We performed the fast flush test 130 times; an inadequate dynamic response of the transducer-tubing system was observed 55 times: in 45 cases the arterial signal was underdumped and in 10 cases was overdumped. The arterial dumping was correctly detected by the medical staff in 95% of cases, by nursing staff and postgraduates in 35% of cases.
Conclusion
The bias between invasive and noninvasive ABP measure can be relevant and mislead in the therapeutic management. These errors can be avoided by identifying the artefacts that affect arterial signal and so the ICU staff must pay attention to the recognition of arterial dumping in critically ill patients.
References
Pickering TG: Principles and techniques of blood pressure measurement. Cardiol Clin 2002, 20: 207-223. 10.1016/S0733-8651(01)00009-1
Author information
Authors and Affiliations
Rights and permissions
This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Franchi, F., De Palo, V., Faltoni, A. et al. Errors in the arterial blood pressure measurement. Crit Care 16 (Suppl 1), P212 (2012). https://doi.org/10.1186/cc10819
Published:
DOI: https://doi.org/10.1186/cc10819
Keywords
- Arterial Blood Pressure
- Diastolic Arterial Pressure
- Invasive Blood Pressure
- Arterial Waveform
- Arterial Signal